No. There is a QTOF specific collision cell EPC module. The software needs to recognize it.
G3850-60039 | Agilent EPC module
G3470-60550 | Agilent , Fitting and 1/16" tubing to go from the EPC module to the QTOF CC 1/16" Swagelok connection.
Is…
Durjoy123 would you please start a new forum post with your questions? Please include as much useful information as possible - instrument model number and age of the system, MassHunter Acquisition software version, which high vacuum pump type, the vacuum…
Hello,
I am going to move this post to the GC forum as they will have a better chance of helping. I assume you want to know how to change the front panel language to some other language.
Marty
Hi Kenforce,
In this case, I would run the pressure decay test, as you could have a small leak that is causing the inconsistency. I would also consider increasing your oven equilibration time by a 0.5min or so, just to ensure the column phase is at…
Hi Marty,
We have a FID and an MS connected to the same system. We only want to acquire the signal of FID, but if we want to report the method parameters we are “forced” to acquire both (we do not really know why and we have already ask the question…
Sorry for the delayed response. I've found that either works fine. I typically try to keep injection volume low (0.2-1 uL for 1uM protein). I've mostly been using 50mM HEPES 150mM NaCl and get great deconvolution results, even with a 3 min short method…
I am getting message "draw command failed, EE34205" while running batch in G7129B infinity 2 autosampler. I have tried replacing transport assembly, but error is same. I have used another module G7129A Infinity autosampler with this LC, but same error…
This Information Applies To: Agilent 4200, 4150 and 2200 TapeStation.
Issue
Is it possible to analyze single stranded DNA (ssDNA) on the TapeStation? Which kit is the most suitable for analysis of ssDNA?
Resolution
Agilent does not offer any...
This Information Applies To: Agilent 4200 and 4150 TapeStation systems
Issue
How to Run the software installation verification on 4200 and 4150 TapeStation systems.
Background
The 4150 TapeStation model G2992A and 4200 TapeStation model G2991B...
This Information Applies To: Agilent 5200/5300/5400 Fragment Analyzer, and Femto Pulse
Issue
Perform a capillary alignment when a new array is installed in your Agilent Parallel Capillary Electrophoresis instrument. An alignment may also be performed...
This Information Applies To: Agilent 4200 and 4150 TapeStation
Issue
Are the Agilent DNA and RNA ScreenTape assays denaturing?
Resolution
The DNA ScreenTape assays in the TapeStation portfolio are not denaturing. Only the RNA and High Sensitivity...
This Information Applies To: Agilent LC pumps with manual purge valve.
Issue
A good practice before using a system is to purge the system to remove air. This is done by observing the clear tubing from the solvent bottle as well as the waste line...
This Information Applies To: ICP-MS MassHunter 4.x and 5.x
Issue
Unable to open a batch in offline data analysis, or the batch is empty.
If the ICP-MS MassHunter software closes unexpectedly, there is a power outage, or the ICP-MS and computer...
This Information Applies To: Agilent OpenLab CDS versions 2.4 or higher.
Issue
How to preview result files using various Report Templates in OpenLab CDS versions 2.4 or higher.
Requirements
The user has the result set loaded in OpenLab CDS...
This Information Applies To: ICP-MS MassHunter with Spectroscopy Database Administrator (SDA) and Protect Local Data ON.
Issue
An 'Authentication failed' error message appeared after entering user name and password when attempting to download a...
This Information Applies To: ICP-MS MassHunter with Spectroscopy Database Administrator (SDA) and Protect Local Data On.
Issue
With Spectroscopy Database Administrator (SDA) and Protect Local Data turned on, batches are not accessible through file...
This Information Applies To: Agilent AriaMx real-time PCR system
Issue
Scratches on AriaMx optical modules can lead to a reduction in their performance.
Background
Scratches on an AriaMx optical module are shown in Figure 1 .
Figure 1 ....