Thank you for your replay. The version of MassHunter (MH) Data Acquisition and quantitative analysis programs software is B.07.00.
Unfortunately, there isn’t any error code for this problem. But a window with “The head of file could…
This file contains a list of 86 commonly used consumables and parts for 8860 and 8890 GC .
Upload this file to your Smart Alerts software to automatically add these parts to the Master Parts List
H ztarbs,
The air would not be for an inlet supply. This is most likely a detector support gas for FID. Check your method to make sure that the FID is not on in a standby method. FID air consumption is 400 to 450ml/min when on.
Regards
James
Hi again
try the last hope ,
first try to select your reporting method in ( DA report template ) , the sequence accept it but i don't know if it will success to print ,
you may try it
the second thing if the first failed ,then try to edit your…
Hello,
We are having an issue with our 7890B GC back inlet. The inlet temperature and flow were both set to "OFF," but the back inlet kept heating up while were were trying to run the front inlet (in GC-FID mode). This resulted in a thermal shutdown…
We use Agilent NGA Analyzer GC based on 4 column configuration. Hayesep Q pre column, Hayesep Analytical column, molsieve column for TCD and one capillary column for FID. The H2S elution takes place in the Hayesep analytical column. But now a days our…
Hello, all you need to do in the method is to uncheck each of the FID flows, and uncheck the electrometer and uncheck the Flame on. With everything unchecked for the FID settings the software will ignore your FID. Under the Signals section you will also…
Hello blairchall ,
I’m sorry for the delayed reply to your message. Do you still need assistance with this? If you do, please take into consideration the notes below. However, if the issue was resolved, could you please share what the resolution was…
We've been having issues with our 7890B gas chromatograph. We have it connected to a workstation running Windows 7 SP1 and OpenLab/Chemstation version C.01.07 [27]. We have a Gerstel MPS attached as a headspace sampler.
We have been trying to run…
This Information Applies To: Agilent ICP-MS MassHunter Workstation 4.6 or below, and ICP-MS MassHunter Workstation 5.1 and above.
Issue
Guide for identifying the current software version your Agilent ICP-MS system is using, to see if any updates...
This Information Applies To: Agilent OpenLab ChemStation, Agilent OpenLab EZChrom, Agilent OpenLab CDS.
Issue
When users of OpenLab CDS forget their passwords, another user with the right privileges will reset their password, however, sometimes...
This Information Applies To: Agilent ICP-MS systems (7700/7800/7850/7900/8800/8900)
Issue
Error messages related to temperature can cause the plasma to shut off or disable plasma ignition. While a warning message will still allow the plasma to light...
This Information Applies To: Agilent ICP-MS MassHunter Software 4.x and 5.x
Issue
How to view meters in ICP-MS MassHunter Software, or change what meters are selected?
Background
ICP-MS MassHunter allows up to five meters to be displayed at...
This Information Applies To: Agilent ICP-MS (7800/7850/7900/8900)
Issue
How to replace the oil mist filter in the MS40+ or MS40-S?
Background
The oil mist filter in the MS40+ or MS40-S should be replaced yearly.
Warning :
...
This Information Applies To: Agilent ICP-MS (7700/7800/8800)
Issue
How to check the DS 402 foreline pump to determine if the pump needs maintenance.
Background
The oil cleanliness and oil level in the DS 402 foreline pump should be checked...
This Information Applies To: Agilent ICP-MS systems (7700, 7800, 7850, 7900, 8800, 8900)
Issue
After shutting down and restarting the ICP-MS, sometimes the communication between ICP-MS and PC does not automatically reconnect. It may be necessary...
This Information Applies To: Agilent ICP-MS (7700/7800/8800)
Issue
How to replace the Foreline Pump Oil Mist Filter in the DS 402 foreline pump.
Background
The oil mist filter element should be replaced yearly , or when the level of oil in...
This Information Applies To: Agilent ICP-MS (7500/7700/8800)
Issue
How to replace the Forline Pump Oil Mist Filter and Odor Element in the E2M18 or E2M18F?
Background
The oil mist filter and odor element should be replaced yearly, or if the...
This Information Applies To: Agilent ICP-MS (7500/7700/8800)
Issue
How to replace the Forline Pump Oil in the E2M18 or E2M18F?
Background
The E2M18 oil should be replaced at least every three months or when the oil color indicates it needs...