Hello,
I am expeirencing the same issue after having to reload E.02.02.1431. The quant report prints as one job and the chromatogram and the spectra reports (I'm printing a detailed report) print as a second job.
I've done as ron_tackett mentioned…
Instability in the anode voltage will cause baseline problems in your detector signal. For example, if the lamp doesn't light properly, the voltage will reset and you will see periodic changes in the baseline. I have seen old lamps sometimes produce odd…
I can't seem to get our new instrument to pass an EPA DFTPP tune for method 525.2. We recently got an instrument with the hydroinert source and I have only gotten a couple of passing tunes but nothing that is repeatable. I have tried manual tuning myself…
Yes! I checked the line too as according to the instructions (+) How to Disassemble and Assemble the HES (High Efficiency Source) - Articles - GC/MS Portal - Agilent Community
I will take a picture when I have them out next time. Thank you so much for…
Analyze Agilent Seahorse XF Real-Time ATP rate assay data using Agilent Seahorse Analytics
Analyze Agilent Seahorse XF Hu T Cell Activation Assay data using Agilent Seahorse Analytics software
Analyze Agilent Seahorse XF Substrate Oxidation Stress…
Hello APacana ,
The reason for the difference in response between the TIC and the EIC is because when you extract the TIC of a given precursor you are extracting the ion signal from all precursors that create a 72 product ion, using your example. If…
Hello AJohnson13 ,
Here is the PC setup guide for older Windows 10 versions of MassHunter. It includes the PC specifications.
Windows10_Setup.book (agilent.com)
The software installation instructions would be included as part of your original media…
At 1 mL/min pumping water, the pressures are as follows:
Pump + Column = 53 bar
Pump + Column + 0.1 mm ID 30 cm tubing = 70 bar
Pump + Column + Tubing + Flow Cell = 75 bar
I have also confirmed that this pressure fluctuation has nothing to do…
This Information Applies To: Agilent Atomic Absorption Spectroscopy Systems (240/280 Series Flame AA and 240/280Z/Zeeman/GTA 120 Graphite Furnace)
Issue:
This article includes recommended maintenance procedures to be performed daily, weekly or annually...
This Information Applies To: Agilent SpectrAA Software, Agilent 240/280 Flame Atomic Absorption Systems, 240/280 Zeeman Graphite Furnace Systems
Issue:
How to Reset Lamp Usage Counters or Add a New Lamp in SpectrAA
Caution : All physical...
This Information Applies To: Agilent GTA 120, Agilent 240Z AA Spectrometer, Agilent 280Z AA Spectrometer, Agilent 240FS/GTA120/PSD120/UltrAA System
Issue:
It is important to align the sample probe of the Programmable Sampler Dispenser (PSD) 120...
This Information Applies To: Agilent Flame AA Systems: e.g., 55B, 240FS, 280FS
Issue:
The Flame AA burner and burner components require regular cleaning to maintain system performance. Components include the burner, nebulizer, and spray chamber...
This Information Applies To: Agilent Flame Atomic Absorption (AA) Spectrometers (50/55/240/280 AA)
Issue:
How to clean a blocked nebulizer of a Flame AA, which may be causing low sensitivity or a sample uptake rate below 5 ml/min.
Steps to follow...
This Information Applies To: Agilent Graphite Furnace Systems (PSD 120/GTA 120/240Z AA/280Z AA)
Issue
Maintenance procedure for replacing the syringe and plunger of the PSD 120 sample dispenser for the Agilent GTA and Zeeman Graphite Furnace Systems...
This Information Applies To: Agilent AA (AA240, AA280) with Graphite Furnace Accessory (GTA)
Issue:
The Programmable Sample Dispenser (PSD) requires the sampling probe to be accurately aligned before the PSD can be used.
The PSD enables the...
This Information Applies To: All Agilent AA systems: 55B, 240FS, 280FS, 240Z, 280Z
Issue:
Optimizing the position of the lamp maximizes amount of light from the lamp passing through the optics to the detector, improving the signal-to-noise and the...
This Information Applies To: Agilent VGA 77, AA, VGA 77P, ICP
Issue:
The Agilent Vapor Generation Accessories VGA 77 and VGA 77P use high concentrations acids. The system needs to be thoroughly cleaned after each use.
The Vapor Generation Accessory...
This Information Applies To: All Agilent Atomic Absorption Spectrometry Systems (AA50, AA55, AA240, AA280)
Issue: This article describes when and how to clean an AA burner.
Background
For optimum and safe performance, burners (see Figure 1 ...