Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
  • Home
  • Technical Areas
    Technical Areas
    • AssayMap
    • Atomic Spectroscopy
    • Automated Electrophoresis
    • Cell Analysis
    • Consumables
    • CrossLab Connect
      CrossLab Connect
      • Service Management
      • Smart Alerts
    • Dissolution
    • GC
    • GC/MS
    • LC
    • LC/MS
    • Life Sciences Reagents
    • Molecular Spectroscopy
      Molecular Spectroscopy
      • Molecular Spectroscopy - UV-Vis & UV-Vis-NIR, FTIR, Raman & Fluorescence
      • 8700 LDIR User Group
      • Transmission Raman User Group
    • Real-Time PCR (qPCR)
    • Sample Preparation
    • Software
      Software
      • Chromatography Software
        OpenLab CDS, OpenLab ChemStation, OpenLab EZChrom
      • Data Management Software
        OpenLab ECM, OpenLab ECM XT, OpenLab Server
      • Mass Spectrometry Software
        Data Analysis, Data Acquisition, Productivity App, WalkUp Software
    • Vacuum
  • Applications Areas
    Applications Areas
    • Basic Research
    • Energy and Chemicals
    • Environmental
    • Food and Agriculture
  • Agilent Knowledge Portal
  • Links
    Links
    • Collection of Support Resources
    • Webinar Notifications
    • Contact Support
    • User Contributions
  • Getting Started
    Community Help and Resources
  • Browse
    Browse
    • People
    • Places
    • Your View
    • Your Community Inbox
Automated Electrophoresis Portal
  • Agilent Knowledge Portal
Automated Electrophoresis Portal
Articles Run Aborted and Target Z Position Error on Parallel Capillary Instruments
  • Articles
  • Files
  • Forum
  • Announcements
  • More
  • Cancel
  • Agilent Knowledge Portal
  • Automated Electrophoresis Portal Home
  • Automated Electrophoresis Portal Articles
  • Automated Electrophoresis Portal Files
  • More
  • Cancel
  • Created :  10 Nov 2022
  • Modified :  7 Feb 2023
  • Category :  Agilent Knowledge Portal
  • Entry Type :  Article
  • Product Type :  Automated Electrophoresis Systems
  • Component :  Femto Pulse Systems Fragment Analyzer Systems Oligo Pro II Systems ZAG DNA Analyzer Systems
  • Product Name :  5200 Fragment Analyzer System 5300 Fragment Analyzer System 5400 Fragment Analyzer System Femto Pulse System Oligo Pro II System ZAG DNA Analyzer system
  • Task :  Troubleshooting
Related Articles
  • Failed to Build Pressure Error on a Parallel Capillary Electrophoresis Instrument

    This Information Applies To: 5200/5300/5400 Fragment Analyzer, Femto Pulse, ZAG DNA Analyzer and Oligo Pro II instruments, and their respective controller software. Issue Failed to build pressure error comes up shortly after starting a separation. Resolution Failure to build pressure error occurs when the system is trying to build pressure when liquid is being pumped during conditioning or flushing of the instrument. Possible causes and suggestions to address the issue are: 1. Empty reagent...
  • No Data From One or More Capillary on Parallel Capillary Electrophoresis Instruments

    This Information Applies To: 5200, 5300 and 5400 Fragment Analyzer, Femto Pulse and ZAG DNA analyzer Issue When no data is available in the Agilent ProSize data analysis software for one or more capillaries, the capillary(ies) could be misaligned, blocked, require extra conditioning, or the electrodes might be encrusted. Steps to Follow Check the capillary alignment Check the alignment of the capillaries in ProSize. Checking the capillary alignment is briefly described in this section...
  • Agilent Capillary Array Installation on Parallel Capillary Electrophoresis Instruments

    This Information Applies To: 5200/5300/5400 Fragment Analyzer, Femto Pulse, ZAG, Oligo Pro II Issue: This article explains the steps to take after installing a brand new array in the Agilent 5200/5300/5400 Fragment Analyzer, Femto Pulse, ZAG, and Oligo Pro II systems. Steps to follow Capillary array replacement steps are detailed in chapter 8 of the Fragment Analyzer user manual , Femto Pulse user manual , ZAG DNA analyzer user manual , and chapter 9 of the Oligo Pro II user manual...
  • Insert Disk Error in the Agilent Fragment Analyzer Controller Software

