1 uL is injected for PAH.
We do change the liner often and are using the one from all the app notes (with glass wool). We haven't tried a dimpled but do use a dimpled with the PBDE method.
After doing a bunch of troubleshooting over the last few…
I'd like to bump this thread up to the top because in our lab we have had the same problem, and our amazing IT guy was able to solve it!
It turned out that on the PC where the Data Analysis (same version as olabanalyst ) was faulty, it had been installed…
Hi everyone. There are some doubts that I would like to clarify because we're experiencing some drift and sensitivity problems and it may help.
We've already cleaned the cones, lenses and sample introduction system, including the torch.
1) There is…
The peak to the left in the Lithium window is Li6. This is from internal standard solution. Li6 abundance is typically very low and should not be larger than the Li7 peak. When you run start the plasma, is the internal standard (ISTD) line in a nitric…
Hi kalle-h,
Has the method worked well for you in the past? Any changes recently like column trims? Sometimes small pieces of column can get caught inside the modulator fittings and impede flow, which has negative effects on our data.
If your peaks…
Hello brandon.mendoza.ardena ,
It would be hard to tell, not seeing it or hearing it in person. But a good start would be to make sure there is nothing in the carousel that might cause that (broken vials, caps, or various other debris).
Could also…
Hello Durjoy123 ,
The logic board for 7890A+ does not have the same part number as the 7890A. I wouldn't hazard a guess as whether it will "work" or not. It seems like it does in your case.
But as a general rule I would stick to the board part number…
Check out our new article titled Mastering Pressure Management in GPC/SEC Systems , which has been published in the April issue of the Column . The authors are Dr. Daniela Held and Dr. Michael Krämer .
This Information Applies To: Agilent Atomic Absorption Spectroscopy Systems (240/280 Series Flame AA and 240/280Z/Zeeman/GTA 120 Graphite Furnace)
Issue:
This article includes recommended maintenance procedures to be performed daily, weekly or annually...
This Information Applies To: Agilent SpectrAA Software, Agilent 240/280 Flame Atomic Absorption Systems, 240/280 Zeeman Graphite Furnace Systems
Issue:
How to Reset Lamp Usage Counters or Add a New Lamp in SpectrAA
Caution : All physical...
This Information Applies To: Agilent GTA 120, Agilent 240Z AA Spectrometer, Agilent 280Z AA Spectrometer, Agilent 240FS/GTA120/PSD120/UltrAA System
Issue:
It is important to align the sample probe of the Programmable Sampler Dispenser (PSD) 120...
This Information Applies To: Agilent Flame AA Systems: e.g., 55B, 240FS, 280FS
Issue:
The Flame AA burner and burner components require regular cleaning to maintain system performance. Components include the burner, nebulizer, and spray chamber...
This Information Applies To: Agilent Flame Atomic Absorption (AA) Spectrometers (50/55/240/280 AA)
Issue:
How to clean a blocked nebulizer of a Flame AA, which may be causing low sensitivity or a sample uptake rate below 5 ml/min.
Steps to follow...
This Information Applies To: Agilent Graphite Furnace Systems (PSD 120/GTA 120/240Z AA/280Z AA)
Issue
Maintenance procedure for replacing the syringe and plunger of the PSD 120 sample dispenser for the Agilent GTA and Zeeman Graphite Furnace Systems...
This Information Applies To: Agilent AA (AA240, AA280) with Graphite Furnace Accessory (GTA)
Issue:
The Programmable Sample Dispenser (PSD) requires the sampling probe to be accurately aligned before the PSD can be used.
The PSD enables the...
This Information Applies To: All Agilent AA systems: 55B, 240FS, 280FS, 240Z, 280Z
Issue:
Optimizing the position of the lamp maximizes amount of light from the lamp passing through the optics to the detector, improving the signal-to-noise and the...
This Information Applies To: Agilent VGA 77, AA, VGA 77P, ICP
Issue:
The Agilent Vapor Generation Accessories VGA 77 and VGA 77P use high concentrations acids. The system needs to be thoroughly cleaned after each use.
The Vapor Generation Accessory...
This Information Applies To: All Agilent Atomic Absorption Spectrometry Systems (AA50, AA55, AA240, AA280)
Issue: This article describes when and how to clean an AA burner.
Background
For optimum and safe performance, burners (see Figure 1 ...