Using the Agilent 7890A. Column installed with no leaks detected. Both the front and back detector are set to 300°C, gas Flow Rates - Hydrogen: 45mL/min, Air 450mL/min and He 1.0mL/min. The split flow is 80mL/min. Following initial start up the front…
We have a Duo 2.5 foreline pump that was replaced in June of this year (2023). The previous Duo 2.5 installed had a sudden catastrophic leak, thus the replacement in June. The oil was filled up halfway and I set a marker with tape to monitor the level…
Hi byoo,
Do you get any failed to communicate or any other message when you start up the software? You are correct if the busy light should be constantly flashing. This means you got a successfuk firmware load from the PC from ethernet connection. Can…
Hi vitiin96,
this is very generic and any book on GC troubleshooting or the videos on the Agilent website could help. For the 7890 there is a troubleshooting manual available.
Good practice in GC is when you get a (new) column, you have an own suitability…
I have started using GC-FID (Agilent 8890) since March for the first time. We are trying to measure the methane. I was trying to make a standard surve, so I injected 1000 ppm methane standard at first. Just to let you guys know no samples has been run…
no you don't need to disconnect it from the FID, ,
bake the column for enough time usually more than 1 hour according to your column status , you can keep the FID working to monitor the baseline .
..
keep temperature 25 degrees less than the maximum…
Thanks for the tips on seeing the position in the sequence once Aquisition has been restarted. Unfortunately something I failed to mention is that the sequence also disappears after pressing run but will re-appear once you toggle between another module…
Hi all, I am using a GC 6890 with split/splitless inlet to analyse reaction products. However, a problem has recently happened when the GC is running. I have noticed the appearance of spikes in the baseline of the FID detector. The spikes appear occasionally…
Get a WS asynchronous error : F F60304 The torch ignition has failed when starting a Varian 720 OES The torch lights and stays on for about 5 seconds then shuts down with this error. I can not find reference in the help to resolve this. Thanks, Scott
It looks like it is having to clear something out the inlet, column, detector or head space as the baseline rise is over most of it. I see it's a FID you are using, how long do you leave it ignited before running the first analysis?
This Information Applies To: Agilent 5975 and 5977 series GC/MSD
Issue:
Tuning the Agilent GC/MSD with EI sources, requires the use of perfluorotributylamine (PFTBA). The calibration vial will be depleted eventually due to regular use. This article...
This Information Applies To: Agilent 5975 and 5977 series GC/MSD
Issue:
If the PFDTD signal is absent or low, but the reagent ions are normal, you may have run out of CI Calibrant.
Inspect the calibration vial on the back of the flow controller...
This Information Applies To: Agilent 5975 and 5977 Series GC/MSD
Issue:
For some GC/MSD maintenance or repair asks, it is required to open the side plate (filament exchange, ion source cleaning, electron multiplier exchange, column installation...
This Information Applies To: Agilent GC/MS MassHunter Acquisition Software Agilent 5975/5977 GC/MSD
Issue
Use the Early Maintenance Feedback (EMF) tool to track the GC/MSD's consumable resource and part use, and to configure alerts for when they...
This Information Applies To: Agilent GC/MSD systems
Issue:
After a filament fails, there will be a system error message: "No emission current". Switching to the other filament (except for the CI ion source, which has just one filament), and then...
This Information Applies To: Agilent 7693A automatic liquid sampler tray
Issue:
The orange Align Mode light comes on . T he sample tray of the Agilent automatic liquid sampler 7693A will not operate again until the alignment procedure has successfully...
This Information Applies To: Agilent 7693A Automatic Liquid Sampler
Issue:
The automatic liquid sampler may need to be replaced, in case it is worn or dirty.
Steps to follow
To remove the syringe:
Open the injector door.
Slide...
This Information Applies To: Agilent 7693A Automatic Liquid Sampler
Issue:
Two turrets are provided with the 7693A injector, the 16-sample standalone turret, and the 3-sample transfer turret used with the 150-sample tray (See Figure 1 ). The turret...
This Information Applies To: Agilent 7697A Headspace Sampler
Issue:
There is a potential leak external to the Headspace unit that can cause loss of pressure.
Steps to Follow:
This article is a troubleshooting guide for when you suspect there...
This Information Applies To: Agilent 7693A Automatic Liquid Sampler
Issue:
This document explains how to replace the syringe carrier assembly in case you want to change from standard to high volume syringes, or if an Agilent support representative...