We have a Duo 2.5 foreline pump that was replaced in June of this year (2023). The previous Duo 2.5 installed had a sudden catastrophic leak, thus the replacement in June. The oil was filled up halfway and I set a marker with tape to monitor the level…
Hi John, You are on the right track-- it sounds like your flame is not lit. The reading you are seeing on the Hydrogen line is likely back-pressure from the column flow. That result could indicate a blockage in the jet, but the flow is a little low to…
no you don't need to disconnect it from the FID, ,
bake the column for enough time usually more than 1 hour according to your column status , you can keep the FID working to monitor the baseline .
..
keep temperature 25 degrees less than the maximum…
Hi all, I am using a GC 6890 with split/splitless inlet to analyse reaction products. However, a problem has recently happened when the GC is running. I have noticed the appearance of spikes in the baseline of the FID detector. The spikes appear occasionally…
It looks like it is having to clear something out the inlet, column, detector or head space as the baseline rise is over most of it. I see it's a FID you are using, how long do you leave it ignited before running the first analysis?
Hi, I have a agilent GC 7890B with FID. currently i am a facing a problem regarding this GC, problem is like that - all gases supply is ok, no leakage is found in detector portion. i have cleaned the FID & Jet but flame is not burning. So kindly suggest…
Hi all,
I'm using a 0.18 mm HP-5 GC Column on a Shimadzu GC-2014 connected with a FID. I'm looking to get new ferrules (the old ones had been passed down from previous users) and wanted to discuss some options. I tried using some 0.5 mm graphite ferrules…
good day!
tell my please..
I have the air generator and his out pressure consist 52 psi, and and air consumption is 1000 ml/min. the EPC module GC 7890b work in rang 55-80psi. I have 2 detector FID. Tell my - with this generatorom may be worked ?…
Hi byoo,
Do you get any failed to communicate or any other message when you start up the software? You are correct if the busy light should be constantly flashing. This means you got a successfuk firmware load from the PC from ethernet connection. Can…
Hello, im having a problem with a 2d GC using the Deans Switch system. Its a 8890 GC with 2 FIDs. The problem is that, when cutting peaks from the 1st dimension to the 2nd dimension (2nd column), all peaks on the 2nd dimensions get split into 2 peaks…
This Information Applies To: Agilent Atomic Absorption Spectroscopy Systems (240/280 Series Flame AA and 240/280Z/Zeeman/GTA 120 Graphite Furnace)
Issue:
This article includes recommended maintenance procedures to be performed daily, weekly or annually...
This Information Applies To: Agilent SpectrAA Software, Agilent 240/280 Flame Atomic Absorption Systems, 240/280 Zeeman Graphite Furnace Systems
Issue:
How to Reset Lamp Usage Counters or Add a New Lamp in SpectrAA
Caution : All physical...
This Information Applies To: Agilent GTA 120, Agilent 240Z AA Spectrometer, Agilent 280Z AA Spectrometer, Agilent 240FS/GTA120/PSD120/UltrAA System
Issue:
It is important to align the sample probe of the Programmable Sampler Dispenser (PSD) 120...
This Information Applies To: Agilent Flame AA Systems: e.g., 55B, 240FS, 280FS
Issue:
The Flame AA burner and burner components require regular cleaning to maintain system performance. Components include the burner, nebulizer, and spray chamber...
This Information Applies To: Agilent Flame Atomic Absorption (AA) Spectrometers (50/55/240/280 AA)
Issue:
How to clean a blocked nebulizer of a Flame AA, which may be causing low sensitivity or a sample uptake rate below 5 ml/min.
Steps to follow...
This Information Applies To: Agilent Graphite Furnace Systems (PSD 120/GTA 120/240Z AA/280Z AA)
Issue
Maintenance procedure for replacing the syringe and plunger of the PSD 120 sample dispenser for the Agilent GTA and Zeeman Graphite Furnace Systems...
This Information Applies To: Agilent AA (AA240, AA280) with Graphite Furnace Accessory (GTA)
Issue:
The Programmable Sample Dispenser (PSD) requires the sampling probe to be accurately aligned before the PSD can be used.
The PSD enables the...
This Information Applies To: All Agilent AA systems: 55B, 240FS, 280FS, 240Z, 280Z
Issue:
Optimizing the position of the lamp maximizes amount of light from the lamp passing through the optics to the detector, improving the signal-to-noise and the...
This Information Applies To: Agilent VGA 77, AA, VGA 77P, ICP
Issue:
The Agilent Vapor Generation Accessories VGA 77 and VGA 77P use high concentrations acids. The system needs to be thoroughly cleaned after each use.
The Vapor Generation Accessory...
This Information Applies To: All Agilent Atomic Absorption Spectrometry Systems (AA50, AA55, AA240, AA280)
Issue: This article describes when and how to clean an AA burner.
Background
For optimum and safe performance, burners (see Figure 1 ...