Hello,
No, you cannot use classic report styles *.frp from Chemstation in the CDS 2.x software. You can use any IR template built in Chemstation in CDS 2.x by copying the .rdl files over to the new project folders.
Marty
The software (except for MicroLab Pharma) comes with a 90-day trial period before the software needs to be registered. If you get in contact with our sales team, they should be able to get you a trial copy of the software. Find the appropriate contact…
Hello, Is it possible to get the Intelligent rinse working if the rinses occur after the analysis. Let's say I have 2min analysis time and I changed the Probe rinse of samples and stds to 120s and the Probe rinse 1-3 after the initial rinse. Then the…
The Digest of upcoming webinars has been updated as of June 23.
Notable additions include:
Cell Analysis (1)
Spectroscopy (1)
Check Out:
Elemental Analysis in Yeast Cells and Microplastics by ICP-MS with Automated Micro-Flow Sample Introduction…
A 5% H2 blend should work also. If you're using multiple calibration standards, make sure the range of the standards matches or exceeds the expected range in your samples.
One thing to watch for: I have seen instances where H2 saturates the TCD at high…
Offsetting the axes, Y or X, is not a feature in MassHunter. You can do head to tail comparison.
Overlaying in different colors and choosing which one is active can be done but there's not a way to do what you're asking even in a custom report.…
Good morning. We analyze fatty acids and The compound response factor of each fatty acid before the softwere changed remained fixed now it changes. I tried to copy the various parameters set in the other gas chromatograph which has its own softwere, but…
This Information Applies To: Agilent Liquid Chromatography (LC) systems
Issue:
Shutting down the HPLC for a long period requires following this general procedure to prevent damage to the system and the column.
Steps to follow:
Store the...
This Information Applies To: Agilent G7129A/B Autosampler with Sample Thermostat/Cooler G7167-60101 or G7167-60005.
Issue:
The autosampler may stop working from time to time, potentially giving off loud noises and switching off the light. The only...
This Information Applies To: Agilent 5977A/B GC/MSD with EI source, EI Extractor Source, or CI Source.
Issue:
During normal operation of the GC/MSD, the source will require cleaning. It is time to clean the source when:
There is a loss in analyte...
This Information Applies To: Agilent HES source, Agilent 5977 Series GC/MSD
Issue:
During normal operation of the GC/MSD, the source will require cleaning. It is time to clean the source when:
There is a loss in analyte response not improving...
This Information Applies To: Agilent 5977 and 5975 Series GC/MSD systems
Issue
The autotune fails and reports the error Tune action stopped: Gain and offset = 0 and peaks are too narrow!
Resolution
The MSD autotune program is run after routine...
This Information Applies To: Agilent ICP-MS systems (7700/7800/7850/7900/8800/8900)
Issue
Condition new or cleaned ICP-MS skimmer cones and sampler cones by aspirating a Conditioning Solution before use.
Background
Drift up can be a sign of...
Turn On the GC/MS Interface and Oven Using OpenLab CDS
Multiple Results
This Information Applies To: Agilent GC/MS instruments controlled by OpenLab CDS software
Issue
Turning on the GC/MS interface heater and Oven using OpenLab CDS software.
Steps to Follow
From the Home tab, click Method > Instrument Setup...
This Information Applies To: Agilent ICP-MS systems (7700/7800/7850/7900)
Issue
The signal monitor function in MassHunter can help with troubleshooting low sensitivity, carry over, contamination, or high background. It can also be used to optimize...
This Information Applies To: Agilent UV detectors (Variable Wavelength Detectors, Multiple Wavelength Detectors, and Diode Array Detectors)
Issue
Checking accumulated hours on the UV lamp.
Steps to Follow
There are four ways to check the accumulated...
This Information Applies To: Agilent 7700/7800/7850/7900/8800/8900 ICP-MS
Issue:
Poor internal standard (ISTD) mixing can cause high relative standard deviation (RSD).
The recommendations in this article are applicable if:
It takes a long...