Thank you for contacting Cell Analysis Technical Support. This will depend on your instrument type. See this link as well as the screenshot below. Keep in mind that these ranges are pre-normalization.
For Cell Mito stress test, Seahorse Cell Mito Stress Test Kit, the recommended working concentrations are 1.5 µM for Oligomycin and 0.5 µM for Rot/AA. The FCCP concentration needs to be optimized (0.5 µM-2 µM). The optimal final compound concentration…
In general, the %CV across a group and from plate to plate should both be < 20%. Higher CVs in this case was due to edge effect on outside of the plate.
GC/MS Photo Parts List for SQ and TQ sources
Stainless Steel, Inert, Extractor, HydroInert, Chemical Ionization, and the High Efficiency Source
These diagrams and parts lists are included in the Troubleshooting & Maintenance manual for the 5977 which…
This Information Applies To: Agilent ICP-MS systems (7700/7800/7850/7900/8800/8900)
Issue
How to power on the Agilent ICP-MS Foreline Pump when fully powering on the Agilent ICP-MS System, or after performing maintenance.
Background
The ICP...
This Information Applies To: Agilent ICP-MS MassHunter Software version 5.1 and above
Issue
The user manuals for Agilent ICP-MS Systems for MassHunter 5.1 and above are included in the ICP-MS MassHunter Help and Learning Center. There are manuals...
This Information Applies To: Agilent ICP-MS systems (7700/7800/7850/7900/8800/8900)
Issue
Elevated oxide ratio values are defined as greater than 1.5% for x-lens configuration or greater than 3% for s-lens configuration. Possible causes are contamination...
This Information Applies To: Agilent GC/MS Systems
Issue
The software version is a valuable information usually required for troubleshooting, software updates preparation or to check installed options.
Table of contents
MassHunter GC/MS Acquisition...
This Information Applies To: Agilent LC modules with RFID tag readers
Issue
Agilent uses RFID (radio frequency identification tag) tag readers in some HPLC components such as detector lamps and flow cells, pump heads, purge valve and mixers. However...
This Information Applies To: Agilent LC systems with Lab Advisor software.
Issue
In Lab Advisor, the Logs and Results register is available for gathering history and additional information on the system. It presents data collected from the configured...
This Information Applies To: Agilent ICP-MS systems (7700/7800/7850/7900/8800/8900)
Issue
What information does the ICP-MS instrument status indicator provide?
Background
The ICP-MS Instrument Status Indicator on the instrument, and in the...
This Information Applies To: Agilent ICP-MS systems (7700/7800/7850/7900/8800/8900)
Issue
How to power off Agilent ICP-MS systems before maintenance, for long-term shutdown, or for power cycling the instrument. Power cycling may be required when...
This Information Applies To: Agilent OpenLab CDS
Issue
Steps to close an instrument connection to a primary instrument controller, to start a successful connection to a secondary instrument controller, or to configure the instrument.
Background...
This Information Applies To: Agilent HES source, Agilent 5977 Series GC/MSD
Issue:
During normal operation of the GC/MSD, the ion source will need to be disassembled to proceed with abrasive ion source cleaning and / or replacement of consumable...