For Cell Mito stress test, Seahorse Cell Mito Stress Test Kit, the recommended working concentrations are 1.5 µM for Oligomycin and 0.5 µM for Rot/AA. The FCCP concentration needs to be optimized (0.5 µM-2 µM). The optimal final compound concentration…
In general, the %CV across a group and from plate to plate should both be < 20%. Higher CVs in this case was due to edge effect on outside of the plate.
GC/MS Photo Parts List for SQ and TQ sources
Stainless Steel, Inert, Extractor, HydroInert, Chemical Ionization, and the High Efficiency Source
These diagrams and parts lists are included in the Troubleshooting & Maintenance manual for the 5977 which…
The free Glycan Peak Assignment Tool was designed to provide additional confidence in identifying InstantPC or 2AB N-glycan peaks by predicted retention times of over 100 glycans. The new tool does not require users to upload sensitive data from N-glycan…
This Information Applies To: Agilent ICP-MS MassHunter Workstation 4.6 or below, and ICP-MS MassHunter Workstation 5.1 and above.
Issue
Guide for identifying the current software version your Agilent ICP-MS system is using, to see if any updates...
This Information Applies To: Agilent OpenLab ChemStation, Agilent OpenLab EZChrom, Agilent OpenLab CDS.
Issue
When users of OpenLab CDS forget their passwords, another user with the right privileges will reset their password, however, sometimes...
This Information Applies To: Agilent ICP-MS systems (7700/7800/7850/7900/8800/8900)
Issue
Error messages related to temperature can cause the plasma to shut off or disable plasma ignition. While a warning message will still allow the plasma to light...
This Information Applies To: Agilent ICP-MS MassHunter Software 4.x and 5.x
Issue
How to view meters in ICP-MS MassHunter Software, or change what meters are selected?
Background
ICP-MS MassHunter allows up to five meters to be displayed at...
This Information Applies To: Agilent ICP-MS (7800/7850/7900/8900)
Issue
How to replace the oil mist filter in the MS40+ or MS40-S?
Background
The oil mist filter in the MS40+ or MS40-S should be replaced yearly.
Warning :
...
This Information Applies To: Agilent ICP-MS (7700/7800/8800)
Issue
How to check the DS 402 foreline pump to determine if the pump needs maintenance.
Background
The oil cleanliness and oil level in the DS 402 foreline pump should be checked...
This Information Applies To: Agilent ICP-MS systems (7700, 7800, 7850, 7900, 8800, 8900)
Issue
After shutting down and restarting the ICP-MS, sometimes the communication between ICP-MS and PC does not automatically reconnect. It may be necessary...
This Information Applies To: Agilent ICP-MS (7700/7800/8800)
Issue
How to replace the Foreline Pump Oil Mist Filter in the DS 402 foreline pump.
Background
The oil mist filter element should be replaced yearly , or when the level of oil in...
This Information Applies To: Agilent ICP-MS (7500/7700/8800)
Issue
How to replace the Forline Pump Oil Mist Filter and Odor Element in the E2M18 or E2M18F?
Background
The oil mist filter and odor element should be replaced yearly, or if the...
This Information Applies To: Agilent ICP-MS (7500/7700/8800)
Issue
How to replace the Forline Pump Oil in the E2M18 or E2M18F?
Background
The E2M18 oil should be replaced at least every three months or when the oil color indicates it needs...