Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
Liquid Chromatography Portal
  • Agilent Knowledge Portal
Liquid Chromatography Portal
LC Articles How to Flush an HPLC Flow Cell to Troubleshoot Baseline Noise
  • LC Articles
  • LC Videos
  • LC Forum
  • More
  • Cancel

Browse the Agilent Knowledge Portal

  • Chromatography
    • Gas Chromatography
    • Liquid Chromatography
  • More
  • Cancel
  • GC/MS
  • More
  • Cancel
  • Spectroscopy
    • Atomic Absorption
    • ICP-OES
  • More
  • Cancel
  • Automated Electrophoresis
  • More
  • Cancel
  • Software and Informatics
    • OpenLab CDS
    • MassHunter
  • More
  • Cancel
  • -Liquid Chromatography
    • +LC Operation
    • +LC Maintenance
    • -LC Troubleshooting
      • Troubleshooting High Pressure Issues in Agilent LC Systems
      • Troubleshooting Peak Area Variation Between Injections
      • Forced Cold Start Procedure
      • Troubleshooting Remote/Cluster Partner Missing Error
      • How to Access Diagnostic Buffer File from LC Modules for Advanced Troubleshooting
      • How to Free up Licenses After License Check Message on Agilent Lab Advisor Software
      • "Could not open channel RD" error message in Lab Advisor
      • +LC Pumps Troubleshooting
      • +LC Injectors Troubleshooting
      • +LC Column Compartments Troubleshooting
      • -LC Detectors Troubleshooting
        • New UV Lamp Not Working - Lamp Tag Not Ready Error
        • Troubleshooting a failed ASTM Drift and Noise Test for the Agilent RID
        • How to Flush an HPLC Flow Cell to Troubleshoot Baseline Noise
        • Removing a Clog from a Flow Cell
    • +LC Best Practices
    • +LC Firmware

How to Flush an HPLC Flow Cell to Troubleshoot Baseline Noise

Created by Paola Labarbuta Paola Labarbuta over 1 year ago | Last modified by Paola Labarbuta Paola Labarbuta 5 months ago

This Information Applies To: Agilent UV LC detectors: Diode Array Detectors (DAD), Multiple Wavelength Detectors (MWD), Variable Wavelength Detectors (VWD)


Issue

The following procedure is recommended to flush a dirty flow cell and remove baseline noise.


  

Caution: disconnect the column before following the procedure. Place a Union instead of the column.

Requirements

  • Reverse the cell flow path by swapping the inlet and outlet lines. 
  • If you suspect a clog or a partial obstruction of the cell, normally associated to an increase in pressure, follow the procedure described at this link first.

Steps to follow

Reversed phase applications

  1. Open the Purge Valve of your pump. Set a flow of 5 ml/min and purge each channel for 5 min with HPLC-quality Water. During step 1, check that the solvent moves into the solvent line correctly and it goes out the Purge Valve. If it does not, check the state of the Solvent Filters and replace them.
  2. Close the Purge Valve and flush the system. Set the flow rate at 1 ml/min. Let it run for 1 hour, with HPLC-quality Water. Make sure that the pressure does not exceed 60 bars during this step.
  3. Flush the system with 100% Isopropanol (IPA) 1 ml/min, for at least 1 hour. Make sure that the pressure does not exceed 60 bars during this step.
  4. Repeat step 2 then continue with step 5.
  5. Swap the inlet and outlet lines of the cell flow and equilibrate the system with your solvents.
  6. Check if the baseline issue observed initially disappeared. If so, reconnect the column and proceed in using the system. If the issue persists, please contact Agilent Technical Support.

Normal phase applications

  1. Open the Purge Valve of your pump and purge each solvent line, one by one, with HPLC-quality Isopropanol (IPA). Set a flow of 5 mL/min and purge for 5 min. During step 1, check that the solvent moves into the solvent line correctly and it goes out the Purge Valve. If it does not, check the state of the Solvent Filters and replace them if needed.
  2. Close the Purge Valve and flush the system with IPA and a flow rate of 1 ml/min. Let it run for at least 2 hours, with HPLC-quality IPA. Make sure that the pressure does not exceed 60 bars during this step.
  3. Swap the inlet and outlet lines of the flow cell. 
  4. Equilibrate the system with your solvents. 
  5. Check if the baseline issue observed initially disappeared. If so, reconnect the column and proceed in using the system. If the issue persists, please contact Agilent Technical Support.
 Tip 
Learn how to effectively troubleshoot your Agilent UV Detector:
HPLC-0GEN-2201e - Chromatographic Troubleshooting for HPLC e-learning course available from Agilent education

  • variable wavelength detector
  • 1220
  • Detectors
  • diode array detector
  • 1290 infinity
  • 1120 and 1220 Series
  • Multiple Wavelength Detector
  • 1260 Infinity
  • 1260 Infinity II
  • 1200 Series
  • 1120
  • Analytical LC
  • 1290 Infinity II
  • Share
  • History
  • More
  • Cancel
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel

Contact Agilent Support

 Still Need Assistance?

 Contact Agilent Support >

Related
Recommended for you
Privacy Statement
Terms of Use
Contact Us
Site Help