Update: FID output signal was very low even when trying to ignite, so we checked the interconnect spring and noticed it was shorter than the ones in the other GCs (not that close to the detector base; dont know if it was always like that or something…
Hi,
We are struggling to solve an issue with an Agilent 8860 GC-FID we purchased in 2019. It is used to quantify propane in headspace samples that are manually injected. It is possible but unlikely that some of those headspace samplings result in the…
Tip for other users that may encounter an FID Ignition problem: try blowing on the FID when it is trying to light. If the FID lights - there is an issue with the gas flows. It is best to use a flow meter like the ADM Flow Meter for gas chromatography…
Tip for other users that may encounter an FID Ignition problem: try blowing on the FID when it is trying to light. If the FID lights - there is an issue with the gas flows. It is best to use a flow meter like the ADM Flow Meter for gas chromatography…
Hey. Johnson we solved the Problem by removing the igniter Castor with collector nut and the spring locks Info the right place. Another Problem could be the malfunction electrometer.
Greetings Agilent Community,
As with many of the errors I encounter with my system, these two happened after I foolishly left the lab to pursue a life outside the world of LCMS.
(1) From the instrument log, the first error read "Failed to run acquisition…
When attempting to submit a sequence in EZChrom, it fails to submit and returns this error message:
This happens each time a sequence is submitted. Any help would be appreciated!
Hi Agilent community.
I'm a LC-MS equipment user and I had a problem with the MassHunter Acquisition B 08.01 software.
When I try to open the Acquisition software, appears the software logo but immediately appears an error dialog box. This error dialog…
Hi everyone,
We have a 6546 QTOF, the following error keeps popping up during acquisition; Failed to start logbook server. A shared violation has occurred. How can I resolve this?. I'll appreciate your help. Thanks
This Information Applies To: Agilent CI Source
Issue:
During normal operation of the GC/MS, the CI ion source will need to be disassembled to proceed with abrasive ion source cleaning and / or routine replacement of consumable parts (filament and...
This Information Applies To: Agilent OpenLab CDS
Issue
This article will explain how to add a sequence summary report line to a sequence, in Agilent OpenLab CDS, as well as how to configure it so it matches your reporting needs.
Background ...
This Information Applies To: Agilent OpenLab CDS, OpenLab CDS ChemStation, OpenLab CDS EZChrom.
Issue
Process for creating a desktop shortcut for your instruments, with OpenLab software, to enable quick access to your acquisition window.
Background...
This Information Applies To: Agilent LC Systems
Issue
Agilent modules will go into an error status (red status light) when an error condition specific to that module occurs. For example, metering home failed will be specific for the autosampler...
This Information Applies To: Agilent Analytical LC Autosamplers
Issue
To increase the maximum injection volume of an autosampler, there are two ways to do it. While both will provide larger injection volumes, there are advantages and disadvantages...
How to Recover a SubscribeNet Password
Multiple Results
This Information Applies To: Agilent Software products with SubscribeNet license.
Issue
Resolutions for password recovery use if you have: lost your password, been locked out of the account, or you don't know if you have an account with SubscribeNet...
This Information Applies To: Agilent OpenLab CDS, OpenLab CDS EZChrom
Issue
Steps to create a Data Analysis shortcut for one or more of your Projects defined in the OpenLab Control Panel .
Steps to Follow
Open the Control Panel of your...
This Information Applies To: Agilent 6100, 6200, 6400 and 6500 LC/MS Systems
Issue
Error on the Mainboard reporting “Leak Detected”.
Background
The fault “Mainboard #1: Leak Detected” is reported in the Data Acquisition Software Logs ( Figure...
This Information Applies To: Agilent LC/MS Systems
Issue
The MS solvent selection valve (SSV) rotor seal is replaced annually as part of the Preventive Maintenance. It can also be replaced as part of troubleshooting a leak, a flow blockage, or when...
This Information Applies To: All Agilent HPLC modules.
Issue:
The firmware is stored on a HPLC module's mainboard. It's a set of parameters, low-level commands, and functions stored on the internal memory of a HPLC module and required for its operation...