Hello LC92 ,
Did you run the installer as an Administrator? That might cause issues like this.
Another thing to try is running the Software Verification Tool found in the Start menu under Agilent Technologies. See if there are any missing files. You…
Learn about how to check and extend the lifetime of peristaltic pump tubing; prevent and remove nebulizer blockages; clean the nebulizer and improve precision and reduce memory effects with spray chamber cleaning.
Preventative maintenance is crucial…
Learn more about accurate ICP-MS standard preparation, sample introduction, diagnosis of sample introduction issues, and preventive ICP-MS maintenance guidelines. In this video we provide simple steps to improve sensitivity and accuracy, precision, and…
If you have data associated with a worksheet you cannot make modifications to it. Depending on your software version, you will need to either do "Save As" and give it another name, or create a new worksheet from template. Then the new worksheet will not…
Hello. You have to adjust Nitrogen pressure to get the blue bar inside green area. Pressure regulator is located on the right side of the 4107. It should be ~ 600kPa.
Hello mipastor,
Thank you for your response. I think we do have problem with the MS2 heater. Agilent customer support hinted that it might be sensor problem or the board so service engineer will have to take a look.
The Digest of upcoming webinars has been updated as of April 23.
The new Agilent Community Group, GPC/SEC Solutions Group is hosting a webinar series in partnership with Dow in April: A Journey through SEC, Light Scattering, and 2D-LC for Macromolecules…
In the previous post, I talked about the merits of automated sampling in the dissolution workflow . Here, I will provide some information about how to validate your automated dissolution system, using the 850-DS Sampling station from Agilent as an example…
You might be able to use HPLC-SEC on those with an Aquagel-like column. Check these literature links for help:
https://doi.org/10.1016/0021-9673(71)80010-9
https://doi.org/10.1208/s12248-017-0126-0
You do this with a Restek column! You cannot get a better resolution by changing just parameters with this cphase. You need a different selectivity to have a chance, maybe use a DB-WAX UltraInert.
This Information Applies To: Agilent Liquid Chromatography (LC) systems
Issue:
Shutting down the HPLC for a long period requires following this general procedure to prevent damage to the system and the column.
Steps to follow:
Store the...
This Information Applies To: Agilent G7129A/B Autosampler with Sample Thermostat/Cooler G7167-60101 or G7167-60005.
Issue:
The autosampler may stop working from time to time, potentially giving off loud noises and switching off the light. The only...
This Information Applies To: Agilent 5977A/B GC/MSD with EI source, EI Extractor Source, or CI Source.
Issue:
During normal operation of the GC/MSD, the source will require cleaning. It is time to clean the source when:
There is a loss in analyte...
This Information Applies To: Agilent HES source, Agilent 5977 Series GC/MSD
Issue:
During normal operation of the GC/MSD, the source will require cleaning. It is time to clean the source when:
There is a loss in analyte response not improving...
This Information Applies To: Agilent 5977 and 5975 Series GC/MSD systems
Issue
The autotune fails and reports the error Tune action stopped: Gain and offset = 0 and peaks are too narrow!
Resolution
The MSD autotune program is run after routine...
This Information Applies To: Agilent ICP-MS systems (7700/7800/7850/7900/8800/8900)
Issue
Condition new or cleaned ICP-MS skimmer cones and sampler cones by aspirating a Conditioning Solution before use.
Background
Drift up can be a sign of...
Turn On the GC/MS Interface and Oven Using OpenLab CDS
Multiple Results
This Information Applies To: Agilent GC/MS instruments controlled by OpenLab CDS software
Issue
Turning on the GC/MS interface heater and Oven using OpenLab CDS software.
Steps to Follow
From the Home tab, click Method > Instrument Setup...
This Information Applies To: Agilent ICP-MS systems (7700/7800/7850/7900)
Issue
The signal monitor function in MassHunter can help with troubleshooting low sensitivity, carry over, contamination, or high background. It can also be used to optimize...
This Information Applies To: Agilent UV detectors (Variable Wavelength Detectors, Multiple Wavelength Detectors, and Diode Array Detectors)
Issue
Checking accumulated hours on the UV lamp.
Steps to Follow
There are four ways to check the accumulated...
This Information Applies To: Agilent 7700/7800/7850/7900/8800/8900 ICP-MS
Issue:
Poor internal standard (ISTD) mixing can cause high relative standard deviation (RSD).
The recommendations in this article are applicable if:
It takes a long...