This Information Applies To: Agilent GC/MSD Systems using the Agilent IDP-3 dry scroll pump
Issue
Scroll pump requires replacement two tip seals and one O-ring after approximately one year of use (9000 hours).
Requirements
Items Required
Figure 1. IDP-3 tip seal maintenance kit
Part Label | Part Name | Quantity |
A | 4 mm T-handle Allen wrench | 1 |
B | Low lint cleaning wipes | 1 pkg |
C | Abrasive pad 9” by 6” | 1 |
D | Low Lint Cleaning Swabs | 1 pkg |
E | Nitrile gloves (color may vary) | One pair |
F | O-ring | 1 |
G | Tip seal 1 set | 2 |
H | Silencer Filter | 1 |
I | Dust Mask | 1 |
Table 1. IDP-3 tip seal maintenance kit contents details
Steps to Follow
Tips: For additional information, see the full video: Agilent IDP-3 Dry Scroll Pump for the Agilent 5977, 5975, and 5973 Series GC/MSD - Pump Maintenance Video
|