This Information Applies To: Agilent ICP-OES 5100, 5110, 5800, 5900
Issue:
Over time the water inlet filter can become clogged and reduce water flow. Reduced water flow limits the system cooling and can result in the instrument shutting down during analysis.
Recommended maintenance schedule: Agilent recommends cleaning the water inlet filter every three months.
|
ICP-OES water inlet filter overview:
Figure 1: ICP-OES water inlet filter components
1 Water outlet connected to water chiller, 2 Water inlet connection from water chiller, 3 Water filter housing
Tool required:
Steps to follow:
|
Caution: Always turn off your instrument before attempting to replace a filter and do not operate the instrument while the filter is removed.
|
Turn off the ICP-OES instrument
Turn off the water chiller
Gently pull the back of the right side panel towards you and then slide it off the instrument
Place a beaker or paper towel under the water filter
Unscrew the water filter drain plug with a 17 mm open ended wrench
Remove the drain plug
Remove the filter from the housing
Soak the filter in water or a mild detergent and then rinse thoroughly with tap water
Insert the filter into the drain plug
Insert the water drain plug into the housing and then tighten with a wrench
Turn on the chiller
Turn on the instrument.
|