LC Articles How to Prepare your HPLC for a Long Time Shut Down
  • LC Articles
  • LC Videos
  • LC Forum
  • More
  • Cancel

Browse the Agilent Knowledge Portal

  • Chromatography
    • Gas Chromatography
    • Liquid Chromatography
  • More
  • Cancel
  • GC/MS
  • More
  • Cancel
  • Spectroscopy
    • Atomic Absorption
    • ICP-OES
  • More
  • Cancel
  • Automated Electrophoresis
  • More
  • Cancel
  • Software and Informatics
    • OpenLab CDS
    • MassHunter
  • More
  • Cancel
  • -Liquid Chromatography
    • +LC Operation
    • +LC Maintenance
    • +LC Troubleshooting
    • -LC Best Practices
      • How to prevent blockages on solvent filters
      • Good practices for use of the Multi-Channel Gradient Valve in Agilent quaternary pump modules
      • Mobile Phase Preparation for Isocratic LC Systems
      • General Column Regeneration Procedure
      • General Recommendations for Using an Agilent RID
      • How to Prepare your HPLC for a Long Time Shut Down
    • +LC Firmware

How to Prepare your HPLC for a Long Time Shut Down

Created by Sarah Scheirs Sarah Scheirs over 2 years ago | Last modified by Carlos Vargas Carlos Vargas over 1 year ago

This Information Applies To: Agilent Liquid Chromatography (LC) systems


Issue:

Shutting down the HPLC for a long period requires following this general procedure to prevent damage to the system and the column.


Steps to follow:

  1. Store the column in an appropriate solvent (check column information sheet)
  2. When you remove the column, add a column union so the detector gets flushed as well in the following steps.
  3. For a system in reversed phase -> Purge all lines with hot HPLC-grade water (60 to 70 °C) to remove salts or ACN polymeric by-products.
    You may use the following solvents for storing the instrument: 100% HPLC grade IPA (Isopropanol), IPA: Water mix > 10:1 or a (50:50) MeOH: Water mix. No salts should be left in the system.
  4. For a system in Normal phase, do not purge with water -> purge with IPA directly and then move on to store in 100% HPLC grade IPA or IPA: Hexane mix > 10:1.   
  5. Fill all solvent lines with IPA or the storing solvent.
  6.  Plug (or seal) all open connections.
  7. Emptying waste containers and mobile phase bottles is recommended to avoid any contamination later on. 
  8. Power off all modules and the PC / control software. 
  9. If you have an evaporative light scattering detector (ELSD), close all regulators and the gas tank. 

 Tip 
Learn how to maintain your Agilent LC system:
Collection of LC (HPLC) Resources in the Agilent Community

  • Best practices
  • Analytical LC
  • Share
  • History
  • More
  • Cancel
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel

Contact Agilent Support

 Still Need Assistance?

 Contact Agilent Support >

Related
Recommended for you
Privacy Statement
Terms of Use
Contact Us
Site Help