High background for mass 7, 89 and 205 on tune (no gas mode) ICP-MS 7700

Hello all,

I have an issue with the tune in the batch on ICP-MS 7700 in no gas mode.

When I launch a Start-Up, the Backgroung on mass 7, 89 and 205 is good but not in tune version (very high compared last time).

For example, the tune in the batch (in no gas mode) indicate :

Mass 7 : 12937CPS, %RSD 3,9, Background 18,8
Mass 89 : 65696CPS, %RSD 2,4, Background 10,9
Mass 205 : 33990CPS, %RSD 3,4, Background 21,1

Oxide ratio 156/140 : 0,923%
Doubly charged 70/140 : 1,256%

In the past, the value for these mass were below 5 in background.
When I launch the performance report in the start-up (no gas mode too), these values are good.

For information, the cones and glassware have been cleaned  and new tubings were installed.

What is the problem with the high background in no gas mode ? 


Thanks in advance for your return,

Have a good day !

Parents
  • The Background values are so called "off mass background". 

    This could indicate an electronics or contamination problem of the quad.

    In your case, the values are still quite ok. 

    I would guess, one of the lenses has a setting which is causing this. 

    Have you tried, creating a new batch (from scratch) and tune it?

    Best Regards

    chris

Reply
  • The Background values are so called "off mass background". 

    This could indicate an electronics or contamination problem of the quad.

    In your case, the values are still quite ok. 

    I would guess, one of the lenses has a setting which is causing this. 

    Have you tried, creating a new batch (from scratch) and tune it?

    Best Regards

    chris