Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
  • Home
  • Technical Areas
    Technical Areas
    • AssayMap
    • Atomic Spectroscopy
    • Automated Electrophoresis
    • Cell Analysis
    • Consumables
    • CrossLab Connect
      CrossLab Connect
      • Service Management
      • Smart Alerts
    • Dissolution
    • GC
    • GC/MS
    • LC
    • LC/MS
    • Life Sciences Reagents
    • Molecular Spectroscopy
      Molecular Spectroscopy
      • Molecular Spectroscopy - UV-Vis & UV-Vis-NIR, FTIR, Raman & Fluorescence
      • 8700 LDIR User Group
      • Transmission Raman User Group
    • Real-Time PCR (qPCR)
    • Sample Preparation
    • Software
      Software
      • Chromatography Software
        OpenLab CDS, OpenLab ChemStation, OpenLab EZChrom
      • Data Management Software
        OpenLab ECM, OpenLab ECM XT, OpenLab Server
      • Mass Spectrometry Software
        Data Analysis, Data Acquisition, Productivity App, WalkUp Software
    • Vacuum
  • Applications Areas
    Applications Areas
    • Basic Research
    • Energy and Chemicals
    • Environmental
    • Food and Agriculture
  • Agilent Knowledge Portal
  • Links
    Links
    • Collection of Support Resources
    • Webinar Notifications
    • Contact Support
    • User Contributions
  • Getting Started
    Community Help and Resources
  • Browse
    Browse
    • People
    • Places
    • Your View
    • Your Community Inbox
LC Portal
  • Agilent Knowledge Portal
LC Portal
Articles Quick way to Manually Determine Gradient Composition Accuracy
  • Articles
  • Videos
  • Forum
  • Announcements
  • More
  • Cancel
  • Agilent Knowledge Portal
  • LC Portal Home
  • LC Portal Articles
  • LC Portal Videos
  • More
  • Cancel
  • Created :  31 Mar 2023
  • Modified :  30 Jun 2023
  • Category :  Agilent Knowledge Portal
  • Entry Type :  Article
  • Product Type :  Analytical LC Systems
  • Component :  LC Pumps
  • Product Name :  1100 Series LC System 1120 Compact LC 1200 Series LC System 1220 infinity II LC System 1260 Infinity II Bio-Inert LC System 1260 Infinity II LC System 1260 Infinity II Prime Bio LC System 1260 Infinity LC System 1260 Infinity II Prime LC System 1290 Infinity II Bio LC System 1290 Infinity II LC System 1290 Infinity LC System
  • Task :  Troubleshooting
Related Articles
  • What Components determine Gradient Composition and Flow Accuracy in Binary and Quaternary Pumps

    This Information Applies To: Agilent Binary and Quaternary LC Pumps Issue When performing a gradient analysis, it’s important to know what component determines composition if it becomes necessary to troubleshoot chromatographic inconsistencies. Those inconsistencies can manifest themselves as peak retention time shifts (late or early peaks) and merged or missing peaks. Troubleshooting strategy depends on the pump type being used. This article will describe how binary and quaternary pumps create...
  • Troubleshooting Unstable Retention Time Issues - Cleaning the Multichannel Gradient Valve (MCGV) on Agilent Quaternary Pump Modules

    This Information Applies To: All Agilent LC pump modules with MCGV installed (quaternary pumps) Issue: Shifting retention times: The composition of the mobile phase in liquid chromatography directly influences retention times. If any of the channels of the MCGV are partially or fully blocked, the gradient produced will be different from the analytical method. Clogging of the MCGV may occur when the system is not rinsed properly or stored incorrectly. If the following steps don’t solve the...
  • No Pressure, No Flow, RT shift Issues: Cleaning Acetonitrile (ACN) Aggregations in Agilent Pumps

    This Information Applies To: Agilent LC pumps Issue: Stuck or obstructed check valves (Active Inlet Valve (AIV), Passive Inlet Valve (PIV), Outlet Ball Valve (OBV)) can cause issues such as no-flow or no-pressure, moderate to strong drops in pressure or retention time shifts. You can see the article How to Troubleshoot no flow or no pressure issues: Binary/Quaternary/Isocratic Pumps - G131XX, and G711XX to troubleshoot the issue. If one of the valves is the source of the blockage, and acetonitrile...
  • How to Stop Pump Flow When Out of Solvent

    This Information Applies To: Agilent LC Pumps Issue Modern Chromatography software will allow the users to determine what to do at the end of a sequence run. The entire system can be shut down or only certain modules can be turned off, such as turning off the detector lamp. In the case where the user forgets, there are other ways to at least stop the pump from continuously running when the solvent bottle is empty. A dry pump, which continues to pump will wear out the pistons and piston seals...
  • How to Integrate Merged Peaks in OpenLab CDS (ver. 2.x)

    This Information Applies To: Agilent OpenLab CDS (ver 2.x) Issue: How to integrate non-ideally separated peaks starting above the zero baseline in OpenLab CDS Resolution Integrating merged peaks is a classical challenge in chromatography in cases where two compounds cannot be fully separated. This post outlines some key steps and references to helpful resources on how to address this problem. What are merged peaks? Merged peaks occur when a new peak begins before the end of peak is...

Quick way to Manually Determine Gradient Composition Accuracy

Answer

This Information Applies To: Agilent LC Pumps


Issue

If a user suspects that the gradient composition isn’t accurate and causes poor chromatography such as retention time shifts, late, early, no peaks or merging peaks. This article provides a quick way to determine if the pump can accurately proportion the requested composition. The test is done by manually changing the composition and noting the responses.

Note: This procedure can only be done with a UV detector, not valid for non-UV systems.


Steps to Follow

  1. Place 200 mL of LC-grade water into a clean reservoir. For a binary pump, place it in channel A1 and for a quaternary pump, use channel A.
  2. Prepare 200 mL of a 0.4% acetone aqueous solution and place it into another reservoir. If using a binary pump, place it in channel B1. If using a quaternary pump, place it in channel B.
  3. Set the detector wavelength to 265 nm.
  4. If a column is installed, remove it, and replace it with a union.
  5. Flush the system thoroughly with water.
  6. Balance (zero) the detector and ensure that the baseline is flat and stable. It is important to patiently wait for a stable baseline and the detector may need to be balanced several times.
  7. Set the flow at 2.0 mL/min and flush for about 5 minutes. The baseline must be flat and above zero. If not, manually balance the detector.
  8. Change the composition to 100% acetone aqueous solution.
  9. The baseline will rapidly rise when the acetone solution gets into the detector cell. Wait until it flattens and is stable. Note and record the absorbance. You can zoom on the plateau to determine the height. A good alternative would be to use the Agilent Instant Pilot to accurately determine the absorbance value. If the response is above 1000mAU, reduce the acetone percentage of the solution, and start again from step 2.
  10. Now manually change to 50:50 water/acetone composition. The baseline will slowly decrease to an expected response of half the response from step 9. Note and record this response as well.
  11. If the response at 100% of acetone solution was 1000 mAU in step 9, then the expected response in step 10 should be around 500 mAU. The composition accuracy for Agilent pumps may vary depending on the specific pump model.
  12. You can repeat and test at other compositions, however, avoid testing at below 5% composition of either solvent as this test is not significant at those compositions.

If the test fails, ensure you have followed the procedure carefully then contact your local Agilent service representative.