For glucose a 87H type is ok as it uses Ion Exclusion Chromatography for the separation, the phases contain benzenesulfonic acid group (and will interact with anions in ion exchange mode). Phosphoric acid does not interact with sulfonic acid groups as…
thanks for sharing the information. It saved my day. But to delete the mentioned file I had to call helpdesk to grant me permission. No way to delete with my regular credentials.
Agilent offers these two services for the IDP-7 Dry Scroll Pump: 1. The Agilent Rapid Exchange Service replaces your pump with a refurbished one • You can send your unit back (exchange) once the replacement unit arrives • Warranty period is 1 year from…
hello,
How to use the Add-ins Blank substraction in Mass Hunter V10.1 without a calibration curve? It seems like it is not possible without a calibration curve but I'm doing semi-quantitative work so I don't have a calibration curve.
Will that feature…
Hello afzishle ,
I'm not aware of any issues with PDF reporting. For troubleshooting I would try manually printing the report with preview enabled. That should give you some idea if it is an issue with saving the file or just report generation.
Hello,
The test and measurement equipment went to Keysight when it was spun off from Agilent in 2014. You would need to check with Keysight for information on that product.
Marty
Agilent Technologies, Inc. - Agilent Technologies Spins Off Its…
Ok, this indicates a network problem.
Please check the energy settings from windows, that the network card is never going to "sleep".
Also check, that no backup process is running, when masshunter is collecting data.
This Information Applies To: Agilent Liquid Chromatography (LC) systems
Issue:
Shutting down the HPLC for a long period requires following this general procedure to prevent damage to the system and the column.
Steps to follow:
Store the...
This Information Applies To: Agilent G7129A/B Autosampler with Sample Thermostat/Cooler G7167-60101 or G7167-60005.
Issue:
The autosampler may stop working from time to time, potentially giving off loud noises and switching off the light. The only...
This Information Applies To: Agilent 5977A/B GC/MSD with EI source, EI Extractor Source, or CI Source.
Issue:
During normal operation of the GC/MSD, the source will require cleaning. It is time to clean the source when:
There is a loss in analyte...
This Information Applies To: Agilent HES source, Agilent 5977 Series GC/MSD
Issue:
During normal operation of the GC/MSD, the source will require cleaning. It is time to clean the source when:
There is a loss in analyte response not improving...
This Information Applies To: Agilent 5977 and 5975 Series GC/MSD systems
Issue
The autotune fails and reports the error Tune action stopped: Gain and offset = 0 and peaks are too narrow!
Resolution
The MSD autotune program is run after routine...
This Information Applies To: Agilent ICP-MS systems (7700/7800/7850/7900/8800/8900)
Issue
Condition new or cleaned ICP-MS skimmer cones and sampler cones by aspirating a Conditioning Solution before use.
Background
Drift up can be a sign of...
Turn On the GC/MS Interface and Oven Using OpenLab CDS
Multiple Results
This Information Applies To: Agilent GC/MS instruments controlled by OpenLab CDS software
Issue
Turning on the GC/MS interface heater and Oven using OpenLab CDS software.
Steps to Follow
From the Home tab, click Method > Instrument Setup...
This Information Applies To: Agilent ICP-MS systems (7700/7800/7850/7900)
Issue
The signal monitor function in MassHunter can help with troubleshooting low sensitivity, carry over, contamination, or high background. It can also be used to optimize...
This Information Applies To: Agilent UV detectors (Variable Wavelength Detectors, Multiple Wavelength Detectors, and Diode Array Detectors)
Issue
Checking accumulated hours on the UV lamp.
Steps to Follow
There are four ways to check the accumulated...
This Information Applies To: Agilent 7700/7800/7850/7900/8800/8900 ICP-MS
Issue:
Poor internal standard (ISTD) mixing can cause high relative standard deviation (RSD).
The recommendations in this article are applicable if:
It takes a long...