Hi, thank you for your patience! I hope the following instructions help you with your question; not sure what exactly is meant by "processing time". Unfortunately, I cannot see your attached picture.
Argon quality for the Agilent 700, 5100 and 5110 icp-oes systems is 5.0 or better. However the 5800/5900 will work with the 4.5 (99.995%) argon quality. For the optional oxygen addition, needed for some organic matrix measurements like kerosine a mix…
Hi, yes we did resolve the issue and it was a faulty Z-motor. Motor was replaced and "Z-axis" error went away. The "unexpected collision" error was due to me using the wrong vial bottoms with the vial tray I had. I needed to use rounded bottoms instead…
Hello Martin and Andy! I found out what were wrong. In my sequence two injections doesn't have a DA Method linked. When I pin only the injections with DA Method linked, the "No Data Found" was gone.
Thanks for letting us know. I will look into this and determine why it is not working. In the meantime, the file can be downloaded from the 7820A product page at https://www.agilent.com/en/product/gas-chromatography/gc-systems/7820a-gc-system . Scroll…
Thank you very much for your quick answer. It's very important for me to understand the concept well, so I have some consistency problems with the response of the compounds.
The Digest of upcoming webinars has been updated as of May 23.
The new Agilent Community Group, GPC/SEC Solutions Group hosted a webinar series in partnership with Dow in April: A Journey through SEC, Light Scattering, and 2D-LC for Macromolecules …
For more information, check out the application note Five Keys to Unlock Maximum Performance in the Analysis of Over 200 Pesticides in Challenging Food Matrices .
Recommended parts from the video:
Self-tightening, collared column nut for use with…
This Information Applies To: Agilent 5975 and 5977 series GC/MSD
Issue:
Tuning the Agilent GC/MSD with EI sources, requires the use of perfluorotributylamine (PFTBA). The calibration vial will be depleted eventually due to regular use. This article...
This Information Applies To: Agilent 5975 and 5977 series GC/MSD
Issue:
If the PFDTD signal is absent or low, but the reagent ions are normal, you may have run out of CI Calibrant.
Inspect the calibration vial on the back of the flow controller...
This Information Applies To: Agilent 5975 and 5977 Series GC/MSD
Issue:
For some GC/MSD maintenance or repair asks, it is required to open the side plate (filament exchange, ion source cleaning, electron multiplier exchange, column installation...
This Information Applies To: Agilent GC/MS MassHunter Acquisition Software Agilent 5975/5977 GC/MSD
Issue
Use the Early Maintenance Feedback (EMF) tool to track the GC/MSD's consumable resource and part use, and to configure alerts for when they...
This Information Applies To: Agilent GC/MSD systems
Issue:
After a filament fails, there will be a system error message: "No emission current". Switching to the other filament (except for the CI ion source, which has just one filament), and then...
This Information Applies To: Agilent 7693A automatic liquid sampler tray
Issue:
The orange Align Mode light comes on . T he sample tray of the Agilent automatic liquid sampler 7693A will not operate again until the alignment procedure has successfully...
This Information Applies To: Agilent 7693A Automatic Liquid Sampler
Issue:
The automatic liquid sampler may need to be replaced, in case it is worn or dirty.
Steps to follow
To remove the syringe:
Open the injector door.
Slide...
This Information Applies To: Agilent 7693A Automatic Liquid Sampler
Issue:
Two turrets are provided with the 7693A injector, the 16-sample standalone turret, and the 3-sample transfer turret used with the 150-sample tray (See Figure 1 ). The turret...
This Information Applies To: Agilent 7697A Headspace Sampler
Issue:
There is a potential leak external to the Headspace unit that can cause loss of pressure.
Steps to Follow:
This article is a troubleshooting guide for when you suspect there...
This Information Applies To: Agilent 7693A Automatic Liquid Sampler
Issue:
This document explains how to replace the syringe carrier assembly in case you want to change from standard to high volume syringes, or if an Agilent support representative...