Problem solved. The format in Region settings should be set to "English (United States)", nothing else. As a result, the Decimal symbol under Numbers in Additional settings becomes a point (Anglo-Saxon decimalization) instead of a comma (Western European…
Hi Helene,
Thank you for contacting Cell Analysis Technical Support. Sodium bicarbonate cannot be used because it is a buffering agent with a very high pKa and, unless incubated in a CO2 incubator, will raise the pH of the assay medium to well above…
Great. The 219 and 502 are quite high. Make sure that the extractor lens insulator is clean, the extractor pin isn't touching the body, the brown extractor lens connector wire is connected well and not crimped, and that the transferline tip seal ceramic…
Hi AVEdsell,
Site prep clearly states that system MUST be vented, this is a snippet from the TQ site Prep Guide about Venting. I do not know waht MS you have but this is common for all :
Bonjour Fabrice,
Juste pour ajouter a l'information que Kristen a deja donne:
Si vous selectionne la langue comme Francais, vous pouvez trouver le manuel. La description, de mon cote, a l'air d'etre en anglais, mais si vous cliquer sur le lien…
Hello lperry ,
For Cool on Column inlets, the septum and the inserts types are dependent on the column size. It is quite possible that either the septum, the insert, or both, are not matching the column size.
Are the needles getting bent too? Does…
Hello Samehalsharawy ,
Can you please elaborate on the statement we uninstalled the cylinder expecting the flow to come down to zero but it didn't. It stayed at 50?
If no physical tank is connected to the GC, the GC gas connections are capped…
This Information Applies To: Agilent ICP-MS MassHunter Workstation 4.6 or below, and ICP-MS MassHunter Workstation 5.1 and above.
Issue
Guide for identifying the current software version your Agilent ICP-MS system is using, to see if any updates...
This Information Applies To: Agilent OpenLab ChemStation, Agilent OpenLab EZChrom, Agilent OpenLab CDS.
Issue
When users of OpenLab CDS forget their passwords, another user with the right privileges will reset their password, however, sometimes...
This Information Applies To: Agilent ICP-MS systems (7700/7800/7850/7900/8800/8900)
Issue
Error messages related to temperature can cause the plasma to shut off or disable plasma ignition. While a warning message will still allow the plasma to light...
This Information Applies To: Agilent ICP-MS MassHunter Software 4.x and 5.x
Issue
How to view meters in ICP-MS MassHunter Software, or change what meters are selected?
Background
ICP-MS MassHunter allows up to five meters to be displayed at...
This Information Applies To: Agilent ICP-MS (7800/7850/7900/8900)
Issue
How to replace the oil mist filter in the MS40+ or MS40-S?
Background
The oil mist filter in the MS40+ or MS40-S should be replaced yearly.
Warning :
...
This Information Applies To: Agilent ICP-MS (7700/7800/8800)
Issue
How to check the DS 402 foreline pump to determine if the pump needs maintenance.
Background
The oil cleanliness and oil level in the DS 402 foreline pump should be checked...
This Information Applies To: Agilent ICP-MS systems (7700, 7800, 7850, 7900, 8800, 8900)
Issue
After shutting down and restarting the ICP-MS, sometimes the communication between ICP-MS and PC does not automatically reconnect. It may be necessary...
This Information Applies To: Agilent ICP-MS (7700/7800/8800)
Issue
How to replace the Foreline Pump Oil Mist Filter in the DS 402 foreline pump.
Background
The oil mist filter element should be replaced yearly , or when the level of oil in...
This Information Applies To: Agilent ICP-MS (7500/7700/8800)
Issue
How to replace the Forline Pump Oil Mist Filter and Odor Element in the E2M18 or E2M18F?
Background
The oil mist filter and odor element should be replaced yearly, or if the...
This Information Applies To: Agilent ICP-MS (7500/7700/8800)
Issue
How to replace the Forline Pump Oil in the E2M18 or E2M18F?
Background
The E2M18 oil should be replaced at least every three months or when the oil color indicates it needs...