This Information Applies To: Agilent 5000 Series ICP-OES, AVS 6/7
Issue
This article shows the required tubing connections and part numbers between the AVS 6/7 and ICP-OES
ICP-OES sample introduction system:
Figure 1. Sample introduction system and AVS 7 (shown as 1)
Figure 2. Connections to AVS: 1 to AVS Pump, 2 from Autosampler probe (rinse), 3 Argon gas, 4 Rinse/Carrier
Figure 3. AVS (showing tubing removed): (A) AVS valve, (B) Sample loop holder, (C) AVS pump inlet, (D) AVS pump outlet, (E) Bubble injector
Table 1 shows connections points to AVS valve and part numbers of tubing
Valve port number |
Tubing kit part number |
Description |
1 |
Sample Loop, 1 mL |
|
2 |
Nebulizer Connection Kit |
|
3 |
Sample Inlet Connection Kit |
|
4 |
Sample Loop, 1 mL |
|
5 |
Autosampler Connection Kit |
|
6 |
Pump Outlet Connection Kit |
|
7 |
Internal Standard Connection Kit for AVS 7 only |
Table 1. AVS connections
Figure 4. AVS (showing tubing connected): (A) to Nebulizer, (B) to Autosampler, (C) to Internal Standard (AVS 7 only), (D) to AVS pump inlet, (E) to Rinse/Carrier
Tip: Refer to the Help section of the Agilent ICP Expert Instrument Software for full details of tubing connection points and parts required |
After replacing all required components, perform the Advanced Valve System Test as detailed in the Help section of the Agilent ICP Expert Instrument Software to check performance of the AVS.
Learn how to effectively maintain your Agilent AVS: Agilent 5000 Series ICP-OES with ICP Expert e-learning path available from Agilent education |