At 310° C ion source temperature, the 502 will always be lower than typical. Also - please reduce the quad temperature down to 150° or maybe as high as 165° C... High quad temperature reduces the lifetime of the quad and only very few applications need…
Hi Andreja,
For calculating area % of two different peaks you must set one baseline left and one baseline right from the peaks. Then you have to enable "Information Pane" when clicking "View" in the menu bar and the results are shown as given in the…
Hi,
Yes it is right, please orientate on the last peak before the impurities start.
The exact baseline settings can be seen when enabling the peak information section below the analysis view (highlighted in green):
Manual integration limits…
We were able to configure the plate, we are good for now.
The 1290 Infinity II we have is coupled to Sciex 6500 MS and is controlled by Analyst Software using the Analyst Device Driver.
Hello,
No, you cannot use classic report styles *.frp from Chemstation in the CDS 2.x software. You can use any IR template built in Chemstation in CDS 2.x by copying the .rdl files over to the new project folders.
Marty
The software (except for MicroLab Pharma) comes with a 90-day trial period before the software needs to be registered. If you get in contact with our sales team, they should be able to get you a trial copy of the software. Find the appropriate contact…
Hello, Is it possible to get the Intelligent rinse working if the rinses occur after the analysis. Let's say I have 2min analysis time and I changed the Probe rinse of samples and stds to 120s and the Probe rinse 1-3 after the initial rinse. Then the…
The Digest of upcoming webinars has been updated as of June 23.
Notable additions include:
Cell Analysis (1)
Spectroscopy (1)
Check Out:
Elemental Analysis in Yeast Cells and Microplastics by ICP-MS with Automated Micro-Flow Sample Introduction…
A 5% H2 blend should work also. If you're using multiple calibration standards, make sure the range of the standards matches or exceeds the expected range in your samples.
One thing to watch for: I have seen instances where H2 saturates the TCD at high…
This Information Applies To: Agilent 4200, 4150 and 2200 TapeStation.
Issue
Is it possible to analyze single stranded DNA (ssDNA) on the TapeStation? Which kit is the most suitable for analysis of ssDNA?
Resolution
Agilent does not offer any...
This Information Applies To: Agilent 4200 and 4150 TapeStation systems
Issue
How to Run the software installation verification on 4200 and 4150 TapeStation systems.
Background
The 4150 TapeStation model G2992A and 4200 TapeStation model G2991B...
This Information Applies To: Agilent 5200/5300/5400 Fragment Analyzer, and Femto Pulse
Issue
Perform a capillary alignment when a new array is installed in your Agilent Parallel Capillary Electrophoresis instrument. An alignment may also be performed...
This Information Applies To: Agilent 4200 and 4150 TapeStation
Issue
Are the Agilent DNA and RNA ScreenTape assays denaturing?
Resolution
The DNA ScreenTape assays in the TapeStation portfolio are not denaturing. Only the RNA and High Sensitivity...
This Information Applies To: Agilent LC pumps with manual purge valve.
Issue
A good practice before using a system is to purge the system to remove air. This is done by observing the clear tubing from the solvent bottle as well as the waste line...
This Information Applies To: ICP-MS MassHunter 4.x and 5.x
Issue
Unable to open a batch in offline data analysis, or the batch is empty.
If the ICP-MS MassHunter software closes unexpectedly, there is a power outage, or the ICP-MS and computer...
This Information Applies To: Agilent OpenLab CDS versions 2.4 or higher.
Issue
How to preview result files using various Report Templates in OpenLab CDS versions 2.4 or higher.
Requirements
The user has the result set loaded in OpenLab CDS...
This Information Applies To: ICP-MS MassHunter with Spectroscopy Database Administrator (SDA) and Protect Local Data ON.
Issue
An 'Authentication failed' error message appeared after entering user name and password when attempting to download a...
This Information Applies To: ICP-MS MassHunter with Spectroscopy Database Administrator (SDA) and Protect Local Data On.
Issue
With Spectroscopy Database Administrator (SDA) and Protect Local Data turned on, batches are not accessible through file...
This Information Applies To: Agilent AriaMx real-time PCR system
Issue
Scratches on AriaMx optical modules can lead to a reduction in their performance.
Background
Scratches on an AriaMx optical module are shown in Figure 1 .
Figure 1 ....