Yes, the ms40+ start. They go up to 3.2E0 Torr.
However, PS1 does not light.
PS2 only light the power light, not the other two.
Are the Turbos connected. Does Turbo2 need Turbo 1 to work? Because if independent, then the problem seems to affect…
Which Electron Ionization Source Part 1 has questions. Which Electron Ionization Source Part 2 has a few answers. Which Electron Ionization Source Part 3 has the particulars of each source.
This segment - Which Electron Ionization Ion Source Part 4…
sammyk I would recommend upgrading Firmware/Embedded Software to ESW 9.4.71. This may help address your issue that you have been experiencing.
The Agilent Call Center may be able to guide you through this. The ESW may be downloaded through Agilent Subscribenet…
Hi there,
Thank you for contacting Cell Analysis Technical Support. Regarding calculations, please see the below chart for the GRA Assay. Additionally, here is a link to the GRA White Paper.
For your second question, I would have to take a…
On the GC, is the Not Ready LED lit? If so, press the Status button and the GC will tell you what is not ready.
Split or Splitless? If it's Splitless the GC may need to be set to AutoPrepRun if the CTC sampler is not configured to send the prep run…
We have not tested nor validated the use of Oligo Pro II data on ProSize so we cannot comment on why it would work for your colleagues. As for why only a single row is showing, please collect the zip file for your data as outlined in my previous message…
This Information Applies To: Agilent ICP-MS systems (7700/7800/7850/7900/8800/8900)
Issue
The torch, bonnet, and platinum shield plate must be installed after performing maintenance, or replacing with new components.
Items Required
Standard...
This Information Applies To: Agilent ICP-MS systems (7700/7800/7850/7900/8800/8900)
Issue
The torch, bonnet, and shield plate must be removed and inspected during routine maintenance, and when troubleshooting plasma ignition problems. Faults with...
This Information Applies To: Agilent ICP-MS systems (7700/7800/7850/7900/8800/8900)
Issue
This procedure tests for an air leak in the sample introduction system, or fault with spray chamber draining, by sealing the end of the torch with a piece...
This Information Applies To: 7700, 7800, 7850, 7900 Agilent ICP-MS instruments
Issue
Error messages related to gas flows can cause the plasma to shut off or disable plasma ignition.
The troubleshooting procedure in this article is applicable for...
This Information Applies To: Agilent ICP-MS systems (7700/7800/7850/7900/8800/8900)
Issue
Error messages related to water flow or water valve issues can cause the plasma to shut off or disable plasma ignition. While a warning message will still...
This Information Applies To: Agilent Fluorescence Detectors (FLD)
Issue
Sometimes it’s necessary to change FLD EX and EM wavelengths during a run since not all components will excite or emit at the same wavelengths. However, when the wavelength...
This Information Applies To: Agilent Analytical Scale LC Systems
Issue
To troubleshoot high backpressure within an HPLC system, it’s necessary to understand the expected pressure throughout an operating system. The total pressure for an LC is measured...
This Information Applies To: Agilent LC Samplers except dual needle multisamplers
Issue
High pressure in the LC sampler is sometimes due to a clog in the needle seat assembly, the needle, or the sample loop. The needle can pick up sample particulates...
This Information Applies To: All OpenLab CDS ChemStation Edition software-Standalone Workstation
Issue
After launching OpenLab ChemStation Online, the following error is shown: “ChemStation ChemMain has stopped working” . OpenLab ChemStation Online...
This Information Applies To: Agilent AriaMx, Software version 1.8 and later
Issue
The Aria Warm up Feature was introduced with Aria SW/FW 1.8 to reduce baseline drift. The warm-up feature is used to speed up the heating process that would normally...