upgrade to MH 11.1

Hello,

i upgraded my MH QUANT 10.1 to 11.1 and since then the batch.script doesn't work for me, we are using a modified component.config, do you know where the error could be? This component.config worked with all 10 versions of masshunter (10,10.1 and 10.2)

Thank you.

Parents
  • This is written in log..> [ERROR] ComponentManager: Script processing error: Compiler executable csc.exe not found.

  • Hello  ,

    I'm not aware of this error specifically in regard to Quant or MassHunter, but generally this type of error points to either a corrupted or missing Microsoft .NET framework component. Make certain both the 3.5 and 4.x versions of .NET Framework are enabled in your Windows Features section or try repairing them. 

  • Hello,

    we have latest framework 4.8 and framework 3.5 SP1, it works fine with MH 10.1.

    <?xml version="1.0" encoding="utf-8"?>
    <components>
      <defaultnamespace>Agilent.MassSpectrometry.DataAnalysis.Quantitative</defaultnamespace>
      <references>
        <reference framework="true" assembly="System.Data.dll"/>
        <reference framework="true" assembly="System.Data.DataSetExtensions.dll"/>
        <reference framework="true" assembly="System.Xml.dll"/>
        <reference framework="true" assembly="System.Web.Services.dll"/>
        <reference framework="false" assembly="Quantitation.dll"/>
        <reference framework="false" assembly="QuantUtils.dll"/>
        <reference framework="false" assembly="ScriptableInterfaces.dll"/>
        <reference framework="true" assembly="System.Linq.dll"/>
        <reference framework="true" assembly="System.Core.dll"/>
        <reference framework="true" assembly="System.Xml.Linq.dll"/>
        <reference framework="true" assembly="System.Windows.Forms.dll"/>
        <reference framework="true" assembly="System.Threading.Tasks.dll"/>
        <reference framework="true" assembly="System.Drawing.dll"/>
        <reference framework="false" assembly="C:\Program Files\Agilent\MassHunter\Workstation\Quant\bin\Als.OData.dll"/>
        <reference framework="false" assembly="C:\Program Files\Agilent\MassHunter\Workstation\Quant\bin\Microsoft.OData.Client.dll"/>
        <reference framework="false" assembly="C:\Program Files\Agilent\MassHunter\Workstation\Quant\bin\Microsoft.OData.Core.dll"/>
        <reference framework="false" assembly="C:\Program Files\Agilent\MassHunter\Workstation\Quant\bin\Microsoft.OData.Edm.dll"/>
        <reference framework="false" assembly="C:\Program Files\Agilent\MassHunter\Workstation\Quant\bin\Microsoft.Spatial.dll"/>
        <reference framework="true" assembly="System.Net.dll"/>
        <reference framework="true" assembly="System.IO.dll"/>
      </references>
      <interface name="IScriptableSample">
        <implementation assembly="CompoundMath_Script.dll">
          <script name="CompoundMath_Script.txt">
            <references>
              <reference framework="false" assembly="BatchAnalysis.dll"/>
            </references>
          </script>
        </implementation>
      </interface>
      <interface name="ICustomExpressions">
        <implementation assembly="C:\MassHunter\Scripts\Quant\IScriptable\CustomExpressions_Script.dll">
          <script name="C:\MassHunter\Scripts\Quant\IScriptable\CustomExpressions_Script.cs">
            <references>
            </references>
          </script>
        </implementation>
      </interface>
      <!--<interface name="IScriptableBatch">
        <implementation assembly="C:\MassHunter\Scripts\Quant\IScriptable\Batch_Script.dll">
          <script name="C:\MassHunter\Scripts\Quant\IScriptable\Batch_Script.txt">
            <references>
              <reference framework="false" assembly="BatchAnalysis.dll"/>
            </references>
          </script>
        </implementation>
      </interface>-->
      <interface name="IScriptableBatch">
        <implementation assembly="C:\MassHunter\Scripts\Quant\IScriptable\Batch_Script.dll">
          <script name="C:\MassHunter\Scripts\Quant\IScriptable\Batch_Script.txt">
            <references>
              <reference framework="false" assembly="BatchAnalysis.dll"/>
              <reference framework="true" assembly="System.Data.dll"/>
              <reference framework="true" assembly="System.Linq.dll"/>
              <!--<reference framework="true" assembly="System.Linq.Enumerable.dll"/>-->
              <reference framework="true" assembly="System.Data.DataSetExtensions.dll"/>
              <reference framework="true" assembly="System.Collections.dll"/>
              <reference framework="true" assembly="System.Xml.Linq.dll"/>
              <reference framework="true" assembly="System.IO.dll"/>
              <reference framework="true" assembly="System.Xml.dll"/>
              <reference framework="true" assembly="System.Core.dll"/>
              <reference framework="true" assembly="System.Threading.Tasks.dll"/>
              <reference framework="true" assembly="System.Net.dll"/>
            </references>
          </script>
        </implementation>
      </interface>
      <interface name="IScriptableOutlier" instanceName="Accuracy">
        <implementation assembly="OutlierAccuracy_Script.dll">
          <script name="OutlierAccuracy_Script.txt">
            <references>
            </references>
          </script>
        </implementation>
      </interface>
      <interface name="IScriptableSample">
        <implementation assembly="Sample_Script.dll">
          <script name="Sample_Script.txt">
            <references>
            </references>
          </script>
        </implementation>
      </interface>
      <interface name="IScriptableCompound">
        <implementation assembly="Compound_Script.dll">
          <script name="Compound_Script.txt">
            <references>
            </references>
          </script>
        </implementation>
      </interface>
      <interface name="IScriptableQuantifierIon">
        <implementation assembly="QuantifierIon_Script.dll">
          <script name="QuantifierIon_Script.txt">
            <references>
            </references>
          </script>
        </implementation>
      </interface>
      <interface name="IScriptableQualifierIon">
        <implementation assembly="QualifierIon_Script.dll">
          <script name="QualifierIon_Script.txt">
            <references>
            </references>
          </script>
        </implementation>
      </interface>
    </components>

