Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
  • Home
  • Technical Areas
    Technical Areas
    • AssayMap
    • Atomic Spectroscopy
    • Automated Electrophoresis
    • Cell Analysis
    • Consumables
    • CrossLab Connect
      CrossLab Connect
      • Service Management
      • Smart Alerts
    • Dissolution
    • GC
    • GC/MS
    • iLab Operations Software
    • LC
    • LC/MS
    • Life Sciences Reagents
    • Molecular Spectroscopy
      Molecular Spectroscopy
      • Molecular Spectroscopy - UV-Vis & UV-Vis-NIR, FTIR, Raman & Fluorescence
      • 8700 LDIR User Group
      • Transmission Raman User Group
    • Real-Time PCR (qPCR)
    • Sample Preparation
    • Software
      Software
      • Chromatography Software
        OpenLab CDS, OpenLab ChemStation, OpenLab EZChrom
      • Data Management Software
        OpenLab ECM, OpenLab ECM XT, OpenLab Server
      • Mass Spectrometry Software
        Data Analysis, Data Acquisition, Productivity App, WalkUp Software
    • Vacuum
  • Applications Areas
    Applications Areas
    • Basic Research
    • Energy and Chemicals
    • Environmental
    • Food and Agriculture
  • Agilent Knowledge Portal
  • Links
    Links
    • Collection of Support Resources
    • Webinar Notifications
    • Contact Support
    • User Contributions
  • Getting Started
    Community Help and Resources
  • Browse
    Browse
    • People
    • Places
    • Your View
    • Your Community Inbox
ICP-MS Portal
  • Agilent Knowledge Portal
ICP-MS Portal
Articles How to Install the ICP-MS Torch, Bonnet, and Platinum Shield Plate
  • Articles
  • Videos
  • Forum
  • Announcements
  • More
  • Cancel
  • Agilent Knowledge Portal
  • ICP-MS Portal Home
  • ICP-MS Portal Articles
  • ICP-MS Portal Videos
  • More
  • Cancel
  • Created :  7 Sep 2023
  • Modified :  11 Oct 2023
  • Category :  Agilent Knowledge Portal
  • Entry Type :  Article
  • Product Type :  ICP-MS Systems
  • Component :  GC/MS Triple Quadrupole ICP-MS Single Quadrupole
  • Product Name :  7700 ICP-MS 7800 ICP-MS 7850 ICP-MS 7900 ICP-MS 8800 Triple Quadrupole ICP-MS 8900 Triple Quadrupole ICP-MS
  • Task :  Maintenance
Related Articles
  • How to Remove and Inspect the ICP-MS Torch, Bonnet, and Platinum Shield Plate

    This Information Applies To: Agilent ICP-MS systems (7700/7800/7850/7900/8800/8900) Issue The torch, bonnet, and shield plate must be removed and inspected during routine maintenance, and when troubleshooting plasma ignition problems. Faults with the torch, bonnet, or shield plate can cause plasma ignition problems. If the plasma will not light, remove these components, inspect, and replace any damaged parts before trying the parafilm test . Items Required Standard Torch, quartz, 2...
  • ICP-MS Parafilm Test for Plasma Ignition Problems

    This Information Applies To: Agilent ICP-MS systems (7700/7800/7850/7900/8800/8900) Issue This procedure tests for an air leak in the sample introduction system, or fault with spray chamber draining, by sealing the end of the torch with a piece of parafilm and testing plasma ignition. Sealing with parafilm isolates the torch from the rest of the sample introduction system. If the plasma ignites with the parafilm in place, the ICP-MS hardware is operational and the plasma ignition problem is...
  • Routine Maintenance 700 Series ICP-OES Systems

    This Information Applies To: Agilent 700 Series ICP-OES Systems Issue The 700 Series ICP-OES requires regular maintenance, and routine maintenance tasks are able to be performed by the instrument operator. Any maintenance procedures not specifically mentioned in this article should only be performed by Agilent-trained, Agilent-qualified, or Agilent-authorized Field Service Engineers. Warning : Eye Hazard Danger to eyes. The plasma is an intense light source...
  • Cleaning the ICP-OES Nebulizer

    This Information Applies To: Agilent 500 Series ICP-OES (5100, 5110, 5800, 5900) Issue The ICP-OES nebulizer converts the liquid sample into an aerosol as it enters the spray chamber. A dirty or partially blocked nebulizer traps sample, which leads to poor reproducibility and reduced sensitivity. Droplets on the internal surface of the spray chamber can indicate poor conversion of liquid to mist. ICP-OES overview The ICP-OES sample introduction components are detailed in Figure 1 . ...
  • Maintaining the Axial Pre-Optic Window

    This Information Applies To: Agilent 5100, 5110, 5800, 5900 ICP-OES Issue The axial pre-optic window needs checking from time to time to ensure it is clean. If the axial pre-optic window is dirty, it must be cleaned or replaced to ensure adequate sensitivity in the UV region of the emission spectrum. If the axial wavelength calibration fails for wavelengths less than 190 nm, this may indicate that the axial pre-optic window is dirty. Steps to follow Follow the steps in this article...

How to Install the ICP-MS Torch, Bonnet, and Platinum Shield Plate

Answer

This Information Applies To: Agilent ICP-MS systems (7700/7800/7850/7900/8800/8900)


Issue

The torch, bonnet, and platinum shield plate must be installed after performing maintenance, or replacing with new components. 


Items Required

  • Standard Torch, quartz, 2.5 mm id - not suitable for samples containing hydrofluoric acid. (p/n G3280-80053)
    see the Agilent Spectroscopy Supplies Catalog if you use a non-standard torch
  • Bonnet (p/n G1833-65421)
  • Platinum Shield Plate (p/n G1833-65419)

Steps to Follow

  1. Install the platinum shield plate over the end of the torch, ensuring the alignment hole in the shield fits over the torch projection (Figure 1).



    Figure 1. Shield plate installed on torch

  2. Slide the bonnet over the platinum shield plate with the lip facing the base of the torch. The lip of the bonnet will rest against the shield plate torch projection (Figure 2).



    Figure 2. Bonnet installed on shield plate and torch

  3. Guide the end of the torch through the work coil being careful to not drop the bonnet, and ensure the torch alignment projection fits into the torch stand groove (Figure 3).



    Figure 3. Torch installed on torch stand

  4. Lock the torch in place with the torch stand clip (Figure 4).



    Figure 4. Torch stand clip latched

  5. Connect the plasma and auxiliary gas lines to the torch and then tighten the connectors by hand until finger tight (Figure 5). 



    Figure 5. Gas lines connected and tightened

  6. Slide the spray chamber block to the right side, to connect the torch and connection tube and tighten the thumbscrew on the bottom of the spray chamber block (Figure 6).



    Figure 6. Spray chamber block slid to the right and thumb screw tightened

  7. Install the clamp that connects that torch and connection tube (Figure 7).



    Figure 7. Torch and transfer tube clamped

  8. Install the torch box cover and secure the torch box catch clip (Figure 8).


    Torch box cover installed and latched.
    Figure 8. Torch box cover installed and latched

  9. Close the ICP-MS small cover (Figure 9).


    Small cover closed.
    Figure 9. Small cover closed

 

  Learn how to effectively maintain your Agilent ICP-MS System:
Agilent 7800/7850/7900 ICP-MS with MassHunter
Agilent 8900 Triple Quadrupole ICP-MS with MassHunter e-learning courses available from Agilent education