This Information Applies To: Agilent CI Source
Issue:
During normal operation of the GC/MS, the CI ion source will need to be disassembled to proceed with abrasive ion source cleaning and / or routine replacement of consumable parts (filament and ceramic insulators). The ion source will then need to be assembled prior to re-installing. Correct abrasive cleaning, assembly, and installation are essential to robust and reliable operation.
Tools Needed:
Figure 1. Parts Breakdown of the CI Source
Item Number | Item Description | Part Number |
1 | Setscrew | G1999-20022 |
2 | Filament screw | G1999-20021 |
3 | CI interface tip seal | G3870-20542 |
4 | CI repeller insulator | G1999-20433 |
5 | CI lens insulator | G3170-20540 |
6 | CI drawout cylinder | G1999-20444 |
7 | CI drawout plate | G1999-20446 |
8 | CI ion source heater block assembly | G3870-60415* |
9 | Entrance lens | G3170-20126 |
10 | CI ion source body | G3170-20430 |
11 | CI ion focus lens | G1999-20443 |
12 | CI repeller | G1999-20432 |
13 | CI filament | G7005-60072 |
14 | Dummy filament | G1999-60454 |
15/16 | Curved Spring washer | 3050-1374 |
* If replacement is required then contact Agilent for repair options
To Disassemble the CI source:
Figure 2. Remove the CI filament, dummy filament, CI repeller and CI insulator from the CI heater-sensor block.
Figure 3. Remove and disassemble the lens, drawout plate, drawout cylinder and insulators from the source body and heater source block.
To Assemble the CI Source:
Figure 4. Assemble the drawout plate, drawout cylinder, lens and insulators into the source body.
|
Caution: Do not overtighten the repeller nut or the ceramic repeller insulators will break when the source heats up. The nut should only be finger-tight
|
Figure 5. Assemble the repeller, filament, dummy filament, repeller and insulators into the source body and heater source block.
|