This Information Applies to: Agilent 5200/5300/5400 Fragment Analyzer, Femto Pulse, and ZAG DNA Analyzer
Issue
The kit components that are stored from 2 to 8 °C were not chilled during transportation from Agilent to the user.
Resolution
The chilled reagents always arrive without being chilled...
This Information Applies to: Agilent 2100 Bioanalyzer system, RNA 6000 Nano kit, RNA 6000 Pico kit, and Small RNA kit
Issue
The Bioanalyzer RNA Ladder and/or samples are degraded.
Background
Two examples of degraded RNA Ladders are shown in Figure 1 .
Figure 1 . Electropherogram of a partially...
This Information Applies to: Agilent 5200/5300/5400 Fragment Analyzer, Femto Pulse, and ZAG DNA Analyzer systems
Issue
One or more electropherograms have baseline noise.
Background
An example of baseline noise is shown in Figure 1 .
Figure 1 . Unexpected peaks and baseline noise in a...
This Information Applies To: Agilent 5200/5300/5400 Fragment Analyzer, Femto Pulse, Oligo Pro II, ZAG DNA Analyzer and ProteoAnalyzer systems
Issue
The Agilent parallel capillary electrophoresis assay quick guides contain the guidance to allow reagents to equilibrate at room temperature before...
This Information Applies To: Agilent 5200/5300/5400 Fragment Analyzer, Femto Pulse, and ProteoAnalyzer systems
Issue
If the bottles in the Fragment Analyzer, Femto Pulse, or ProteoAnalyzer side compartment are damaged or lost, they would need to be replaced.
Background
An example of the bottles...