This Information Applies To: Agilent GC Systems 8890/8860/Intuvo 9000 /7890/7820/6850/6890
Issue
General guidance on GC column flow rates, outlining the correct column flow rate for optimal compound separation depends on the column internal diameter.
The column flow rates in Table 1 are typical for all column temperatures. To further refine suitable flow rates for the particular application, see this article on how to use Agilent GC Calculators.
Column type | Column size | Carrier flow rate, mL/min | ||
Hydrogen | Helium | Nitrogen | ||
Packed | 1/8-inch | Do not use | 30 | 20 |
1/4-inch | Do not use | 60 | 40 | |
Capillary | 0.05 mm id | 0.5 | 0.4 | NA |
0.10 mm id | 1.0 | 0.8 | NA | |
0.20 mm id | 2.0 | 1.6 | 0.25 | |
0.25 mm id | 2.5 | 2.0 | 0.5 | |
0.32 mm id | 3.2 | 2.6 | 0.75 | |
0.53 mm id | 5.3 | 4.2 | 1.5 |
Table 1: Recommended column flow rates