At 310° C ion source temperature, the 502 will always be lower than typical. Also - please reduce the quad temperature down to 150° or maybe as high as 165° C... High quad temperature reduces the lifetime of the quad and only very few applications need…
Hi Andreja,
For calculating area % of two different peaks you must set one baseline left and one baseline right from the peaks. Then you have to enable "Information Pane" when clicking "View" in the menu bar and the results are shown as given in the…
Hi,
Yes it is right, please orientate on the last peak before the impurities start.
The exact baseline settings can be seen when enabling the peak information section below the analysis view (highlighted in green):
Manual integration limits…
We were able to configure the plate, we are good for now.
The 1290 Infinity II we have is coupled to Sciex 6500 MS and is controlled by Analyst Software using the Analyst Device Driver.
Hello,
No, you cannot use classic report styles *.frp from Chemstation in the CDS 2.x software. You can use any IR template built in Chemstation in CDS 2.x by copying the .rdl files over to the new project folders.
Marty
The software (except for MicroLab Pharma) comes with a 90-day trial period before the software needs to be registered. If you get in contact with our sales team, they should be able to get you a trial copy of the software. Find the appropriate contact…
Hello, Is it possible to get the Intelligent rinse working if the rinses occur after the analysis. Let's say I have 2min analysis time and I changed the Probe rinse of samples and stds to 120s and the Probe rinse 1-3 after the initial rinse. Then the…
The Digest of upcoming webinars has been updated as of June 23.
Notable additions include:
Cell Analysis (1)
Spectroscopy (1)
Check Out:
Elemental Analysis in Yeast Cells and Microplastics by ICP-MS with Automated Micro-Flow Sample Introduction…
A 5% H2 blend should work also. If you're using multiple calibration standards, make sure the range of the standards matches or exceeds the expected range in your samples.
One thing to watch for: I have seen instances where H2 saturates the TCD at high…
This Information Applies To: Agilent 6820, 6890, 7820A GC systems with a Split/Splitless (SSL) inlet
Issue
Non-volatile sample matrix components will accumulate in the inlet liner and impact on analytical performance. The inlet liner should be changed...
This Information Applies To: Agilent headspace samplers (HS)
Issue
GC cycle time needs to be calculated correctly. Too short a GC cycle time will reduce sample quality and possibly stop the analysis prematurely. Too long GC cycle time can cause...
This Information Applies To: Agilent 7820A GC
Issue
7820A GC is designed with simplified keypad that comes with five buttons. They are Stop, Prep Run , Start, Up and Down arrow buttons. Changing the IP address requires a different method compared...
This Information Applies To: Agilent Single Needle Autosamplers
Background
Sometimes the user may observe liquid flowing from Port four on the autosampler injection valve. During an injection sequence, it is normal to have liquid from port four...
This Information Applies To: Agilent 4200, 4150, 2200 TapeStation systems
Issue
Can I run an assay using the ladder or sample buffer from a different assay if I have run out of the appropriate assay ladder or sample buffer? Is it possible to run...
This Information Applies To: Agilent ICP-MS 7700, 7800, 7850, 7900, 8800, and 8900 systems
Issue
Peristaltic pump tubing that has lost its elasticity will not deliver the sample to the nebulizer correctly. When the pump tubing appears limp or flattened...
This Information Applies To: Agilent AriaMx real-time PCR system
Issue
AriaMx touch screen displays error code 90 after the Instrument Health Check (IHC), which happens when the AriaMx is switched on, or after running the onboard diagnostic tests...
This Information Applies To: Agilent AriaMx real-time PCR system, AriaMx optical modules
Issue
AriaMx users can change which AriaMx optical modules are installed to upgrade the instrument capabilities or to troubleshoot.
Items Required
Any...
This Information Applies To: AriaMx real-time PCR system
Issue
Error 43 is displayed on the AriaMx touch screen. Error 43 is a voltage error and it normally indicates that the cartridge LED voltage or motor voltage being detected is abnormal. Sometimes...
This Information Applies To: Agilent 2100 Bioanalyzer system
Issue
Can I run qPCR samples containing Sybr® Green I on the Bioanalyzer?
Background
Sybr® Green I is an Intercalating Dye commonly used to detect amplification products in real...