Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
  • Home
  • Technical Areas
    Technical Areas
    • AssayMap
    • Atomic Spectroscopy
    • Automated Electrophoresis
    • Bioreagents
    • Cell Analysis
    • Consumables
    • CrossLab Connect
      CrossLab Connect
      • Service Management
      • Smart Alerts
    • Dissolution
    • GC
    • GC/MS
    • LC/MS
    • LC
    • Molecular Spectroscopy
      Molecular Spectroscopy
      • Molecular Spectroscopy - UV-Vis & UV-Vis-NIR, FTIR, Raman & Fluorescence
      • 8700 LDIR User Group
    • Sample Preparation
    • Software
      Software
      • Chromatography Software
        OpenLab CDS, OpenLab ChemStation, OpenLab EZChrom
      • Data Management Software
        OpenLab ECM, OpenLab ECM XT, OpenLab Server
      • Mass Spectrometry Software
        Data Analysis, Data Acquisition, Productivity App, WalkUp Software
    • Vacuum
  • Applications Areas
    Applications Areas
    • Basic Research
    • Energy and Chemicals
    • Environmental
    • Food and Agriculture
  • Agilent Knowledge Portal
  • Links
    Links
    • Collection of Support Resources
    • Webinar Notifications
    • Contact Support
    • User Contributions
  • Getting Started
    Community Help and Resources
  • Browse
    Browse
    • People
    • Places
    • Your View
    • Your Community Inbox
ICP-OES Portal
  • Agilent Knowledge Portal
ICP-OES Portal
Articles Replacing the AVS 6/7 Rotor
  • Articles
  • Videos
  • Forum
  • Announcements
  • More
  • Cancel
  • Agilent Knowledge Portal
  • ICP-OES Portal Home
  • ICP-OES Portal Articles
  • ICP-OES Portal Videos
  • More
  • Cancel
  • Created :  10 Nov 2022
  • Modified :  8 Feb 2023
  • Category :  Agilent Knowledge Portal
  • Entry Type :  Article
  • Product Type :  Atomic Spectroscopy Accessories ICP OES Systems
  • Component :  Atomic Spectroscopy Sample Introduction Accessories
  • Product Name :  AVS 6/7
  • Task :  Maintenance
Related Articles
  • Bypassing the AVS for the 5000 Series ICP-OES

    This Information Applies To: 5000 series ICP-OES (5100,5110,5800,5900), AVS 4/6/7 Issue Steps to bypass the AVS to troubleshoot poor performance on the ICP-OES Tip : These steps should be completed after the Subsystem Communication, Air Flow, Water Flow, Gas Flows, RF Generator, Camera, and Optics Tests have been completed with a status of ‘Done’ or ‘Pass ’ Bypass (or skip) the AVS to run a worksheet and troubleshoot instrument performance issues such as sensitivity...
  • AVS 6/7 Tubing Connections and Part Numbers

    This Information Applies To: Agilent 5000 Series ICP-OES, AVS 6/7 Issue This article shows the required t ubing connections and part numbers between the AVS 6/7 and ICP-OES ICP-OES sample introduction system: Figure 1. Sample introduction system and AVS 7 (shown as 1) Figure 2. Connections to AVS: 1 to AVS Pump, 2 from Autosampler probe (rinse), 3 Argon gas, 4 Rinse/Carrier Figure 3. AVS (showing tubing removed): (A) AVS valve, (B) Sample loop holder, (C) AVS pump inlet, (D) AVS...
  • Set up an ICP-OES Worksheet with Internal Standard while Bypassing the AVS

    This Information Applies To: 5000 series ICP-OES (5100,5110,5800,5900), AVS 4/6/7, ICP Expert version 7.X Issue The AVS can contribute to instrument performance issues such as sensitivity and precision. During troubleshooting, bypassing (or skipping) the AVS and running a worksheet during can identify if the AVS is the root cause of the issue. Other root causes could be the method application, or an instrument fault. Items Required If an internal standard line is required, install the...
  • AVS 6/7 Maintenance Checklist

    This Information Applies To: Agilent 5000 Series ICP-OES, AVS 6/7 Issue The AVS 6 and AVS 7 require routine maintenance. This article provides a maintenance checklist. Figure 1. 1 AVS 7 Warning : Always rinse all tubing and the valve before disconnecting the tubing from the switching valve to avoid contact with potentially dangerous chemicals. Appropriate personal protective equipment (PPE) should be worn when handling components contaminated with chemicals. ...
  • How to Reset the AVS 6/7 on Agilent ICP-OES Instruments

    This Information Applies To: 5000 series ICP-OES (5100/5110/5800/5900), ICP Expert Issue How to Reset the AVS 6/7 after maintenance or when a red LED is visible Steps to follow Tips : Instructions for changing the AVS switching valve can be found in the ICP Expert Help under “Hardware How to” and “Replace the AVS 4/6/7 Tubing”. Figure 1. ICP Expert Help After replacing the AVS valve and tubing, a red LED is showing on the AVS plate. Figure 2 shows the location...

Replacing the AVS 6/7 Rotor

Answer

This Information Applies To: Agilent 5000 Series ICP-OES Systems, AVS 6/7


Issue 

Over time the AVS 6/7 rotor can become worn or damaged by chemicals. Agilent recommends inspecting and/or replacing the rotor every 5000 AVS switches or when analytical results or precision fall outside expected limits.


 

Steps to Follow:

 Tip 
Tip: Refer to the Help section of the Agilent ICP Expert Instrument Software for further detailed instructions.

 

 Warning  Warning:
Always wear appropriate PPE when handling AVS 6/7 components and tubing to avoid contact with potentially dangerous chemicals. 
  1. It is not necessary to disconnect the tubing from the AVS valve to replace the rotor however, the tubing can be removed and rinsed when there is suspected system contamination. Figure 1 shows the AVS with the tubing removed. 


    Figure 1. 1 AVS7 with the tubing removed

  2. Disassemble the AVS 6/7. Use the 7/16 inch hex key provided with the valve to remove the AVS body. 

  3. Inspect the contact surface of the rotor for signs of wear and replace if there are visible score marks or damage. AVS 6/7 rotor part number G8494-60002, AVS 6 stator and rotor part number G8494-60003. AVS 7 stator and rotor part number G8495-60003

     

     

  4. Inspect the reverse face of the stator for signs of wear and replace if there are visible score marks or damage.

     

  5. Re-assemble the AVS7 valve.

     Tip  Tip: When reassembling the AVS, it is critical that each screw is tightened evenly in turn to ensure consistent pressure. If the pressure on the three screws is not even or the screws are too tight / too loose, air leaks may occur.

  6. When finished reassembling the AVS, it is recommended to run the Advanced Valve System Test as detailed in the Help section of the Agilent ICP Expert Instrument Software to check performance of the AVS. 

 Tip 
Learn how to effectively maintain your Agilent AVS:
Agilent 5000 Series ICP-OES with ICP Expert e-learning path available from Agilent education
 
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel
Contact Agilent Support

Still Need Assistance?

Contact Agilent Support >

Privacy Statement
Terms of Use
Contact Us
Site Help