    This Information Applies To: 5200/5300/5400 Fragment Analyzer controller software, Femto Pulse controller software, ZAG DNA Analyzer controller software, Oligo Pro II controller software Issue: The Agilent Fragment Analyzer, Femto Pulse, ZAG, or Oligo Pro II controller software does not open when the desktop icon is double-clicked. Instead, one of the following errors come up: "Insert the 'Fragment Analyzer' disk and click OK" "The feature you are trying to use is on a network resource...
  • Agilent Parallel Capillary Electrophoresis Compatible Plasticware

    This Information Applies To: 5200/5300/5400 Fragment Analyzer, Femto Pulse, Oligo Pro II, and ZAG Issue: Only plasticware from the approved consumables list should be used in the Agilent Parallel Capillary Electrophoresis systems. The use of plasticware with different dimensions could lead to decreased injection quality and consistency as well as damage to the tips of the instrument capillaries. Resolution: Compatible plates and tubes are listed in the following manuals: 5200, 5300...

Run Aborted and Target Z Position Error on Parallel Capillary Instruments

Answer

 

This Information Applies To: Agilent 5200/5300/5400 Fragment Analyzer, Agilent Femto Pulse, Agilent ZAG DNA Analyzer


Issue: The instrument encounters a run error, Run Aborted, Position fault: Did not reach target Z position (Figure 1). This error can occur if the drawer gets opened when the stage is still moving. The error occurs soon after starting the run. 

   

Figure 1. Run Error:CtrlTaskStage Error 


Resolution: 

System Reboot: 

Perform an entire system reboot as described in the Agilent Knowledge Portal article Full Power Cycle for Agilent Parallel Capillary Electrophoresis Instruments and as mentioned in the following steps: 

  1. Turn off the instrument.
  2. Turn off the computer.
  3. Unplug the two USB cables labeled CCD and HUB from the back of the computer.
  4. Wait for approximately 10 minutes before reconnecting the USB cables.
  5. Reconnect USB cables to different, nonadjacent ports and reconnect power cables to the instrument and PC. We recommend keeping at least one port empty (unused) between the two USB cables.
  6. Turn on the computer and wait for it to complete startup.
  7. Turn on the instrument.
  8. Launch the controller software.

If the issue persists, please perform a stage reset as described in the following procedure.

Reset Stage:

Please ensure the instrument is connected and that the stage is not holding any plate. This can be checked by opening each drawer and making sure all the white plate carriers are in the drawers. Then perform a stage reset from the controller software:

  1. From the instrument controller software, open the Utilities drop-down menu (Figure 2).
  2. Select the Hardware I/O option and locate the stage section of the newly created window (Figure 3).
  3. Click Reset stage and ensure that the holding number directly above the reset button goes to zero (Figure 3). Before you reset the stage, please ensure that the stage is not holding a plate, then click Yes to proceed (Figure 4).

     

Figure 2: Select Hardware I/O from the Utilities menu of the controller software. The Fragment Analyzer controller software is shown as an example in this figure.  

    

Figure 3: Reset stage to holding number value to zero.

      

Figure 4: Confirm that you wish to proceed with the stage reset after checking that the stage is not holding a tray.

 If the error persists: 

  1. Take a screenshot of the error message.
  2. Bypass the error(s) and collect the event and error reports from the instrument controller software one by one. To collect the reports, open the instrument controller software (Login as an Administrator), click Admin, select and collect the Event Report, then select the Error Report (Figure 5). 
  3. Please send the screenshots, event report and error report to your local Agilent support representative.

       

Figure 5: Steps to collect the event and error reports. The Fragment Analyzer controller software is shown as an example in this figure. 

 

 Caution 
Caution: Before you reset the stage, ensure that it is not holding a plate. If the stage is holding a plate, the instrument might be damaged and the stage will not get reset. 

 

 

 

 Tip 

Learn how to effectively maintain your Agilent parallel capillary electrophoresis systems:
5200, 5300 and 5400 Fragment Analyzer User Manual 
Femto Pulse User Manual 
ZAG DNA Analyzer User Manual 
Oligo Pro II User Manual
Prosize Data Analysis Software User Manual

For Research Use Only. Not for use in diagnostic procedures.