Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
Atomic Spectroscopy
  • Technical Areas
  • More
Atomic Spectroscopy
Wiki Long-term shutdown of Vista/700 Series ICPOES instruments and Eventual Start-up
  • Announcements
  • Forum
  • Files
  • Videos
  • Wiki
  • More
  • Cancel
  • New
  • Atomic Spectroscopy
  • Alignment of the Sample Probe - PSD 120
  • How to Resolve SpectrAA Software Issues due to Adobe Flash Failure
  • Basic Identification of Gas Box Type in AAS instrument for New Customers
  • Collection of AAS & MP-AES Resources
  • Collection of ICP-OES Resources
  • Collection of ICPMS Resources
  • Explore Atomic Spectroscopy eLearning Courses at Agilent University
  • ICP-MS Active Cleaning
  • ICP-OES: How to Clean a Blocked Nebulizer
  • ICP-OES:  Checking for a Blocked Nebulizer
  • ICPMSD Cone Conditioning Procedure
  • Instructions for Long-term shut down of ICP-MS instruments
  • Long-term shutdown of Vista/700 Series ICPOES instruments and Eventual Start-up
  • Maintaining the Axial Pre-Optic Window
  • Tips and Tricks for ICP-MS General Applications
  • Troubleshooting and Maintenance of AA Systems: Part 1 - Hollow Cathode Lamps
  • Troubleshooting and Maintenance of AA Systems: Part 2A- Cleaning Flame AA Components
  • Troubleshooting and Maintenance of AA Systems: Part 2B- Flame Optimization
  • Troubleshooting Failed Plasma Ignition for Agilent 5000 Series ICP-OES
  • Troubleshooting ICP-OES and ICP-MS Chillers (G3292A and G8481A)

Long-term shutdown of Vista/700 Series ICPOES instruments and Eventual Start-up

Shutting Down Vista/700 Series Instrument for a Lengthy Period of Time:

  1. Turn the peristaltic pump on, and pump all liquid from the sample lines, nebulizer and spray chamber. Ensure you pump all the liquid out of the sample introduction components.
  2. Turn off the heat exchanger/chiller (water supply), then wait a few minutes.
  3. Turn off argon supply.
  4. Close the software and shut the instrument PC down.
  5. Switch off the power to the instrument (front left power switch).

 

Starting Up Vista/700 Series Instruments after a Lengthy Period of Shut-down:

  1. Switch the instrument power on (front left switch).
  2. Open the software.
  3. Turn on the argon, then wait a few minutes. (When argon is on, the minimum purge of 0.7L/min will be in effect while software is open. If needing to analyze wavelengths less than 190nm, turn the polyboost purge on in the software.)
  4. Turn on the heat exchanger/chiller (water supply).

 

Notes: When you repower the unit on, the optics will be at room temperature.  Turn the unit power on and allow the optics to warm for 1 day, prior to starting an analysis.

  • shutdown method
  • ICP-OES
  • covid-19
  • icpoes chiller
  • Share
  • History
  • More
  • Cancel
Anonymous
Parents
  • arjun
    arjun 11 months ago

    very useful for customer during this COVID-19 Period.

    • Cancel
    • Up 0 Down
    • Reply
    • More
    • Cancel
Comment
  • arjun
    arjun 11 months ago

    very useful for customer during this COVID-19 Period.

    • Cancel
    • Up 0 Down
    • Reply
    • More
    • Cancel
Children
No Data

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel
Related
Recommended
Privacy Statement
Terms of Use
Contact Us
Site Help