Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
  • Home
  • Technical Areas
    Technical Areas
    • AssayMap
    • Atomic Spectroscopy
    • Automated Electrophoresis
    • Cell Analysis
    • Consumables
    • CrossLab Connect
      CrossLab Connect
      • Service Management
      • Smart Alerts
    • Dissolution
    • GC
    • GC/MS
    • iLab Operations Software
    • LC
    • LC/MS
    • Life Sciences Reagents
    • Molecular Spectroscopy
      Molecular Spectroscopy
      • Molecular Spectroscopy - UV-Vis & UV-Vis-NIR, FTIR, Raman & Fluorescence
      • 8700 LDIR User Group
      • Transmission Raman User Group
    • Real-Time PCR (qPCR)
    • Sample Preparation
    • Software
      Software
      • Chromatography Software
        OpenLab CDS, OpenLab ChemStation, OpenLab EZChrom
      • Data Management Software
        OpenLab ECM, OpenLab ECM XT, OpenLab Server
      • Mass Spectrometry Software
        Data Analysis, Data Acquisition, Productivity App, WalkUp Software
    • Vacuum
  • Applications Areas
    Applications Areas
    • Basic Research
    • Energy and Chemicals
    • Environmental
    • Food and Agriculture
  • Agilent Knowledge Portal
  • Links
    Links
    • Collection of Support Resources
    • Webinar Notifications
    • Contact Support
    • User Contributions
  • Getting Started
    Community Help and Resources
  • Browse
    Browse
    • People
    • Places
    • Your View
    • Your Community Inbox
LC Portal
  • Agilent Knowledge Portal
LC Portal
Articles How to Change the Outlet Valve on a 1260 Isocratic or Quaternary Pump
  • Articles
  • Videos
  • Forum
  • Announcements
  • More
  • Cancel
  • Agilent Knowledge Portal
  • LC Portal Home
  • LC Portal Articles
  • LC Portal Videos
  • More
  • Cancel
  • Created :  10 Nov 2022
  • Modified :  6 Feb 2023
  • Category :  Agilent Knowledge Portal
  • Entry Type :  Article
  • Product Type :  Analytical LC Systems
  • Component :  LC Pumps
  • Product Name :  1260 Infinity II LC System 1260 Infinity LC System 1260 Infinity II Prime LC System
  • Task :  Maintenance
Related Articles
  • How to Replace an Inlet Valve in a G7120A Agilent 1290 Infinity II High Speed Pump

    This Information Applies To: All Agilent G7120A 1290 Infinity II High Speed Pump modules. Issue: Required when: The inlet valve is defective. A pressure or leak rate test in Lab Advisor suggests the inlet valve may require replacement. If a large increase in pressure ripple is observed. Steps to follow Tools Required: Wrench, 14 mm 5067–5688 - Torque wrench 1 to 25 Nm with 14 mm wrench Parts Required: G4220-60022 - Inlet valve (primary pump head) ...
  • How to Change the Passive Inlet Valve (PIV) on a 1260 Isocratic or Quaternary Pump

    This Information Applies To: Agilent 1260 Isocratic (G1310B) Quaternary (G1311B) Pumps. Issue Required when: The inlet valve is defective. The pressure or leak test in Lab Advisor suggests the inlet valve may require replacement. A large increase in pressure ripple is observed. Steps to follow Tools Required: 8710–1924 - Open-end wrench 14 mm 5067–5688 - Torque wrench, 1 - 25 Nm with 14 mm wrench Pair of tweezers or small flat blade screwdriver Parts Required: ...
  • How to Troubleshoot Low Pressure or No Flow/Solvent Coming Out of 1100/1200 Series Isocratic/Quaternary/Binary Pumps

    This Information Applies To: Agilent 1100/1200 Series G1310A Isocratic Pump, G1311A Quaternary Pump, and G1312A/B Binary Pump modules Issue This article shows different options to troubleshoot low pressure or no flow issues in 1100/1200 series Isocratic/Quaternary/Binary pumps. Items Required Open-end wrench 14 mm (p/n 8710–1924 ) Open-end wrench 1/4 - 5/16 inch (p/n 8710–0510 ) Pair of Tweezers Steps to follow The following steps will guide you to find out which part is causing...
  • How to Change the Needle and Needle Seat in an Agilent G7129x Vialsampler

    This Information Applies To: Agilent G7129A/B/C Vialsampler Issue: Replacing the needle and needle seat of the Vialsampler is required when: The needle is visibly damaged. Leaks or blockages in the vialsampler are observed, especially between the needle and needle seat. The limit for the needle interaction EMF counter is exceeded. The needle needs to be replaced as part of yearly maintenance. Over time, deposits can block the needle and the needle seat causing pressure issues...
  • Nebulizer Needle Replacement in an Agilent Jet Stream or Standard Electrospray Ion Source

    This Information Applies To: Agilent 6000 Series LC/MS systems. Issue Efficient electrospray ionization (ESI) in your LC/MS ion source requires a properly adjusted nebulizer needle in good condition. The nebulizer needle is a consumable item in your ion source that may need to be replaced when the following signs are observed: Increased LC backpressure associated to the MSD. Off-axis spraying Dripping from the Nebulizer Poor or inconsistent spray shape This article describes how...

How to Change the Outlet Valve on a 1260 Isocratic or Quaternary Pump

Answer

This Information Applies To: Agilent 1260 Isocratic (G1310B) quaternary (G1311B) pump modules 


Issue:

Required when:

  • The outlet valve is defective or leaking.
  • The pressure or leak test in Lab Advisor suggests the outlet valve may require replacement.
  • There is large increase in pressure ripple is observed.

Steps to follow

Tools Required:

  • 8710–1924 - Wrench, open 14 mm
  • 8710–0510 - Wrench, open 1/4 - 5/16 inch
  • 5067–5688 - Torque wrench, 1 - 25 Nm with 14 mm wrench

 Parts Required:

  • G1312-60067 - Valve outlet, 600 Bar 1220/1260
 Warning 
Warning: This procedure risks exposure to hazardous, toxic, or flammable solvents and reagents. Wear appropriate PPE as described in the material handling and safety data sheet supplied by the chemical vendor, and always follow good laboratory practice.

1. Switch off the pump module and remove the pump front cover.

2. If possible, raise the solvent filters in each solvent bottle connect to the pump above the level of solvent in the bottle. This will minimize solvent leaking from the solvent tubes while performing this maintenance.

3. Using a 1/4-inch wrench, disconnect the valve capillary from the outlet valve. Using the 14 mm wrench, loosen the valve and remove it from the pump body (Figure 1).

Figure 1. Removing the valve capillary (1) and outlet valve (2) from the pump head 

 

4. Do not disassemble the outlet valve, to prevent damaging the valve.

5. Install the new outlet valve and tighten the valve using a torque wrench (12 Nm). Reconnect the valve capillary (Figure 2).

 

Figure 2. Installing the Outlet Valve on the pump head

6. Reinstall the front cover and then power on the pump module.

 If possible, run a system pressure test with the outlet of the purge valve blanked off using the appropriate blanking nut (Part No. 01080–83202). This checks that the pump flow path is pressure tight to a pressure higher than your application requires.

 Cause: If old or aged Acetonitrile is left in the outlet ball valve, then the ball inside the valve can stick open or closed causing pressure and flow issues. Always flush pumps after use according to instructions given in the pump user manual.

 Tip 

Learn more on how to effectively troubleshoot your 1260 Agilent HPLC with ChemStation:

Agilent 1260 Infinity LC Systems with OpenLab CDS ChemStation Edition Learning Path available from Agilent Education

 
 