Update: FID output signal was very low even when trying to ignite, so we checked the interconnect spring and noticed it was shorter than the ones in the other GCs (not that close to the detector base; dont know if it was always like that or something…
Hi,
We are struggling to solve an issue with an Agilent 8860 GC-FID we purchased in 2019. It is used to quantify propane in headspace samples that are manually injected. It is possible but unlikely that some of those headspace samplings result in the…
Tip for other users that may encounter an FID Ignition problem: try blowing on the FID when it is trying to light. If the FID lights - there is an issue with the gas flows. It is best to use a flow meter like the ADM Flow Meter for gas chromatography…
Tip for other users that may encounter an FID Ignition problem: try blowing on the FID when it is trying to light. If the FID lights - there is an issue with the gas flows. It is best to use a flow meter like the ADM Flow Meter for gas chromatography…
Hey. Johnson we solved the Problem by removing the igniter Castor with collector nut and the spring locks Info the right place. Another Problem could be the malfunction electrometer.
Greetings Agilent Community,
As with many of the errors I encounter with my system, these two happened after I foolishly left the lab to pursue a life outside the world of LCMS.
(1) From the instrument log, the first error read "Failed to run acquisition…
When attempting to submit a sequence in EZChrom, it fails to submit and returns this error message:
This happens each time a sequence is submitted. Any help would be appreciated!
Hi Agilent community.
I'm a LC-MS equipment user and I had a problem with the MassHunter Acquisition B 08.01 software.
When I try to open the Acquisition software, appears the software logo but immediately appears an error dialog box. This error dialog…
Hi everyone,
We have a 6546 QTOF, the following error keeps popping up during acquisition; Failed to start logbook server. A shared violation has occurred. How can I resolve this?. I'll appreciate your help. Thanks
Performing a GC/MS Manual Injection Using MassHunter Software
Multiple Results
This Information Applies To: Agilent GC/MS systems using MassHunter software
Issue
Manual injections are needed for a GC/MS without an automated sampler. They are also used for troubleshooting specific issues such as no peaks, lower peaks than expected...
This Information Applies To: Agilent GC systems with Gas Clean Filter
Issue
If you have determined that the Gas Clean Filter is saturated, it mus be replaced (see the article When to Replace the Gas Clean Carrier Gas Filter ).
The steps described...
This Information Applies To: Agilent GC Systems 8890/8860/Intuvo 9000 /7890/7820/6850/6890
Issue
General guidance on GC column flow rates, outlining the correct column flow rate for optimal compound separation depends on the column internal diameter...
This Information Applies To: Agilent GC Method Translator
Issue
Detailed instructions on translating methods using the GC Method Translator.
Background
For ideal use of cases and applications with the GC Method Translator, see the article...
This Information Applies To: Agilent Solvent Vent Calculator
Issue
The Solvent Vent Calculator estimates the Agilent solvent vapor exit (SVE) accessory valve timing based on solvent type, oven temperature, vent flow, and pressure.
Background...
This Information Applies To: Use of the Agilent GC Method Translator
Issue
How to correctly use the GC Method Translator to convert a method to suit a different GC configuration or to optimize a method.
Background
The GC Method Translator...
This Information Applies To: GC Pressure Flow Calculator
Issue
Instructions for the correct use of the Pressure Flow Calculator.
Background
The Pressure Flow Calculator uses the column dimensions, oven temperature, and detector outlet pressure...
This Information Applies To: The Agilent GC Vapor Volume Calculator.
Issue
The Vapor Volume Calculator determines the vapor volume based on the injection volume, solvent type, inlet head pressure, and the inlet temperature. This calculated vapor...
This Information Applies To: Agilent GCs
Issue
Agilent provides free GC calculators and tools to assist with method development.
Background
Tips : The calculators are available for downloading at Agilent.com .
Vapor Volume...
This Information Applies To: Agilent GC Systems
Issue
The Agilent Gas Clean Carrier Gas Filter delivers faster stabilization times for increased GC productivity, removes oxygen, moisture, and hydrocarbons from the carrier gas for MS applications...