This Information Applies To: Agilent ICP-MS systems (7700/7800/7850/7900/8800/8900)
Issue
The torch, bonnet, and shield plate must be removed and inspected during routine maintenance, and when troubleshooting plasma ignition problems. Faults with the torch, bonnet, or shield plate can cause plasma ignition problems. If the plasma will not light, remove these components, inspect, and replace any damaged parts before trying the parafilm test.
Items Required
Steps to Follow
|
Warning: The torch remains hot after the plasma is turned off. Do not touch the torch box for 10 minutes after turning off the plasma to let them cool. |
Figure 4. Spray chamber block thumb screw loosened and spray chamber slid to the left
Figure 5. Removal of plasma and auxiliary gas lines from the torch
Figure 6. Removal of torch, bonnet, and platinum shield plate
Figure 8. Damaged platinum shield plates
1. Shield plate torn in half, 2. Scorch marks
Tips: For additional torch maintenance and plasma ignition troubleshooting, see the folloing articles:
How to Clean the ICP-MS Torch ICP-MS Parafilm Test for Plasma Ignition Problems |