Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
  • Home
  • Technical Areas
    Technical Areas
    • AssayMap
    • Atomic Spectroscopy
    • Automated Electrophoresis
    • Bioreagents
    • Cell Analysis
    • Consumables
    • CrossLab Connect
      CrossLab Connect
      • Service Management
      • Smart Alerts
    • Dissolution
    • GC
    • GC/MS
    • LC/MS
    • LC
    • Molecular Spectroscopy
      Molecular Spectroscopy
      • Molecular Spectroscopy - UV-Vis & UV-Vis-NIR, FTIR, Raman & Fluorescence
      • 8700 LDIR User Group
    • Sample Preparation
    • Software
      Software
      • Chromatography Software
        OpenLab CDS, OpenLab ChemStation, OpenLab EZChrom
      • Data Management Software
        OpenLab ECM, OpenLab ECM XT, OpenLab Server
      • Mass Spectrometry Software
        Data Analysis, Data Acquisition, Productivity App, WalkUp Software
    • Vacuum
  • Applications Areas
    Applications Areas
    • Basic Research
    • Energy and Chemicals
    • Environmental
    • Food and Agriculture
  • Agilent Knowledge Portal
  • Links
    Links
    • Collection of Support Resources
    • Webinar Notifications
    • Contact Support
    • User Contributions
  • Getting Started
    Community Help and Resources
  • Browse
    Browse
    • People
    • Places
    • Your View
    • Your Community Inbox
Automated Electrophoresis Portal
  • Agilent Knowledge Portal
Automated Electrophoresis Portal
Articles How to do a NaOH Flush on an Agilent Fragment Analyzer
  • Articles
  • Files
  • Forum
  • Announcements
  • More
  • Cancel
  • Agilent Knowledge Portal
  • Automated Electrophoresis Portal Home
  • Automated Electrophoresis Portal Articles
  • Automated Electrophoresis Portal Files
  • More
  • Cancel
  • Created :  10 Nov 2022
  • Modified :  7 Feb 2023
  • Category :  Agilent Knowledge Portal
  • Entry Type :  Article
  • Product Type :  Automated Electrophoresis Systems
  • Component :  Fragment Analyzer Systems
  • Product Name :  5200 Fragment Analyzer System 5300 Fragment Analyzer System 5400 Fragment Analyzer System
  • Task :  Maintenance
Related Articles
  • No Data From One or More Capillary on Parallel Capillary Electrophoresis Instruments

    This Information Applies To: 5200, 5300 and 5400 Fragment Analyzer, Femto Pulse and ZAG DNA analyzer Issue When no data is available in the Agilent ProSize data analysis software for one or more capillaries, the capillary(ies) could be misaligned, blocked, require extra conditioning, or the electrodes might be encrusted. Steps to Follow Check the capillary alignment Check the alignment of the capillaries in ProSize. Checking the capillary alignment is briefly described in this section...
  • Long-Term Capillary Array Storage on Parallel Capillary Electrophoresis Instruments

    This Information Applies To: Agilent 5200/5300/5400 Fragment Analyzer, Femto Pulse, ZAG DNA Analyzer, and Oligo Pro II Issue Methods for storing capillary array when not in use for longer than two weeks. The array can either be stored installed in the instrument or in the external array docking station. Items Required Capillary Storage Solution For use with the Femto Pulse and Oligo Pro II system (p/n GP-435-0100 ) For use with the Fragment Analyzer and ZAG DNA Analyzer systems...
  • Agilent Capillary Array Installation on Parallel Capillary Electrophoresis Instruments

    This Information Applies To: 5200/5300/5400 Fragment Analyzer, Femto Pulse, ZAG, Oligo Pro II Issue: This article explains the steps to take after installing a brand new array in the Agilent 5200/5300/5400 Fragment Analyzer, Femto Pulse, ZAG, and Oligo Pro II systems. Steps to follow Capillary array replacement steps are detailed in chapter 8 of the Fragment Analyzer user manual , Femto Pulse user manual , ZAG DNA analyzer user manual , and chapter 9 of the Oligo Pro II user manual...
  • Agilent Parallel Capillary Electrophoresis Compatible Plasticware

