upgrade to MH 11.1

Hello,

i upgraded my MH QUANT 10.1 to 11.1 and since then the batch.script doesn't work for me, we are using a modified component.config, do you know where the error could be? This component.config worked with all 10 versions of masshunter (10,10.1 and 10.2)

Thank you.

Parents
  • This is written in log..> [ERROR] ComponentManager: Script processing error: Compiler executable csc.exe not found.

  • Hello  ,

    I'm not aware of this error specifically in regard to Quant or MassHunter, but generally this type of error points to either a corrupted or missing Microsoft .NET framework component. Make certain both the 3.5 and 4.x versions of .NET Framework are enabled in your Windows Features section or try repairing them. 

  • Hello,

    we have latest framework 4.8 and framework 3.5 SP1, it works fine with MH 10.1.

    <?xml version="1.0" encoding="utf-8"?>
    <components>
      <defaultnamespace>Agilent.MassSpectrometry.DataAnalysis.Quantitative</defaultnamespace>
      <references>
        <reference framework="true" assembly="System.Data.dll"/>
        <reference framework="true" assembly="System.Data.DataSetExtensions.dll"/>
        <reference framework="true" assembly="System.Xml.dll"/>
        <reference framework="true" assembly="System.Web.Services.dll"/>
        <reference framework="false" assembly="Quantitation.dll"/>
        <reference framework="false" assembly="QuantUtils.dll"/>
        <reference framework="false" assembly="ScriptableInterfaces.dll"/>
        <reference framework="true" assembly="System.Linq.dll"/>
        <reference framework="true" assembly="System.Core.dll"/>
        <reference framework="true" assembly="System.Xml.Linq.dll"/>
        <reference framework="true" assembly="System.Windows.Forms.dll"/>
        <reference framework="true" assembly="System.Threading.Tasks.dll"/>
        <reference framework="true" assembly="System.Drawing.dll"/>
        <reference framework="false" assembly="C:\Program Files\Agilent\MassHunter\Workstation\Quant\bin\Als.OData.dll"/>
        <reference framework="false" assembly="C:\Program Files\Agilent\MassHunter\Workstation\Quant\bin\Microsoft.OData.Client.dll"/>
        <reference framework="false" assembly="C:\Program Files\Agilent\MassHunter\Workstation\Quant\bin\Microsoft.OData.Core.dll"/>
        <reference framework="false" assembly="C:\Program Files\Agilent\MassHunter\Workstation\Quant\bin\Microsoft.OData.Edm.dll"/>
        <reference framework="false" assembly="C:\Program Files\Agilent\MassHunter\Workstation\Quant\bin\Microsoft.Spatial.dll"/>
        <reference framework="true" assembly="System.Net.dll"/>
        <reference framework="true" assembly="System.IO.dll"/>
      </references>
      <interface name="IScriptableSample">
        <implementation assembly="CompoundMath_Script.dll">
          <script name="CompoundMath_Script.txt">
            <references>
              <reference framework="false" assembly="BatchAnalysis.dll"/>
            </references>
          </script>
        </implementation>
      </interface>
      <interface name="ICustomExpressions">
        <implementation assembly="C:\MassHunter\Scripts\Quant\IScriptable\CustomExpressions_Script.dll">
          <script name="C:\MassHunter\Scripts\Quant\IScriptable\CustomExpressions_Script.cs">
            <references>
            </references>
          </script>
        </implementation>
      </interface>
      <!--<interface name="IScriptableBatch">
        <implementation assembly="C:\MassHunter\Scripts\Quant\IScriptable\Batch_Script.dll">
          <script name="C:\MassHunter\Scripts\Quant\IScriptable\Batch_Script.txt">
            <references>
              <reference framework="false" assembly="BatchAnalysis.dll"/>
            </references>
          </script>
        </implementation>
      </interface>-->
      <interface name="IScriptableBatch">
        <implementation assembly="C:\MassHunter\Scripts\Quant\IScriptable\Batch_Script.dll">
          <script name="C:\MassHunter\Scripts\Quant\IScriptable\Batch_Script.txt">
            <references>
              <reference framework="false" assembly="BatchAnalysis.dll"/>
              <reference framework="true" assembly="System.Data.dll"/>
              <reference framework="true" assembly="System.Linq.dll"/>
              <!--<reference framework="true" assembly="System.Linq.Enumerable.dll"/>-->
              <reference framework="true" assembly="System.Data.DataSetExtensions.dll"/>
              <reference framework="true" assembly="System.Collections.dll"/>
              <reference framework="true" assembly="System.Xml.Linq.dll"/>
              <reference framework="true" assembly="System.IO.dll"/>
              <reference framework="true" assembly="System.Xml.dll"/>
              <reference framework="true" assembly="System.Core.dll"/>
              <reference framework="true" assembly="System.Threading.Tasks.dll"/>
              <reference framework="true" assembly="System.Net.dll"/>
            </references>
          </script>
        </implementation>
      </interface>
      <interface name="IScriptableOutlier" instanceName="Accuracy">
        <implementation assembly="OutlierAccuracy_Script.dll">
          <script name="OutlierAccuracy_Script.txt">
            <references>
            </references>
          </script>
        </implementation>
      </interface>
      <interface name="IScriptableSample">
        <implementation assembly="Sample_Script.dll">
          <script name="Sample_Script.txt">
            <references>
            </references>
          </script>
        </implementation>
      </interface>
      <interface name="IScriptableCompound">
        <implementation assembly="Compound_Script.dll">
          <script name="Compound_Script.txt">
            <references>
            </references>
          </script>
        </implementation>
      </interface>
      <interface name="IScriptableQuantifierIon">
        <implementation assembly="QuantifierIon_Script.dll">
          <script name="QuantifierIon_Script.txt">
            <references>
            </references>
          </script>
        </implementation>
      </interface>
      <interface name="IScriptableQualifierIon">
        <implementation assembly="QualifierIon_Script.dll">
          <script name="QualifierIon_Script.txt">
            <references>
            </references>
          </script>
        </implementation>
      </interface>
    </components>

Reply
  • Hello,

    we have latest framework 4.8 and framework 3.5 SP1, it works fine with MH 10.1.

