Good day . i have another issue to my 8890. after replacing the fused silica in transfer line. tha air flow in FID is suddenly shutting down. then i try the detector diagnostic in the system it said test failed and need to replace the jet.
after replacing…
Hello,
this look like your plasma is not align with the cones. For the EM to failed it's difficult. You have to check for noise when it's doing it and if the counts are low when it's using mass 80. You can try to tune EM with mass 37 if you know where…
Try to ignite the plasma, without fixing the shield plate. Due to poor quality of argon gas or leak on gas line, reflected power will be above the limit and prevent the plasma sustain. Sometimes, the plasma sensor fiber optics tip is damaged, needs to…
Hello,
I am currently developing a residual solvent analysis method by Gas chromatography equipped with a flame ionization detector. My sample is an API with a phosphate salt attached. One of the residual solvents of interest is triethylamine. When…
8/7/19
Hello,
I have had a fair amount of success with posted questions in this forum, so here is another one. I am interested in knowing more about whether the flame ionization detector I have equipped on my HP/Agilent 6890 GC is the correct detector…
We have a lot of 1260s and 1290s in our lab, and as part of our sustainability initiatives are looking for ways to reduce our energy usage. Currently, all the instruments are left on (standby), and most of them also have the lamps on when not in use.…
When I need information about detectors I use the following books:
Raymond P. W. Scott "Chromatographic Detectors - Design, Function, and Operation", Chromatographic Science Series Volume 73, Marcel Dekker, New York (1996). ISBN 0-8247-9779-5.
Raymond…
Hello Inês,
Typically we use the Performance report as a general health report of the instrument. Changing plasma parameters may cause the test to fail, unless you correct the limits. Moving the sample depth will decrease sensitivity and can increase…
Hi Guys,
I have a GC that does not seperate natural gas(maily C1 and C2 with small traces of C3+), it comes as almost one peak, and for condensate (C3+) it seperates C6+, and does not seperate C3÷C5,
I Have,
1: Readjusted the split ratio Many Many…
Hello
I'm trying to setup report to have values in column "PASS/FAIL" green or red depends on result.
Can I do it as conditional formatting for column "PASS/FAIL" to mark FAIL as red based on text?
This column "PASS/FAIL" is custom field and…
This Information Applies To: Agilent 5975 and 5977 series GC/MSD
Issue:
Tuning the Agilent GC/MSD with EI sources, requires the use of perfluorotributylamine (PFTBA). The calibration vial will be depleted eventually due to regular use. This article...
This Information Applies To: Agilent 5975 and 5977 series GC/MSD
Issue:
If the PFDTD signal is absent or low, but the reagent ions are normal, you may have run out of CI Calibrant.
Inspect the calibration vial on the back of the flow controller...
This Information Applies To: Agilent 5975 and 5977 Series GC/MSD
Issue:
For some GC/MSD maintenance or repair asks, it is required to open the side plate (filament exchange, ion source cleaning, electron multiplier exchange, column installation...
This Information Applies To: Agilent GC/MS MassHunter Acquisition Software Agilent 5975/5977 GC/MSD
Issue
Use the Early Maintenance Feedback (EMF) tool to track the GC/MSD's consumable resource and part use, and to configure alerts for when they...
This Information Applies To: Agilent GC/MSD systems
Issue:
After a filament fails, there will be a system error message: "No emission current". Switching to the other filament (except for the CI ion source, which has just one filament), and then...
This Information Applies To: Agilent 7693A automatic liquid sampler tray
Issue:
The orange Align Mode light comes on . T he sample tray of the Agilent automatic liquid sampler 7693A will not operate again until the alignment procedure has successfully...
This Information Applies To: Agilent 7693A Automatic Liquid Sampler
Issue:
The automatic liquid sampler may need to be replaced, in case it is worn or dirty.
Steps to follow
To remove the syringe:
Open the injector door.
Slide...
This Information Applies To: Agilent 7693A Automatic Liquid Sampler
Issue:
Two turrets are provided with the 7693A injector, the 16-sample standalone turret, and the 3-sample transfer turret used with the 150-sample tray (See Figure 1 ). The turret...
This Information Applies To: Agilent 7697A Headspace Sampler
Issue:
There is a potential leak external to the Headspace unit that can cause loss of pressure.
Steps to Follow:
This article is a troubleshooting guide for when you suspect there...
This Information Applies To: Agilent 7693A Automatic Liquid Sampler
Issue:
This document explains how to replace the syringe carrier assembly in case you want to change from standard to high volume syringes, or if an Agilent support representative...