This Information Applies To: Agilent Spectroscopy Systems, SPS 4 Autosampler
Issue
The sample probe may be displaced from its correct position due to the probe hitting a sample vial or the sides of the wash reservoir. User calibration allows for minor adjustments to center a misaligned probe using...
This Information Applies To: Agilent 5800/5900 ICP-OES
Issue:
The instrument requires clean, dry, noncorrosive air for cooling purposes. The air enters the instrument through an air supply vent located beside the chimney on the top of the instrument. The vent is fitted with a dust filter to screen...
This Information Applies To: Agilent 700 Series ICP-OES instruments
Issue:
It is important that a torch alignment is performed to ensure that the optics are optimized and are focused on the analytical zone of the plasma. This alignment procedure will ensure maximum sensitivity and signal-to-noise...
This Information Applies To: SPS 4/SPS 3 autosampler, 5000 series (5100/5110/5800/5900) ICP-OES
Issue:
How to change the autosampler probe due to maintenance, blockage or damage.
Steps to follow:
Warning : Always ensure that the power to the autosampler is turned off before installing...
This Information Applies To: Agilent 5100, 5110, 5800, 5900 ICP-OES
Issue
The axial pre-optic window needs checking from time to time to ensure it is clean.
If the axial pre-optic window is dirty, it must be cleaned or replaced to ensure adequate sensitivity in the UV region of the emission spectrum...