    <?xml version="1.0" encoding="utf-8"?>
    <components>
      <defaultnamespace>Agilent.MassSpectrometry.DataAnalysis.Quantitative</defaultnamespace>
      <references>
        <reference framework="true" assembly="System.Data.dll"/>
        <reference framework="true" assembly="System.Data.DataSetExtensions.dll"/>
        <reference framework="true" assembly="System.Xml.dll"/>
        <reference framework="true" assembly="System.Web.Services.dll"/>
        <reference framework="false" assembly="Quantitation.dll"/>
        <reference framework="false" assembly="QuantUtils.dll"/>
        <reference framework="false" assembly="ScriptableInterfaces.dll"/>
        <reference framework="true" assembly="System.Linq.dll"/>
        <reference framework="true" assembly="System.Core.dll"/>
        <reference framework="true" assembly="System.Xml.Linq.dll"/>
        <reference framework="true" assembly="System.Windows.Forms.dll"/>
        <reference framework="true" assembly="System.Threading.Tasks.dll"/>
        <reference framework="true" assembly="System.Drawing.dll"/>
        <reference framework="false" assembly="C:\Program Files\Agilent\MassHunter\Workstation\Quant\bin\Als.OData.dll"/>
        <reference framework="false" assembly="C:\Program Files\Agilent\MassHunter\Workstation\Quant\bin\Microsoft.OData.Client.dll"/>
        <reference framework="false" assembly="C:\Program Files\Agilent\MassHunter\Workstation\Quant\bin\Microsoft.OData.Core.dll"/>
        <reference framework="false" assembly="C:\Program Files\Agilent\MassHunter\Workstation\Quant\bin\Microsoft.OData.Edm.dll"/>
        <reference framework="false" assembly="C:\Program Files\Agilent\MassHunter\Workstation\Quant\bin\Microsoft.Spatial.dll"/>
        <reference framework="true" assembly="System.Net.dll"/>
        <reference framework="true" assembly="System.IO.dll"/>
      </references>
      <interface name="IScriptableSample">
        <implementation assembly="CompoundMath_Script.dll">
          <script name="CompoundMath_Script.txt">
            <references>
              <reference framework="false" assembly="BatchAnalysis.dll"/>
            </references>
          </script>
        </implementation>
      </interface>
      <interface name="ICustomExpressions">
        <implementation assembly="C:\MassHunter\Scripts\Quant\IScriptable\CustomExpressions_Script.dll">
          <script name="C:\MassHunter\Scripts\Quant\IScriptable\CustomExpressions_Script.cs">
            <references>
            </references>
          </script>
        </implementation>
      </interface>
      <!--<interface name="IScriptableBatch">
        <implementation assembly="C:\MassHunter\Scripts\Quant\IScriptable\Batch_Script.dll">
          <script name="C:\MassHunter\Scripts\Quant\IScriptable\Batch_Script.txt">
            <references>
              <reference framework="false" assembly="BatchAnalysis.dll"/>
            </references>
          </script>
        </implementation>
      </interface>-->
      <interface name="IScriptableBatch">
        <implementation assembly="C:\MassHunter\Scripts\Quant\IScriptable\Batch_Script.dll">
          <script name="C:\MassHunter\Scripts\Quant\IScriptable\Batch_Script.txt">
            <references>
              <reference framework="false" assembly="BatchAnalysis.dll"/>
              <reference framework="true" assembly="System.Data.dll"/>
              <reference framework="true" assembly="System.Linq.dll"/>
              <!--<reference framework="true" assembly="System.Linq.Enumerable.dll"/>-->
              <reference framework="true" assembly="System.Data.DataSetExtensions.dll"/>
              <reference framework="true" assembly="System.Collections.dll"/>
              <reference framework="true" assembly="System.Xml.Linq.dll"/>
              <reference framework="true" assembly="System.IO.dll"/>
              <reference framework="true" assembly="System.Xml.dll"/>
              <reference framework="true" assembly="System.Core.dll"/>
              <reference framework="true" assembly="System.Threading.Tasks.dll"/>
              <reference framework="true" assembly="System.Net.dll"/>
            </references>
          </script>
        </implementation>
      </interface>
      <interface name="IScriptableOutlier" instanceName="Accuracy">
        <implementation assembly="OutlierAccuracy_Script.dll">
          <script name="OutlierAccuracy_Script.txt">
            <references>
            </references>
          </script>
        </implementation>
      </interface>
      <interface name="IScriptableSample">
        <implementation assembly="Sample_Script.dll">
          <script name="Sample_Script.txt">
            <references>
            </references>
          </script>
        </implementation>
      </interface>
      <interface name="IScriptableCompound">
        <implementation assembly="Compound_Script.dll">
          <script name="Compound_Script.txt">
            <references>
            </references>
          </script>
        </implementation>
      </interface>
      <interface name="IScriptableQuantifierIon">
        <implementation assembly="QuantifierIon_Script.dll">
          <script name="QuantifierIon_Script.txt">
            <references>
            </references>
          </script>
        </implementation>
      </interface>
      <interface name="IScriptableQualifierIon">
        <implementation assembly="QualifierIon_Script.dll">
          <script name="QualifierIon_Script.txt">
            <references>
            </references>
          </script>
        </implementation>
      </interface>
    </components>