Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
  • Home
  • Technical Areas
    Technical Areas
    • AssayMap
    • Atomic Spectroscopy
    • Automated Electrophoresis
    • Cell Analysis
    • Consumables
    • CrossLab Connect
      CrossLab Connect
      • Service Management
      • Smart Alerts
    • Dissolution
    • GC
    • GC/MS
    • LC
    • LC/MS
    • Life Sciences Reagents
    • Molecular Spectroscopy
      Molecular Spectroscopy
      • Molecular Spectroscopy - UV-Vis & UV-Vis-NIR, FTIR, Raman & Fluorescence
      • 8700 LDIR User Group
      • Transmission Raman User Group
    • Real-Time PCR (qPCR)
    • Sample Preparation
    • Software
      Software
      • Chromatography Software
        OpenLab CDS, OpenLab ChemStation, OpenLab EZChrom
      • Data Management Software
        OpenLab ECM, OpenLab ECM XT, OpenLab Server
      • Mass Spectrometry Software
        Data Analysis, Data Acquisition, Productivity App, WalkUp Software
    • Vacuum
  • Applications Areas
    Applications Areas
    • Basic Research
    • Energy and Chemicals
    • Environmental
    • Food and Agriculture
  • Agilent Knowledge Portal
  • Links
    Links
    • Collection of Support Resources
    • Webinar Notifications
    • Contact Support
    • User Contributions
  • Getting Started
    Community Help and Resources
  • Browse
    Browse
    • People
    • Places
    • Your View
    • Your Community Inbox
ICP-OES Portal
  • Agilent Knowledge Portal
ICP-OES Portal
Articles AVS 6/7 Tubing Connections and Part Numbers
  • Articles
  • Videos
  • Forum
  • Announcements
  • More
  • Cancel
  • Agilent Knowledge Portal
  • ICP-OES Portal Home
  • ICP-OES Portal Articles
  • ICP-OES Portal Videos
  • More
  • Cancel
  • Created :  10 Nov 2022
  • Modified :  8 Feb 2023
  • Category :  Agilent Knowledge Portal
  • Entry Type :  Article
  • Product Type :  Atomic Spectroscopy Accessories ICP OES Systems
  • Component :  Atomic Spectroscopy Sample Introduction Accessories
  • Product Name :  5100 ICP-OES 5110 ICP-OES 5800 ICP-OES 5900 ICP-OES AVS 6/7
  • Task :  Maintenance Operation
Related Articles
  • Bypassing the AVS for the 5000 Series ICP-OES

    This Information Applies To: 5000 series ICP-OES (5100,5110,5800,5900), AVS 4/6/7 Issue Steps to bypass the AVS to troubleshoot poor performance on the ICP-OES Tip : These steps should be completed after the Subsystem Communication, Air Flow, Water Flow, Gas Flows, RF Generator, Camera, and Optics Tests have been completed with a status of ‘Done’ or ‘Pass ’ Bypass (or skip) the AVS to run a worksheet and troubleshoot instrument performance issues such as sensitivity...
  • ICP-OES Detector and Instrument Calibration

    This Information Applies To: Agilent 5000 Series ICP-OES (5100, 5110, 5800, 5900) Issue How to calibrate the ICP-OES ICP-OES calibration is a two-step procedure: The detector calibration performs a dark current calibration to correct for detector and electronic background signals. The instrument calibration uses a specified calibration solution to calibrate the wavelengths of the polychromator. ICP-OES wavelength calibration solution is available as either Agilent ICP-OES wavelength...
  • Replacing the AVS 6/7 Rotor

    This Information Applies To: Agilent 5000 Series ICP-OES Systems, AVS 6/7 Issue Over time the AVS 6/7 rotor can become worn or damaged by chemicals. Agilent recommends inspecting and/or replacing the rotor every 5000 AVS switches or when analytical results or precision fall outside expected limits. Steps to Follow: Tip : Refer to the Help section of the Agilent ICP Expert Instrument Software for further detailed instructions. Warning : Always wear appropriate...
  • AVS 6/7 Maintenance Checklist

    This Information Applies To: Agilent 5000 Series ICP-OES, AVS 6/7 Issue The AVS 6 and AVS 7 require routine maintenance. This article provides a maintenance checklist. Figure 1. 1 AVS 7 Warning : Always rinse all tubing and the valve before disconnecting the tubing from the switching valve to avoid contact with potentially dangerous chemicals. Appropriate personal protective equipment (PPE) should be worn when handling components contaminated with chemicals. ...
  • Set up an ICP-OES Worksheet with Internal Standard while Bypassing the AVS

    This Information Applies To: 5000 series ICP-OES (5100,5110,5800,5900), AVS 4/6/7, ICP Expert version 7.X Issue The AVS can contribute to instrument performance issues such as sensitivity and precision. During troubleshooting, bypassing (or skipping) the AVS and running a worksheet during can identify if the AVS is the root cause of the issue. Other root causes could be the method application, or an instrument fault. Items Required If an internal standard line is required, install the...

AVS 6/7 Tubing Connections and Part Numbers

Answer

This Information Applies To: Agilent 5000 Series ICP-OES, AVS 6/7


Issue

This article shows the required tubing connections and part numbers between the AVS 6/7 and ICP-OES


ICP-OES sample introduction system:

 Figure 1. Sample introduction system and AVS 7 (shown as 1)


Figure 2. 
Connections to AVS:  1 to AVS Pump, 2 from Autosampler probe (rinse), 3 Argon gas, 4 Rinse/Carrier

Figure 3. AVS (showing tubing removed): (A)  AVS valve, (B) Sample loop holder, (C) AVS pump inlet, (D) AVS pump outlet, (E) Bubble injector

Table 1 shows connections points to AVS valve and part numbers of tubing

Valve port number
(from position in A in Figure 3)

Tubing kit part number

Description

1

G8010-60305

Sample Loop, 1 mL

2

G8010-60304

Nebulizer Connection Kit

3

G8010-60303

Sample Inlet Connection Kit

4

G8010-60305

Sample Loop, 1 mL

5

G8010-60306

Autosampler Connection Kit

6

G8010-60308

Pump Outlet Connection Kit

7

G8010-60307

Internal Standard Connection Kit for AVS 7 only

Table 1. AVS connections

Figure 4. AVS (showing tubing connected): (A) to Nebulizer, (B) to Autosampler, (C) to Internal Standard (AVS 7 only), (D) to AVS pump inlet, (E) to Rinse/Carrier

 Tip  Tip: Refer to the Help section of the Agilent ICP Expert Instrument Software for full details of tubing connection points and parts required

After replacing all required components, perform the Advanced Valve System Test as detailed in the Help section of the Agilent ICP Expert Instrument Software to check performance of the AVS.

 

 Tip  Learn how to effectively maintain your Agilent AVS:
 Agilent 5000 Series ICP-OES with ICP Expert e-learning path available from Agilent education
 
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel
Privacy Statement
Terms of Use
Contact Us
Site Help