This Information Applies To: Agilent Analytical Scale LC Systems
Issue
To troubleshoot high backpressure within an HPLC system, it’s necessary to understand the expected pressure throughout an operating system. The total pressure for an LC is measured by a pressure sensor in the pump. In the hypothetical LC schematic in Figure 1, pressure sensors are placed at various points to give an understanding of what each hypothetical sensor would read.
Background
Assume that the total backpressure on the system is 350 bar.
Figure 1. Locations in a typical LC flow path and the expected pressure readings with a total backpressure of 350 bar
For the description and expected pressure at the numbered locations, see Table 1.
Pressure sensor |
Location Description |
Pressure reading |
1 |
In the Pump |
350 |
2 |
From Pump before in line filter |
348 |
3 |
After the inline filter and before Sampler |
345 |
4a |
At Sampler while in Mainpass |
339 |
4b |
At Sampler while in Bypass |
327 |
5 |
After Sampler and before in line filter |
322 |
6 |
After in line filter and before Guard column |
319 |
7 |
After Guard column and before Analytical Column |
304 |
8 |
After Analytical Column and before Detector |
8 |
9 |
After Detector and to Waste |
3 |
Some useful considerations from the pressure readings.
Learn how to effectively operate your Agilent 1260 LC System:
Agilent 1260 Infinity LC systems with OpenLab CDS eLearning path available from Agilent education |