Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
  • Home
  • Technical Areas
    Technical Areas
    • AssayMap
    • Atomic Spectroscopy
    • Automated Electrophoresis
    • Cell Analysis
    • Consumables
    • CrossLab Connect
      CrossLab Connect
      • Service Management
      • Smart Alerts
    • Dissolution
    • GC
    • GC/MS
    • LC
    • LC/MS
    • Life Sciences Reagents
    • Molecular Spectroscopy
      Molecular Spectroscopy
      • Molecular Spectroscopy - UV-Vis & UV-Vis-NIR, FTIR, Raman & Fluorescence
      • 8700 LDIR User Group
      • Transmission Raman User Group
    • Real-Time PCR (qPCR)
    • Sample Preparation
    • Software
      Software
      • Chromatography Software
        OpenLab CDS, OpenLab ChemStation, OpenLab EZChrom
      • Data Management Software
        OpenLab ECM, OpenLab ECM XT, OpenLab Server
      • Mass Spectrometry Software
        Data Analysis, Data Acquisition, Productivity App, WalkUp Software
    • Vacuum
  • Applications Areas
    Applications Areas
    • Basic Research
    • Energy and Chemicals
    • Environmental
    • Food and Agriculture
  • Agilent Knowledge Portal
  • Links
    Links
    • Collection of Support Resources
    • Webinar Notifications
    • Contact Support
    • User Contributions
  • Getting Started
    Community Help and Resources
  • Browse
    Browse
    • People
    • Places
    • Your View
    • Your Community Inbox
LC/MS Portal
  • Agilent Knowledge Portal
LC/MS Portal
Articles Installing the Ion Transfer Capillary in Agilent 6000 Series LC/MS Systems
  • Articles
  • Videos
  • Forum
  • Announcements
  • More
  • Cancel
  • Agilent Knowledge Portal
  • LC/MS Portal Home
  • LC/MS Portal Articles
  • LC/MS Portal Videos
  • More
  • Cancel
  • Created :  10 Nov 2022
  • Modified :  11 Jul 2023
  • Category :  Agilent Knowledge Portal
  • Entry Type :  Article
  • Product Type :  LC/MS Systems
  • Component :  LC/MS Quadrupole TOF LC/MS Single Quadrupole LC/MS Time of Flight LC/MS Triple Quadrupole
  • Product Name :  6220A Time of Flight LC/MS 6224A Time of Flight LC/MS 6230 Time of Flight LC/MS 6410 Triple Quadripole LC/MS 6420 Triple Quadripole LC/MS 6430 Triple Quadripole LC/MS 6460 Triple Quadripole LC/MS 6470 Triple Quadripole LC/MS 6490 Triple Quadripole LC/MS 6510 Q-TOF LC/MS 6520 Q-TOF LC/MS 6530 Q-TOF LC/MS 6538 Q-TOF LC/MS 6540 LC/Q-TOF 6545 LC/Q-TOF 6545XT AdvanceBio LC/Q-TOF 6546 LC/Q-TOF 6550 iFunnel LC/Q-TOF 6560 Ion Mobility LC/Q-TOF
  • Task :  Maintenance
Related Articles
  • Removing the Ion Transfer Capillary from Agilent 6000 Series LC/MS Systems

    This Information Applies To: All Agilent 6000 Series LC/MS Systems without a gate valve for ion capillary installation/removal. For instructions for removing an ion capillary on LC/MS instruments with a gate valve installed, please click here Issue: Removing the Ion Transfer Capillary for maintenance. Steps to follow: Tips : The figures in this article show an LC/MS with an Agilent Jet Stream (AJS) source installed. The process of capillary removal is the same for LC/MS instruments...
  • Removing the Ion Transfer Capillary from Triple Quadrupole LC/MS Systems Equipped With a Gate Valve

    This Information Applies To: Agilent 6400 Series triple quadrupole LC/MS systems equipped with a gate valve. Issue Removing the Ion Transfer Capillary for maintenance on instruments with gate valve. The gate valve allows ion capillary removal and installation without having to vent the instrument. Required Items Clean powder-free nitrile gloves Capillary puller tool (supplied with your triple quadrupole LC/MS with a gate valve installed) Steps to follow Caution : Handling...
  • Installing an Ion Capillary in Triple Quadrupole LC/MS Systems Equipped With a Gate Valve

    This Information Applies To: Agilent 6400 Series triple quadrupole LC/MS systems equipped with a gate valve Issue Installing the Ion Transfer Capillary after cleaning or replacement in Agilent LC/MS instruments with a gate valve. Required Items Clean powder-free nitrile gloves LC/MS-grade isopropanol Caution : Ensure the source heated zones are cool before working on the instrument. Before installing the new ion capillary, the used ion capillary must be removed...
  • What to do if my LC-QQQ Autotune Fails?