  •  ,

    I am not familiar with those types of modifications to quant. In general, the error for not finding csc.exe means that the expected version of .NET Framework has not been found or that the installation for that version is corrupt and missing csc.exe or other necessary files. The file csc.exe should be located in C:\Windows\Microsoft.NET\Framework or \Framework64 and then the .NET version. Not all version folders will contain csc.exe.

    I would try repairing or verifying your .NET installations. Microsoft provides tools for doing this.

    If that doesn't work, then for troubleshooting purposes, I would suggest temporarily changing the Windows language to English (United States) just to rule out any issues that might cause with quant. It is unlikely, but language settings have caused unexpected issues and it should be quick enough to try and rule this out.

    Finally, I would suggest seeking assistance from whoever provided the custom .dlls and component.config.xml file. If this was provided with assistance from Agilent, then I would suggest reaching out to your local support team or the contact that provided the files. Contact Us | Agilent

Reply
  •  ,

    I am not familiar with those types of modifications to quant. In general, the error for not finding csc.exe means that the expected version of .NET Framework has not been found or that the installation for that version is corrupt and missing csc.exe or other necessary files. The file csc.exe should be located in C:\Windows\Microsoft.NET\Framework or \Framework64 and then the .NET version. Not all version folders will contain csc.exe.

    I would try repairing or verifying your .NET installations. Microsoft provides tools for doing this.

    If that doesn't work, then for troubleshooting purposes, I would suggest temporarily changing the Windows language to English (United States) just to rule out any issues that might cause with quant. It is unlikely, but language settings have caused unexpected issues and it should be quick enough to try and rule this out.

    Finally, I would suggest seeking assistance from whoever provided the custom .dlls and component.config.xml file. If this was provided with assistance from Agilent, then I would suggest reaching out to your local support team or the contact that provided the files. Contact Us | Agilent