Hello Lucas_S ,
Based on your follow-up explanation in this discussion -
(+) Macro for Batch Export of CSV in OpenLab CDS C.01.10 to a unique archive. - Forum - Chromatography Software - Agilent Community
I don't believe this macro could be easily…
For others who may find this interesting. Looking at the GPIB set up with NI MAX you see 2 'instruments' with primary addresses of 9 and 20. The instrument at '9' seems to work fine. Thus I can see the machine warming up and going through the wavelengths…
I am using an Agilent 6460 Triple Quad and have been dealing with a terribly low signal for the past few days, the signal appears at an abundance of 30 to 50 as a baseline only. I analyzed different compounds with different concentrations, the Tune mix…
Hi Sabbas, thank you for your answer. I'm sorry I couldn't have answered you sooner.
We're using Cary 3500 with Cary UV workstation version 1.3.5, so we couldn't perform like you said. We have to lower the threshold to find peaks and it is quite inconvenient…
Joining associations that specialize in ultra-high voltage can be beneficial for PenflexCorp. This may include associations of electrical or power engineers, as well as groups specializing in the design and standardization of equipment for high-voltage…
The Digest of upcoming webinars has been updated as of April 11.
The new Agilent Community Group, GPC/SEC Solutions Group is hosting a webinar series in partnership with Dow in April: A Journey through SEC, Light Scattering, and 2D-LC for Macromolecules…
The Background values are so called "off mass background".
This could indicate an electronics or contamination problem of the quad.
In your case, the values are still quite ok.
I would guess, one of the lenses has a setting which is causing this…
Hello everybody!
We have the issue that in our run (ca. 100 - 200 samples, 7 min each) for like 20 samples there is no signal, just the technical background noise. Not even the peak that is in the Blank appears. An the it sowly comes back over 2 samples…
This Information Applies To: Agilent Atomic Absorption Spectroscopy Systems (240/280 Series Flame AA and 240/280Z/Zeeman/GTA 120 Graphite Furnace)
Issue:
This article includes recommended maintenance procedures to be performed daily, weekly or annually...
This Information Applies To: Agilent SpectrAA Software, Agilent 240/280 Flame Atomic Absorption Systems, 240/280 Zeeman Graphite Furnace Systems
Issue:
How to Reset Lamp Usage Counters or Add a New Lamp in SpectrAA
Caution : All physical...
This Information Applies To: Agilent GTA 120, Agilent 240Z AA Spectrometer, Agilent 280Z AA Spectrometer, Agilent 240FS/GTA120/PSD120/UltrAA System
Issue:
It is important to align the sample probe of the Programmable Sampler Dispenser (PSD) 120...
This Information Applies To: Agilent Flame AA Systems: e.g., 55B, 240FS, 280FS
Issue:
The Flame AA burner and burner components require regular cleaning to maintain system performance. Components include the burner, nebulizer, and spray chamber...
This Information Applies To: Agilent Flame Atomic Absorption (AA) Spectrometers (50/55/240/280 AA)
Issue:
How to clean a blocked nebulizer of a Flame AA, which may be causing low sensitivity or a sample uptake rate below 5 ml/min.
Steps to follow...
This Information Applies To: Agilent Graphite Furnace Systems (PSD 120/GTA 120/240Z AA/280Z AA)
Issue
Maintenance procedure for replacing the syringe and plunger of the PSD 120 sample dispenser for the Agilent GTA and Zeeman Graphite Furnace Systems...
This Information Applies To: Agilent AA (AA240, AA280) with Graphite Furnace Accessory (GTA)
Issue:
The Programmable Sample Dispenser (PSD) requires the sampling probe to be accurately aligned before the PSD can be used.
The PSD enables the...
This Information Applies To: All Agilent AA systems: 55B, 240FS, 280FS, 240Z, 280Z
Issue:
Optimizing the position of the lamp maximizes amount of light from the lamp passing through the optics to the detector, improving the signal-to-noise and the...
This Information Applies To: Agilent VGA 77, AA, VGA 77P, ICP
Issue:
The Agilent Vapor Generation Accessories VGA 77 and VGA 77P use high concentrations acids. The system needs to be thoroughly cleaned after each use.
The Vapor Generation Accessory...
This Information Applies To: All Agilent Atomic Absorption Spectrometry Systems (AA50, AA55, AA240, AA280)
Issue: This article describes when and how to clean an AA burner.
Background
For optimum and safe performance, burners (see Figure 1 ...