Hello ivynotterpek ,
I'm not aware of any way to do this with just one compound/cal curve. One solution would be to make duplicate compounds and assign one compound the low level standards and the other the high level standards. You can utilize different…
Thank you, unfortunately we tried to set a collection window much larger than the peak duration, but the peak got split again. And in the version of the software we are using no advanced settings allowing to collect different peaks into the same vial…
Hi CMZainea,
Thank you for your feedback. The leak test for the 6 ports valve was passing, so no leaks detected at the valve. Any other leaks were detected from the visible piping and connection from the back panel of the 7697A going to the 6 ways valve…
The two-way splitter without makeup G3181B can be used as a splitter or a combiner to connect two inlets to one detector. Discuss this part with your local sales rep and they can help you get a quote for this part.
https://www.agilent.com/cs/library…
The Digest of upcoming webinars has been updated as of February 20.
Notable additions include:
BioPharma (1)
GC & GCMS (2)
Pharma (1)
Software & Informatics (21)
Check Out:
Tips and Tricks Workshop on ICP-OES: Smarter Method Development…
This Information Applies To: Agilent GC Systems
Issue
Excessive baseline noise can be introduced at different points including sample preparation, sample introduction, separation, and the detector. The excessive background can cause reduced signal...
This Information Applies To: Agilent GC Systems
Issue
Baseline instability that is not reproducible across multiple runs can be introduced at different points including sample introduction, separation, and the detector. The baseline instability...
This Information Applies To: Agilent GC Systems
Issue
Peak fronting is present in a chromatography peak when it has an excessive asymmetry with a leading edge. A normal peak is almost symmetrical.
Other chromatographic problems are identified...
This Information Applies To: Agilent GC Systems
Issue
A shift in retention time can occur for single, multiple or all peaks.
Other chromatographic problems are identified in Basic Troubleshooting for GC Systems . Regular maintenance of your instrument...
This Information Applies To: Agilent GC Systems
Issue
A change in peak size can occur for single, multiple or all peaks.
Other chromatographic problems are identified in Basic Troubleshooting for GC Systems . Regular maintenance of your instrument...
This Information Applies To: Agilent GC Systems
Issue
Peak splitting is when a single compound is shown in the chromatogram as two peaks usually unresolved, or one of the peaks is a shoulder.
Other chromatographic problems are identified in Basic...
This Information Applies To: Agilent GC Systems
Issue
This test isolates the point of contamination to the sample introduction system, which includes the gas supply, gas supply lines, and the sample inlet. It does not include the syringe or autosampler...
This Information Applies To: Agilent GC systems
Issue
Peak tailing is present in a chromatography peak when it has an excessive asymmetry with a trailing edge. A normal peak is almost symmetrical.
Other chromatographic problems are identified...
T his Information Applies To: Agilent 8890/8860/7890/7820/6850/6890 GC Systems with TCD (Thermal Conductivity Detector)
Issue
Problems observed with thermal conductivity detector (TCD) results can have different underlying causes.
Resolution...
This Information Applies To: All GCs, 6890, 6850, 7890, 7820, 8890/8860, Intuvo 9000
Issue:
This article is a troubleshooting guide for when the Flame Ionization Detector (FID) does not ignite or the flame keeps going out.
Cause:
If the FID...