Problem solved. The format in Region settings should be set to "English (United States)", nothing else. As a result, the Decimal symbol under Numbers in Additional settings becomes a point (Anglo-Saxon decimalization) instead of a comma (Western European…
Hi Helene,
Thank you for contacting Cell Analysis Technical Support. Sodium bicarbonate cannot be used because it is a buffering agent with a very high pKa and, unless incubated in a CO2 incubator, will raise the pH of the assay medium to well above…
Great. The 219 and 502 are quite high. Make sure that the extractor lens insulator is clean, the extractor pin isn't touching the body, the brown extractor lens connector wire is connected well and not crimped, and that the transferline tip seal ceramic…
Hi AVEdsell,
Site prep clearly states that system MUST be vented, this is a snippet from the TQ site Prep Guide about Venting. I do not know waht MS you have but this is common for all :
Bonjour Fabrice,
Juste pour ajouter a l'information que Kristen a deja donne:
Si vous selectionne la langue comme Francais, vous pouvez trouver le manuel. La description, de mon cote, a l'air d'etre en anglais, mais si vous cliquer sur le lien…
Hello lperry ,
For Cool on Column inlets, the septum and the inserts types are dependent on the column size. It is quite possible that either the septum, the insert, or both, are not matching the column size.
Are the needles getting bent too? Does…
This Information Applies To: Agilent ICP-MS systems (7700/7800/7850/7900/8800/8900)
Issue
How to power on the Agilent ICP-MS Foreline Pump when fully powering on the Agilent ICP-MS System, or after performing maintenance.
Background
The ICP...
This Information Applies To: Agilent ICP-MS MassHunter Software version 5.1 and above
Issue
The user manuals for Agilent ICP-MS Systems for MassHunter 5.1 and above are included in the ICP-MS MassHunter Help and Learning Center. There are manuals...
This Information Applies To: Agilent ICP-MS systems (7700/7800/7850/7900/8800/8900)
Issue
Elevated oxide ratio values are defined as greater than 1.5% for x-lens configuration or greater than 3% for s-lens configuration. Possible causes are contamination...
This Information Applies To: Agilent GC/MS Systems
Issue
The software version is a valuable information usually required for troubleshooting, software updates preparation or to check installed options.
Table of contents
MassHunter GC/MS Acquisition...
This Information Applies To: Agilent LC modules with RFID tag readers
Issue
Agilent uses RFID (radio frequency identification tag) tag readers in some HPLC components such as detector lamps and flow cells, pump heads, purge valve and mixers. However...
This Information Applies To: Agilent LC systems with Lab Advisor software.
Issue
In Lab Advisor, the Logs and Results register is available for gathering history and additional information on the system. It presents data collected from the configured...
This Information Applies To: Agilent ICP-MS systems (7700/7800/7850/7900/8800/8900)
Issue
What information does the ICP-MS instrument status indicator provide?
Background
The ICP-MS Instrument Status Indicator on the instrument, and in the...
This Information Applies To: Agilent ICP-MS systems (7700/7800/7850/7900/8800/8900)
Issue
How to power off Agilent ICP-MS systems before maintenance, for long-term shutdown, or for power cycling the instrument. Power cycling may be required when...
This Information Applies To: Agilent OpenLab CDS
Issue
Steps to close an instrument connection to a primary instrument controller, to start a successful connection to a secondary instrument controller, or to configure the instrument.
Background...
This Information Applies To: Agilent HES source, Agilent 5977 Series GC/MSD
Issue:
During normal operation of the GC/MSD, the ion source will need to be disassembled to proceed with abrasive ion source cleaning and / or replacement of consumable...