Update: FID output signal was very low even when trying to ignite, so we checked the interconnect spring and noticed it was shorter than the ones in the other GCs (not that close to the detector base; dont know if it was always like that or something…
Hi,
We are struggling to solve an issue with an Agilent 8860 GC-FID we purchased in 2019. It is used to quantify propane in headspace samples that are manually injected. It is possible but unlikely that some of those headspace samplings result in the…
Tip for other users that may encounter an FID Ignition problem: try blowing on the FID when it is trying to light. If the FID lights - there is an issue with the gas flows. It is best to use a flow meter like the ADM Flow Meter for gas chromatography…
Tip for other users that may encounter an FID Ignition problem: try blowing on the FID when it is trying to light. If the FID lights - there is an issue with the gas flows. It is best to use a flow meter like the ADM Flow Meter for gas chromatography…
Hey. Johnson we solved the Problem by removing the igniter Castor with collector nut and the spring locks Info the right place. Another Problem could be the malfunction electrometer.
Greetings Agilent Community,
As with many of the errors I encounter with my system, these two happened after I foolishly left the lab to pursue a life outside the world of LCMS.
(1) From the instrument log, the first error read "Failed to run acquisition…
When attempting to submit a sequence in EZChrom, it fails to submit and returns this error message:
This happens each time a sequence is submitted. Any help would be appreciated!
Hi Agilent community.
I'm a LC-MS equipment user and I had a problem with the MassHunter Acquisition B 08.01 software.
When I try to open the Acquisition software, appears the software logo but immediately appears an error dialog box. This error dialog…
Hi everyone,
We have a 6546 QTOF, the following error keeps popping up during acquisition; Failed to start logbook server. A shared violation has occurred. How can I resolve this?. I'll appreciate your help. Thanks
This Information Applies To: Agilent OpenLab CDS (version 2.x)
Contents:
In the following procedure, you will find steps to create a multipoint ESTD calibration method with OpenLab CDS 2.X.
Steps to follow:
1. Load the Data
a. From Data...
This Information Applies To: Agilent MassHunter Quantitative Analysis Software
Issue
Chromatographic peak integration is a critical part of accurate and reproducible quantitative data analysis. The integrator type and the integration parameters...
This Information Applies To: Agilent 6200/6400/6500 Series LC/MS Systems
Issue
Instructions on how to fill the calibrant bottle on 6200/6400/6500 Series LC/MS Instruments when calibrant levels are running low.
Background
The ability to autotune...
This Information Applies To: All Agilent 6000 Series LC/MS Systems without a gate valve for ion capillary installation/removal.
Issue:
Installing the Ion Transfer Capillary after cleaning or replacement. For instructions for installing an ion capillary...
This Information Applies To: AriaMx Real-time PCR System
Issue
The saturation level is unique for each optical module after the offset of calibration factor and background calibration. The FAM, ROX, HEX, CY3 & CY5 optical modules will saturate at...
This Information Applies To: Agilent LC Systems
Issue
While it is possible to combine modules from different HPLC series such as 1100, 1200, 1260 or 1290 series to create a functioning LC system, there are some important topics to consider.
...
Fixes for the "RPC Server Unavailable" Error
Multiple Results
This Information Applies To: Agilent OpenLab CDS
Issue
When launching the instrument Acquisition screen/instrument session, there is an "RPC Server Unavailable" error.
Resolution
Accessing the Services App (Windows)
There are two ways to...
This Information Applies To: Agilent LC Pumps
Issue
If a user suspects that the gradient composition isn’t accurate and causes poor chromatography such as retention time shifts, late, early, no peaks or merging peaks. This article provides a quick...
This Information Applies To: Agilent 1290 Infinity (G4226A) and Agilent 1260 Infinity (G1367E) HiP Autosampler
Issue
Replacing the Needle Seat of the Autosampler is required when:
The needle seat is visibly damaged.
Leaks or blockages in the...
This Information Applies To: Agilent OpenLab CDS
Issue
This article discusses different types of bracketing modes and how to set up a bracketed calibration in Data Analysis for OpenLab CDS 2.x
Background
Occasionally, the instrument conditions...