    This Information Applies To: Agilent 6400 Series triple quadrupole LC/MS systems Issue Autotune fails to complete. Background Agilent 6400 LC/MS systems use an Autotune algorithm to optimize the sensitivity and mass resolution of the instrument. Sometimes when running the autotune procedure the autotune does not complete and the user is given an Autotune failed message in MassHunter. One of the main reasons for Autotune failure is a lack of sensitivity or abundance of the tuning ions...
  • How To Ping the IP Address of an Agilent GC System

    This Information Applies To: Agilent GC Systems 6890N, 7890A/B, 7820, 8860,8890, Intuvo 9000 Issue The communication between Agilent GC instrument and the computer (PC) requires an active LAN connectivity. Each part of the instrument-PC network has a unique IP Address, a shared Subnet Mask, and an undefined Gate Way. This ping test verifies the LAN connectivity between PC and instrument. If the pinging is not successful, the customer will not be able to configure the new instrument or the configured...

Installing the Ion Transfer Capillary in Agilent 6000 Series LC/MS Systems

Answer

This Information Applies To: All Agilent 6000 Series LC/MS Systems without a gate valve for ion capillary installation/removal. 


Issue:

Installing the Ion Transfer Capillary after cleaning or replacement. For instructions for installing an ion capillary on LC/MS instruments with a gate valve installed, click here


Steps to follow:

 Tip
Tips: The following figures show an LC/MS with an Agilent Jet Stream (AJS) source installed. The process of capillary installation is the same for LC/MS instruments equipped with the standard electrospray ion source installed.
  1. Put on clean powder-free nitrile gloves.

     Caution 
    Caution: Handling Ion Source components or the Ion Transfer Capillary without gloves can cause contamination. Use clean powder-free nitrile gloves during this procedure
  2. Lubricate the ion capillary entrance end with LC/MS-grade isopropanol (IPA) (Figure 1).



    Figure 1: Arrow illustrating the correct end of the resistive capillary to apply IPA

  3. Check the correct orientation of the capillary before inserting it into the instrument. The black band on resistive capillaries is positioned toward the front (spray chamber side) of the instrument.

  4. Carefully insert the capillary straight into the desolvation assembly. Do not turn or twist the capillary when you install it or you can damage the metal plating.

     Caution 
    Caution: Vertical or horizontal pressure on the capillary during installation can break the capillary. 
  5. When 2 to 3 cm of the capillary remains extended from the desolvation assembly, the capillary rests against the rear contract spring, which slightly restricts further insertion of the capillary. Slightly increase the insertion force to push the capillary through the rear contact spring to fully insert the capillary (Figure 2). Continue to apply gentle pressure until approximately 1 cm of capillary remains extended from the desolvation assembly.



    Figure 2: Video showing the correct procedure for inserting the ion capillary into the desolvation assembly of the LC/MS
  6. Install the contact ring and end plate (Figure 3).



    Figure 3: Installing the contact ring and end plate onto the desolvation assembly

  7. Wet the end of the ion capillary with LC/MS-grade IPA, then install the capillary cap onto the end of the capillary. Screw on the spray shield (Figure 4).

     Caution
    Caution: Do not spray IPA directly into the ion capillary entrance. Use a small amount of IPA applied from the side.

     



    Figure 4: Installing the capillary cap and spray shield.

  8. Close the source and pump down the system. When the system reaches normal operating vacuum readings, bring the source back up to your method temperature and gas flow setpoints.

  9. Wait a minimum of 2 hours before using the instrument for analysis to let the heated zones in the instruments fully equilibrate. 

 