    This Information Applies To: 5200/5300/5400 Fragment Analyzer, Femto Pulse, Oligo Pro II, and ZAG Issue: Only plasticware from the approved consumables list should be used in the Agilent Parallel Capillary Electrophoresis systems. The use of plasticware with different dimensions could lead to decreased injection quality and consistency as well as damage to the tips of the instrument capillaries. Resolution: Compatible plates and tubes are listed in the following manuals: 5200, 5300...
  • How to Use Mineral Oil to Prevent Evaporation on Agilent Parallel Capillary Electrophoresis Systems

    This Information Applies To: 5200, 5300, 5400 Fragment Analyzer, Femto Pulse, Oligo Pro II, and ZAG Issue: Adding mineral oil to Agilent parallel capillary electrophoresis DNA marker plates and samples plates prevents evaporation and enables re-use of the plates. Caution : Agilent has not validated adding mineral oil to RNA plates. It is used at the customers own risk. Steps to follow: Tip : Agilent recommends using mineral oil part number FS-SMO15 to overlay...

How to do a NaOH Flush on an Agilent Fragment Analyzer

Answer

This Information Applies To: 5200/5300/5400 Fragment Analyzer, Fragment Analyzer controller software


Issue:

Over time the dynamic coating polymer on the capillaries needs to be refreshed. Failure to do so can lead to a loss of performance. Indicators that the dynamic coating needs refreshing include:

  • Delayed migration
  • Peak tailing

Also, if capillaries are stored dry or the Storage Solution is not replaced they can clog. Indicators of clogged capillaries include no flow through the capillaries leading to no signal. Flushing the capillaries with sodium hydroxide (NaOH) followed by Conditioning Solution rejuvenates the dynamic coating and unclogs capillaries restoring instrument performance.

 Tip  Recommended maintenance schedule:  Agilent recommends performing the NaOH flush as part of troubleshooting procedures and your routine maintenance to restore separation performance. You can add it as part of a regular cleaning schedule weekly or biweekly to ensure optimal performance.

 


Steps to follow:

 Tip  Tips: Further information on the NaOH flush procedure is available in the Fragment Analyzer System Manual Appendix. To remove dried gel or Storage Solution, consider performing Method B before the NaOH flush: submerse capillary ends and electrodes in hot water (65–93 °C). Sometimes, the reservoir vent valve needs to be cleaned as well (for more details, see here).

 

 Warning  Warning: Sodium hydroxide is corrosive and can cause severe burns when it contacts skin. Always wear appropriate personal protective equipment when handling sodium hydroxide. 

 

  1. Prepare 0.5 N NaOH solution. Minimum solution volume required is 15 mL for priming all three lines twice plus the volume required for ten-minute flush as follows:
    ≥12 mL for a 12-capillary array
    ≥25 mL for a 48-capillary array
    ≥50 mL for a 96-capillary array
  2. Place a NaOH solution sequentially in the Gel1, Conditioning Solution, and Gel2 positions and prime the tubing twice to clean out potential debris (click Utilities > Prime). Volume needed per prime step: 2.5 mL. You may leave the tube with NaOH in the Gel2 position. 
  3. Place a bottle or tube containing fresh 1x Conditioning Solution in the Conditioning position and prime the tubing.
  4. Place a bottle or tube containing freshly prepared gel-dye mix in the Gel1 position and prime the tubing.
  5. Perform a 0.5N NaOH flush:
    1. In the main screen Operation tab > Capillary Array - Conditioning click Add to queue 

    2. From the Method drop down menu, select Method C Flush - 0.5 N NaOH - 10min 200psi.mthdc (you can check the parameters by clicking Edit). 

    3. Select OK and run the method from the queue (see Figure 1)


      Figure 1. How to select Method C Flush

       

       
  6. If the Gel2 position is not needed otherwise, you may leave the NaOH there. It will not damage the tubing. 
     Caution 
    Caution: If NaOH is not washed out of the capillaries immediately after the NaOH flush, it can damage the capillary array.
  7. Perform at least a Full Conditioning (see Figure 2) or a separation method with freshly prepared reagents and 24 µL of blank solution (for quantitative kits), marker only, or samples. This step is crucial to flush out residual NaOH and restore the capillaries. 

 
Figure 2. Where to find the Full Conditioning method

 

 Tip  Learn how to effectively maintain your Agilent Fragment Analyzer:
Fragment Analyzer system manual

For Research Use Only. Not for use in diagnostic procedures.

 PR7000-8016

 
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel
Contact Agilent Support

Still Need Assistance?

Contact Agilent Support >

Privacy Statement
Terms of Use
Contact Us
Site Help