Midway through the analysis, the argon ran out, and error code 11011: Vacuum (IF/BK) time out, vacuum too high to start turbo pump, appeared. I've attempted a full system restart, but the same error persists
Midway through the analysis, the argon ran out, and error code 11011: Vacuum (IF/BK) time out, vacuum too high to start turbo pump, appeared. I've attempted a full system restart, but the same error persists
Hello! There is the Error 1101... It could be probably generated due to an incomplete closure of the gate-valve. The gate-valve separates (when closed) the vacuum chamber (where is vacuum) from the outside and is driven by Argon pressure, thus, when the Argon pressure sadly drops, the gate-valve can't close because there is not enough pressure to push it. To check that you have to remove the cones and the assembly of lenses behind the cones. Behind the lenses should be no hole further into the vacuum chamber but just a silvery black wall (circular shape). If the gate is not closed, it could be maneuvered (by a service engineer) into open and then closed state, from the Service interface of Mass Hunter (the Ar should be on and you need to have a proper pressure).
Hello! There is the Error 1101... It could be probably generated due to an incomplete closure of the gate-valve. The gate-valve separates (when closed) the vacuum chamber (where is vacuum) from the outside and is driven by Argon pressure, thus, when the Argon pressure sadly drops, the gate-valve can't close because there is not enough pressure to push it. To check that you have to remove the cones and the assembly of lenses behind the cones. Behind the lenses should be no hole further into the vacuum chamber but just a silvery black wall (circular shape). If the gate is not closed, it could be maneuvered (by a service engineer) into open and then closed state, from the Service interface of Mass Hunter (the Ar should be on and you need to have a proper pressure).
Is there any procedure to perform the closure by myself? I am in Colombia, and the technical service is terrible; I have been waiting for an engineer for over a week. Is it possible for me to do it myself?
You must first check if this is the issue. If it is the case, you can ask the engineer to log remotely to your PC, by TeamViewer or so, and set it on or off remotely from Mass Hunter. The main problem is that you need a Service password to access the Service interface, but since you are not trained by Agilent you can not have access to that interface.
So, remove the lenses and send him, or here, a picture of what you see in that remaining space... we need to see if the valve is on or off...
Amazing! I will do that. Thank you so much
I already removed the lenses and I think the gate valve is closed:
Maybe if I change the O-Ring, it will help? Thank you
The IF/Bk errors are also related with the graphite gasket or the cones, the dimensions of the orifices. But, in this case is strange, because the missing of argon does not influence that... check also with other cones.
Ok, I'm cleaning the cones, changing the graphite gasket and I will replace the oil on the pump as well to see if it works.