This Information Applies To: Agilent SIPS 10/20 and Flame AA
Issue
Replacing and conditioning the SIPS pump tubing.
Background
Whenever new pump tubes have been installed, it is necessary to condition the tubing before the analysis to prevent signal drift as the tubing settles during operation...
This Information Applies To: Agilent SIPS 10/20
Issue
Operational tips for the Sample Introduction Pump System (SIPS) 10/20.
Background
SIPS is an integrated, efficient dilution system for flame AA and is suitable for use with most Agilent (flame) atomic absorption spectrometers. See Figure...
This Information Applies To: Agilent GTA 120 and Zeeman Atomic Absorption instruments
Issue
How to replace and align graphite tubes. Replacement of graphite tubes are a routine maintenance item for all Graphite Furnace instruments.
Items Required
Graphite tubes suitable for your application...
This Information Applies To: Agilent GTA 120, Agilent 240Z AA Spectrometer, Agilent 280Z AA Spectrometer, Agilent 240FS/GTA120/PSD120/UltrAA System
Issue:
It is important to align the sample probe of the Programmable Sampler Dispenser (PSD) 120, so that the sample is injected properly into the...
This Information Applies To: Agilent atomic absorption systems (AA 240/280/55B/240Z/280Z)
Issue
Replacing hollow cathode lamps is a common maintenance procedure and can be used during troubleshooting to isolate the source of an error.
Steps to Follow
If you are developing a method using an...