This Information Applies To: Agilent 5000 Series ICP-OES, AVS 6/7
Issue
This article shows the required tubing connections and part numbers between the AVS 6/7 and ICP-OES
ICP-OES sample introduction system:
Figure 1. Sample introduction system and AVS 7 (shown as 1)
Figure 2. Connections to AVS: 1 to AVS Pump, 2 from Autosampler probe (rinse), 3 Argon gas, 4 Rinse/Carrier
Figure 3. AVS (showing tubing removed): (A) AVS valve, (B) Sample loop holder, (C) AVS pump inlet, (D) AVS pump outlet, (E) Bubble injector
Table 1 shows connections points to AVS valve and part numbers of tubing
Valve port number |
Tubing kit part number |
Description |
1 |
Sample Loop, 1 mL |
|
2 |
Nebulizer Connection Kit |
|
3 |
Sample Inlet Connection Kit |
|
4 |
Sample Loop, 1 mL |
|
5 |
Autosampler Connection Kit |
|
6 |
Pump Outlet Connection Kit |
|
7 |
Internal Standard Connection Kit for AVS 7 only |
Table 1. AVS connections
Figure 4. AVS (showing tubing connected): (A) to Nebulizer, (B) to Autosampler, (C) to Internal Standard (AVS 7 only), (D) to AVS pump inlet, (E) to Rinse/Carrier
|
Tip: Refer to the Help section of the Agilent ICP Expert Instrument Software for full details of tubing connection points and parts required |
After replacing all required components, perform the Advanced Valve System Test as detailed in the Help section of the Agilent ICP Expert Instrument Software to check performance of the AVS.
|
Learn how to effectively maintain your Agilent AVS: Agilent 5000 Series ICP-OES with ICP Expert e-learning path available from Agilent education |