This Information Applies To: Agilent ICP-MS systems (7700/7800/7850/7900/8800/8900)
Issue
Error messages related to water flow or water valve issues can cause the plasma to shut off or disable plasma ignition. While a warning message will still allow the plasma to light and stay on, it is necessary to correct the issue.
Requirements
This troubleshooting procedure is applicable for the following warnings/errors:
Add the following meter to the ICP-MS MassHunter software
Items Required
Agilent Cool Clear (p/n 5799-0037)
Steps to Follow
Figure 1. Rear of G3292A Chiller or G1879B Heat Exchanger
1. Fluid level indicator 2. Reservoir fill port with cap removed
Figure 2: G8496A Chiller
1. Fluid level OK (blue), 2. Fluid level low (orange), 3. Fluid level extremely low (red) 4. Reservoir fill port with cap removed
Figure 4: Sample Introduction Maintenance Menu
1. Enable Temperature Control check box selected, 2. Water RF/WC/IF meter reading
Learn how to effectively troubleshoot your Agilent ICP-MS System:
Agilent 7800/7850/7900 ICP-MS with MassHunter Agilent 8900 Triple Quadrupole ICP-MS with MassHunter e-learning courses available from Agilent |