Hi Tenamyd
Maybe a bit more detail would help to get you an answer. Here's a few questions:
Is the method gradient or isocratic
Do you prepare the mobile phase freshly each day
Is the column maintained at a set temperature using a column oven…
You will need to plot the piston movement as well. The upper plateau is the secondary piston moving. If the spike is at the beginning of the plateau, then it is more likely the Outlet ball valve. If the spike is at the end of the plateau, it is more likely…
Hello,
Sorry for the late reply. Thank you so much, that actually worked perfectly and I can now produce a report template fit for our purpose. Thanks again and I really appreciate the assistance.
This shows where to drill a 2.5" (6.35 cm) hole in the bench to be able to pass the rough pump hose fittings through to be put the rough pump on a shelf below the bench. Information on how to vent the rough pump to a lab exhaust system is also included…
Putting the GCMS rough pump under the bench SQ and TQ 31Jan2024 with DE.pdf
Hi Mike,
I followed your advice but have not been successful getting any of these methods to work. When I try the first method, I get an error message at the bottom of the screen "Syntax Error in cmd/fct EVALUATE" that prevents the data from being exported…
Hello Tom,
Thank you so much for your reply. This is really helpful. I am running MH V4.6. If you could send me the screen captures for this version would be greatly appreciated!
Cheers,
Inês
My Green Lab’s ACT labelling - Accountability, consistency, and transparency
There are many reasons why I like working at Agilent. One of those is Agilent’s progress towards sustainability and ultimately a zero-carbon future. Collaborating with My Green…
The Digest of upcoming webinars has been updated as of January 31.
Check Out:
Part 2: The Mining Lifecycle and Elemental Analysis Workshop Series on February 6
Acceptability and Antioxidant Properties of Vit D and Saskatoon Berry Fortified Rooibos…
Hello,
You can view the completed injections in a sequence in the reprocessing session(offline) of the instrument. If you are talking about viewing the data for the in-progress injection, you will need to take a snapshot. I did not have anything running…
This Information Applies To: Agilent GC systems with Gas Clean Filter
Issue
If you have determined that the Gas Clean Filter is saturated, it mus be replaced (see the article When to Replace the Gas Clean Carrier Gas Filter ).
The steps described...
This Information Applies To: Agilent GC Systems 8890/8860/Intuvo 9000 /7890/7820/6850/6890
Issue
General guidance on GC column flow rates, outlining the correct column flow rate for optimal compound separation depends on the column internal diameter...
This Information Applies To: Agilent GC Method Translator
Issue
Detailed instructions on translating methods using the GC Method Translator.
Background
For ideal use of cases and applications with the GC Method Translator, see the article...
This Information Applies To: Agilent Solvent Vent Calculator
Issue
The Solvent Vent Calculator estimates the Agilent solvent vapor exit (SVE) accessory valve timing based on solvent type, oven temperature, vent flow, and pressure.
Background...
This Information Applies To: Use of the Agilent GC Method Translator
Issue
How to correctly use the GC Method Translator to convert a method to suit a different GC configuration or to optimize a method.
Background
The GC Method Translator...
This Information Applies To: GC Pressure Flow Calculator
Issue
Instructions for the correct use of the Pressure Flow Calculator.
Background
The Pressure Flow Calculator uses the column dimensions, oven temperature, and detector outlet pressure...
This Information Applies To: The Agilent GC Vapor Volume Calculator.
Issue
The Vapor Volume Calculator determines the vapor volume based on the injection volume, solvent type, inlet head pressure, and the inlet temperature. This calculated vapor...
This Information Applies To: Agilent GCs
Issue
Agilent provides free GC calculators and tools to assist with method development.
Background
Tips : The calculators are available for downloading at Agilent.com .
Vapor Volume...
This Information Applies To: Agilent GC Systems
Issue
The Agilent Gas Clean Carrier Gas Filter delivers faster stabilization times for increased GC productivity, removes oxygen, moisture, and hydrocarbons from the carrier gas for MS applications...
This Information Applies To: Agilent GC systems
Issue
When a GC heater cannot reach setpoint, or taking too long to reach setpoint, the GC will enter a "Thermal Shutdown" (or similar heater error) state to protect itself. This state will turn off...