Hello blairchall ,
I’m sorry for the delayed reply to your message. Do you still need assistance with this? If you do, please take into consideration the notes below. However, if the issue was resolved, could you please share what the resolution was…
We use Agilent NGA Analyzer GC based on 4 column configuration. Hayesep Q pre column, Hayesep Analytical column, molsieve column for TCD and one capillary column for FID. The H2S elution takes place in the Hayesep analytical column. But now a days our…
H ztarbs,
The air would not be for an inlet supply. This is most likely a detector support gas for FID. Check your method to make sure that the FID is not on in a standby method. FID air consumption is 400 to 450ml/min when on.
Regards
James
Hello, all you need to do in the method is to uncheck each of the FID flows, and uncheck the electrometer and uncheck the Flame on. With everything unchecked for the FID settings the software will ignore your FID. Under the Signals section you will also…
Thank you for your response Today when you open the GC we noted the increase returned valid but not connected and the Remote LED off we tried to configuration instrument but a message appears(failed to connect to 6890 at 100.100.100.101)but the GC IP…
Hi again
try the last hope ,
first try to select your reporting method in ( DA report template ) , the sequence accept it but i don't know if it will success to print ,
you may try it
the second thing if the first failed ,then try to edit your…
I want to check Total hydrocarbon by FID without identifying the hydrocarbon species . Agilent suggest to give me DB-1 15m, 0.32mm, 0.10u as a non polar column to elute all HC without separation as one peak . Does this column will work for Total hydrocarbon…
Hi silencer2008,
the 1 µL/s rule means the injection speed used for splitless injections, which is very slow compared to split injection where you want to inject as fast as possible.
Two interesting articles would be:
(+) Pre-Columns - the forgotten…
We've been having issues with our 7890B gas chromatograph. We have it connected to a workstation running Windows 7 SP1 and OpenLab/Chemstation version C.01.07 [27]. We have a Gerstel MPS attached as a headspace sampler.
We have been trying to run…
This Information Applies To: Agilent 8890/8860 GC Systems with SSL, MMI, PTV, or VI inlets.
Issue
Split vent trap replacement is recommended as part of the scheduled preventive maintenance. Sometimes, it has to be replaced when the inlet pressure...
This Information Applies To: Agilent 6000 Series LC/MS systems equipped with MS40+ rotary vane pumps
Issue
Instructions on how to replace the filter in an Agilent MS 40+ pump as part of annual pump maintenance
Background
The filter in the...
This Information Applies To: Agilent ICP-OES 5000 Series (5100/5110/5800/5900) with ICP Expert 7.5.x Software
Issue
The Recommended Routine Maintenance Schedule does not adjust for the number of samples analyses or the sample matrix, both of which...
This Information Applies To: Agilent 7890/7820 GC Systems with SSL, MMI, PTV, or VI inlets.
Issue
Split vent trap replacement is recommended as part of the scheduled preventive maintenance. Sometimes, it has to be replaced when the inlet pressure...
This Information Applies To: Agilent LC pumps
Issue:
Stuck or obstructed check valves (Active Inlet Valve (AIV), Passive Inlet Valve (PIV), Outlet Ball Valve (OBV)) can cause issues such as no-flow or no-pressure, moderate to strong drops in pressure...
This Information Applies To: Agilent 1260 Infinity II multisampler, 1290 Infinity II multisampler, 1260 Infinity II SFC multisampler, and 1260 Infinity II bio-Inert multisampler
Issue
The needle bends while moving to the wash port or showing one...
This Information Applies To: Agilent OpenLab CDS 2.x
Issue
This article outlines the step-by-step instruction on how to overlay chromatograms for OpenLab CDS 2.x
Steps to Follow
Load data from result sets or single samples. See Load or add...
This Information Applies To: Agilent Lab Advisor Software, Agilent Multicolumn Thermostat
Issue:
Agilent Lab Advisor is a great tool to diagnose the issue of your HPLC. Running the tests that are relevant to your specific issue in the Service &...
This Information Applies To: Agilent OpenLab CDS (ver 2.x)
Content:
The following procedure shows how to review completed injections in a running sequence in OpenLab CDS (ver 2.x).
Steps to Follow:
Select the sequence from the Run Queue...